Mammogram Introduction

Mammogram Preparation

Breast Cancer Overview




Author: William Teh, MB, ChB, FRCR, Clinical Director, Department of Radiology, Northwick Park Hospital, UK

William Teh is a member of the following medical societies: British Institute of Radiology, British Medical Association, and Royal College of Radiologists

Coauthor(s): Hemant Singhal, MD, MBBS, FRCSE, FRCS(C), Senior Lecturer, Department of Surgery, Imperial College School of Medicine, UK; Consultant Surgeon, Northwick Park and St Marks Hospitals, UK

Editors: John M Lewin, MD, Section Chief, Breast Imaging, Diversified Radiology of Colorado, PC; Associate Clinical Professor, Department of Preventative Medicine and Biometrics, University of Colorado Denver; Bernard D Coombs, MB, ChB, PhD, Consulting Staff, Department of Specialist Rehabilitation Services, Hutt Valley District Health Board, New Zealand; Edward Azavedo, MD, PhD, Director of Clinical Breast Imaging Services, Associate Professor, Department of Radiology, Karolinska University Hospital, Sweden; Robert M Krasny, MD, Consulting Staff, Department of Radiology, The Angeles Clinic and Research Institute; Lawrence M Davis, MD, Assistant Professor of Diagnostic Imaging (Clinical), Department of Diagnostic Imaging, Warren Alpert Medical School at Brown University

Author and Editor Disclosure

Synonyms and related keywords: wire localization, guidewire localization, localization biopsy, hookwire breast localization, hook-wire breast localization

Background

This article discusses the techniques applicable to the localization of breast lesions. As breast cancer screening with mammography increases, many impalpable breast lesions are being detected. These lesions should first and foremost be definitively diagnosed by using image-guided needle biopsy. After needle biopsy, some of these lesions may require diagnostic or therapeutic surgical biopsy. If a malignant or indefinite diagnosis is obtained, surgical excision is indicated. This, in turn, requires accurate localization of the lesion.

Good accuracy of such localizations is required to ensure correct and adequate removal of the lesions and to minimize the degree of cosmetic disfigurement.1

For excellent patient education resources, visit eMedicine's Cancer and Tumors Center. Also, see eMedicine's patient education articles Mammogram and Breast Cancer.

See also the following related Medscape topics:
Resource Center Breast Cancer
CME/CE NCCN Clinical Practice Guidelines in Oncology Symposium: Breast Cancer
CME/CE Refining First-line Treatment for Metastatic Breast Cancer
CME/CE Treatment Options for Early-Stage Breast Cancer: Information for Primary Care Providers

Preferred Examination

The modality of choice depends on operator expertise and the type of lesion. Most palpable lesions do not require image-guided localization. Breast lesions that are visible with ultrasonography can be adequately localized by using a skin marker if the lesion lies fairly superficially within the breast.2 In addition, a surgeon can easily perform ultrasonographic localization in the operating room to identify the lesion and to confirm its adequate removal.3, 4

Other localization techniques include magnetic resonance imaging (MRI) and mammography (by using a grid, a holey plate, or a stereotactic method).5, 6 Although a number of techniques are used to localize nonpalpable breast lesions, needle localization is the most common.7, 8, 9

After successful placement, the patient is then taken to the operating room, where the surgeon commonly makes an incision at the site of the wire insertion and excises the wire and the lesion. Some surgeons place a stiff outer cannula along the wire down to the wire tip and excise the tip along with the lesion.

Impalpable lesions may also be marked by using an injection of methylene blue,10, 11 charcoal suspension,12 or radioisotopes.13

It is essential to obtain a diagnosis with a core biopsy under ultrasonographic or stereotactic guidance before surgery so that a definitive, 1-step surgical procedure can be planned. Once malignancy is diagnosed, any of the following procedures can be used to localize the lesion:

  • Skin-marker localization
  • Needle localization under ultrasonographic guidance
  • Needle localization under stereotactic guidance6
  • Radionuclide localization combined with sentinel-node biopsy11

See also the following related eMedicine topics:
Breast Cancer [Oncology]
Breast Cancer [Plastic Surgery]
Breast Cancer Evaluation
Breast Cancer, Mammography
Breast Cancer, Ultrasonography
Breast, Stereotactic Core Biopsy/Fine Needle Aspiration

Limitations of Techniques

All methods are subject to potential inaccuracies.14 It is important to confirm the correct placement of the guidewire. Ultrasonographic and mammographic wire localizations can be followed by mammography performed in 2 planes (eg, lateromedial and craniocaudal projections) to confirm that the correct lesion has been accurately targeted.

Ideally, the guidewire should transfix the lesion on both projections (see Image 1). The tip of the guidewire should ideally be placed within 1 cm of the target lesion.

Guidewires may be dislodged or may migrate prior to surgery. Complications, including pneumothoraces, have been described. Wire fragments may also be retained following surgery. Published literature show that the rate of needle-localization failure is in the range of 0-20%. Factors associated with failure include the following15, 16, 17:

  • Lesion type
  • Lesion size
  • Increased distance of the needle from the lesion
  • Decreased breast size
  • Decreased specimen volume



Findings

Techniques that involve mammography usually require the upright mammographic attachment on a normal mammographic unit, although localizations with stereotactic prone tables also have been described. Before stereotaxy came into use, a grid (see Image 2) or holey plate was used to calculate the position of needle placement in the X and Y planes. The depth was calculated from the lateromedial projection. The position was then checked according to the superimposition of target, hub, and shaft of the needle, and the required depth was verified on the orthogonal view (see Image 3).

Stereotaxy enables the exact position to be located. The needle is then placed 1 cm beyond the lesion to ensure that it is adequately transfixed. Because of the accordion effect (the thickness of the breast expands when compression is released), the needle tip may migrate, causing the needle to be placed short of the lesion.18 The final depth of the needle is therefore checked on the orthogonal view to ensure that the lesion is adequately transfixed.

Different needles exist, and most are introduced by using a stiffer, co-axial needle. Some needles are then removed, leaving the wire in situ. The wires commonly have a barb or hook that is deployed in the final position to anchor the wire in place. Some needles require the outer cannula to be retained in situ. The choice of needles and wires used are dictated by the preference of the radiologist and the surgeons.

After surgical removal of the lesion, specimen radiography must be performed to ensure that the lesion was adequately excised (see Image 4).



Findings

The use of MRI localization reflects the increasing use of magnetic resonance mammography. Although a portion of MRI-depicted lesions can subsequently be localized with ultrasonography, some of them are mammographically and ultrasonographically occult, which means that they require MRI-guided needle biopsy or localization.5, 19

All of the major manufacturers of MRI units have made biopsy attachments available; these require the patient to be kept in a semiprone position.

MRI is then performed with the intravenous administration of a gadolinium-based contrast agent, and the lesion is localized by using MRI-compatible localization needles. The correct placement of the needle can be confirmed by obtaining a T2-weighted MRI to verify the signal void to transfix the lesion.

Gadolinium-based contrast agents (gadopentetate dimeglumine [Magnevist], gadobenate dimeglumine [MultiHance], gadodiamide [Omniscan], gadoversetamide [OptiMARK], gadoteridol [ProHance]) have recently been linked to the development of nephrogenic systemic fibrosis (NSF) or nephrogenic fibrosing dermopathy (NFD). For more information, see the eMedicine topic Nephrogenic Fibrosing Dermopathy. The disease has occurred in patients with moderate to end-stage renal disease after being given a gadolinium-based contrast agent to enhance MRI or magnetic resonance angiography (MRA) scans.

As of late December 2006, the Food and Drug Administration (FDA) had received reports of 90 such cases. Worldwide, over 200 cases have been reported, according to the FDA. NSF/NFD is a debilitating and sometimes fatal disease. Characteristics include red or dark patches on the skin; burning, itching, swelling, hardening, and tightening of the skin; yellow spots on the whites of the eyes; joint stiffness with trouble moving or straightening the arms, hands, legs, or feet; pain deep in the hip bones or ribs; and muscle weakness. For more information, see the FDA Public Health Advisory or Medscape.

Technical difficulties and challenges include the tendency of the MRI-visible lesion to fade over time (because enhancement with contrast material is transient).5 In addition, because of the amount of time required to perform the procedure (30-60 min), the patient may begin to move, which could cause an error in needle placement. Nevertheless, MRI localizations have generally been as accurate as mammographic localizations, with miss rates of 2-9%. Some have suggested that postoperative MRI should be performed to verify complete excision of the lesion.



Findings

When the lesion is visible, ultrasonography is the modality of choice for needle placement. It provides the advantage of real-time imaging, allowing accurate placement of the needle. Being able to directly visualize the lesion and needle position results in a quicker procedure, reducing the risk of patient morbidity. As in mammographic techniques, the lesion must be transfixed, and orthogonal mammography can be used to confirm that the correct lesion has been localized.

The use of intraoperative ultrasonography by the surgeon in the operating theatre, as well as the employment of skin marking, is described in Preferred Examination.



Findings

Nuclear medicine study was originally used for sentinel lymph node biopsy. Colloidal albumin labeled with technetium-99m (99mTc) is injected directly into the lesion under stereotactic or ultrasonographic guidance.13 The accuracy of isotope placement is checked with scintigraphy. Excision biopsy is then performed by using a gamma probe. After excision, the excised lesion and the cavity are checked for radioactivity, and the specimen is radiographed to ensure the radiographic adequacy of the excision.

Measured doses to the breast and to the surgeon's hand appear to be negligible (0.03 ± 0.02 mGy/MBq and 7.5 ± 5.0 µSv/h, respectively).

Degree of Confidence

Results from a study of 67 consecutive patients showed complete removal of the lesion in 99.5% of them. Thus, this technique is accurate and at least comparable to conventional wire localization.20



The amount of tissue that is excised is dependent on the nature of the surgical procedure. If a diagnostic procedure is performed, a small volume of tissue is removed so that histologic diagnosis can be achieved with minimal scarring. When surgery is performed to treat malignant disease, the aim is to excise the lesion with a clear margin in order to minimize the risk of local recurrence.13



Media file 1:  Mammogram shows a spiculated mass to be transfixed by the guidewire.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  X-RAY

Media file 2:  Grid technique of localization.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  X-RAY

Media file 3:  Orthogonal (mediolateral) projection confirms the position of the needle to be placed beyond the cluster of microcalcification.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  X-RAY

Media file 4:  Specimen radiograph shows the wire and the localized speculated mass in situ, with a good excision margin.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  X-RAY

Media file 5:  Figure shows a poorly attenuating lesion being transfixed with a hook wire under ultrasonographic guidance.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  X-RAY

Media file 6:  Image shows the specimen radiograph with a stellate lesion containing clustered, pleomorphic microcalcification with wire in situ. Note the use of a stiff outer cannula.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  X-RAY



  1. Guidelines for surgeons in the management of symptomatic breast disease in the United Kingdom. Eur J Surg Oncol. Oct 1995;21 Suppl A:1-13. [Medline].
  2. Wilson M, Boggis CR, Mansel RE, et al. Non-invasive ultrasound localization of impalpable breast lesions. Clin Radiol. May 1993;47(5):337-8. [Medline].
  3. Fornage BD, Ross MI, Singletary SE. Localization of impalpable breast masses: value of sonography in the operating room and scanning of excised specimens. AJR Am J Roentgenol. Sep 1994;163(3):569-73. [Medline][Full Text].
  4. Fortunato L, Penteriani R, Farina M, et al. Intraoperative ultrasound is an effective and preferable technique to localize non-palpable breast tumors. Eur J Surg Oncol. Jan 11 2008;[Medline].
  5. Morris EA, Liberman L, Dershaw DD, et al. Preoperative MR imaging-guided needle localization of breast lesions. AJR Am J Roentgenol. May 2002;178(5):1211-20. [Medline][Full Text].
  6. Welch BL, Brem R, Black R, et al. Quality assurance procedure for a gamma guided stereotactic breast biopsy system. Phys Med. 2006;21 Suppl 1:102-5. [Medline].
  7. Silverstein MJ, Gamagami P, Rosser RJ, et al. Hooked-wire-directed breast biopsy and overpenetrated mammography. Cancer. Feb 15 1987;59(4):715-22. [Medline].
  8. Lannin DR, Grube B, Black DS, et al. Breast tattoos for planning surgery following neoadjuvant chemotherapy. Am J Surg. Oct 2007;194(4):518-20. [Medline].
  9. Haid A, Knauer M, Dunzinger S, et al. Intra-operative sonography: a valuable aid during breast-conserving surgery for occult breast cancer. Ann Surg Oncol. Nov 2007;14(11):3090-101. [Medline].
  10. Zografos GC, Doumitriou C, Lappas D, et al. Localization of nonpalpable breast lesions using hook-wire combined with isosulfan blue dye. J Surg Oncol. Jan 2003;82(1):73-4. [Medline].
  11. Varghese P, Abdel-Rahman AT, Akberali S, et al. Methylene blue dye--a safe and effective alternative for sentinel lymph node localization. Breast J. Jan-Feb 2008;14(1):61-7. [Medline].
  12. Canavese G, Catturich A, Vecchio C, et al. Pre-operative localization of non-palpable lesions in breast cancer by charcoal suspension. Eur J Surg Oncol. Feb 1995;21(1):47-9. [Medline].
  13. Luini A, Zurrida S, Paganelli G, et al. Comparison of radioguided excision with wire localization of occult breast lesions. Br J Surg. Apr 1999;86(4):522-5. [Medline].
  14. Hasselgren PO, Hummel RP, Georgian-Smith D, et al. Breast biopsy with needle localization: accuracy of specimen x-ray and management of missed lesions. Surgery. Oct 1993;114(4):836-40; discussion 840-2. [Medline].
  15. Jackman RJ, Marzoni FA Jr. Needle-localized breast biopsy: why do we fail?. Radiology. Sep 1997;204(3):677-84. [Medline][Full Text].
  16. Abrahamson PE, Dunlap LA, Amamoo MA, et al. Factors predicting successful needle-localized breast biopsy. Acad Radiol. Jun 2003;10(6):601-6. [Medline].
  17. Kouskos E, Gui GP, Mantas D, et al. Wire localisation biopsy of non-palpable breast lesions: reasons for unsuccessful excision. Eur J Gynaecol Oncol. 2006;27(3):262-6. [Medline].
  18. Guenin MA. Stereotactic needle localization. AJR Am J Roentgenol. Jan 2001;176(1):254-5. [Medline][Full Text].
  19. van den Bosch MA, Daniel BL, Pal S, et al. MRI-guided needle localization of suspicious breast lesions: results of a freehand technique. Eur Radiol. Aug 2006;16(8):1811-7. [Medline].
  20. Gennari R, Galimberti V, De Cicco C. Use of technetium-99m-labeled colloid albumin for preoperative and intraoperative localization of nonpalpable breast lesions. J Am Coll Surg. Jun 2000;190(6):692-8; discussion 698-9. [Medline].

Breast, Needle Localization excerpt

Article Last Updated: Feb 29, 2008