| Patient Education |
|
Click here for patient education.
|
|
AUTHOR AND EDITOR INFORMATION
Section 1 of 14
Author: Heather Tassone, DO, Department of Emergency Medicine, Albert Einstein School of Medicine, Jacobi and Montefiore Medical Centers, Bronx, NY
Coauthor(s):
Matthew A Silver, MD, Assistant Professor, Department of Emergency Medicine, Albert Einstein College of Medicine; Assistant Residency Director, Albert Einstein College of Medicine Jacobi/Montefiore Emergency Medicine Training Program; Consulting Staff, Department of Emergency Medicine, Montefiore Medical Center
Editors: Luis M Lovato, MD, Assistant Clinical Professor, David Geffen School of Medicine at UCLA; Director of Critical Care, Department of Emergency Medicine, Olive View/UCLA Medical Center; Mary L Windle, PharmD, Adjunct Assistant Professor, University of Nebraska Medical Center College of Pharmacy, Pharmacy Editor, eMedicine.com, Inc; Rick Kulkarni, MD, Medical Director, Assistant Professor of Surgery, Section of Emergency Medicine, Yale-New Haven Hospital
Author and Editor Disclosure
Synonyms and related keywords:
sural nerve, sural nerve block, regional nerve block, peripheral nerve block, local infiltration, local anesthesia, ankle block, ankle anesthesia, ankle injury, foot injury, foot injury repair
Physicians in the emergency department frequently encounter patients who have sustained trauma to the lower leg or foot and require anesthesia for repair. The regional sural nerve block allows for rapid anesthetization of the posterolateral calf and laterodorsal foot, including part of the dorsal fifth digit. Regional blocks have several advantages compared to local infiltration, such as fewer injections required to attain adequate anesthesia, smaller volume of anesthetic required, and less distortion of the wound site. Because of the lower number of injections, regional block is better tolerated by the patient and limits the chance of a needle stick injury to the provider. This procedure, often overlooked in the emergency department, is safe, is relatively easy to perform, and can provide excellent anesthesia to the foot and lower leg. In one study, regional anesthesia of the foot and ankle, when performed by surgeons, was completely successful 95% of the time.1 Because the sural nerve is relatively superficial, it is easily blocked at multiple levels at or above the ankle. In fact, because the nerve is so accessible, sural nerve biopsy specimens have been used to study inflammatory demyelinating peripheral neuropathies. (For information on skin biopsy as a diagnostic tool in peripheral neuropathy, please see this CME course on Medscape.)
Understanding the arborization of the sural nerve is crucial to a regional block of this nerve. The sural nerve has a contribution from both the tibial nerve and the common peroneal nerve, each of which originates from the sciatic nerve. The contribution from the tibial nerve is the medial sural cutaneous nerve; the common peroneal nerve’s contribution is the sural communicating branch. These two contributions come together to form the sural nerve, which arises in the popliteal fossa and courses superficially after piercing the deep fascia in the posterior calf.
 The sural nerve as it travels down the posterior calf.
The sural nerve continues down the posterior calf and supplies the skin of the posterolateral lower third of the lower leg. Entering the foot posterior to the lateral malleolus, this nerve supplies the lateral aspect of the foot, including the lateral fifth digit, via the lateral dorsal cutaneus nerve. It supplies the lateral heel via the lateral calcaneal branches.
 Sural nerve dermatome at the level of the posterior calf.
 Sural nerve dermatome at the level of the sole of the foot.
- Wound repair or exploration of the lateral posterior calf or dorsolateral fifth digit
- As part of an ankle block required to manipulate a fractured or dislocated ankle (A combination of posterior tibial, saphenous, superficial peroneal, deep peroneal, and sural nerve blocks results in complete block of sensory perception beneath the ankle.)
 Areas of anesthetization to complete an ankle block. This block requires anesthetization of 5 nerves for complete sensory block below the ankle. The areas to anesthetize include a line along the anterior ankle for the superficial peroneal nerve (blue line), the deep peroneal nerve (red star), the saphenous nerve (pink star), the sural nerve (green arrow), and the posterior tibial nerve (orange arrow). - Incision and drainage of an abscess in the lateral posterior calf or laterodorsal fifth digit
- Removal of a foreign body in the lateral posterior calf or dorsolateral fifth digit
- Allergy to anesthetic solution or additives (eg, ester, amide)
- Injection through infected tissue
- Severe bleeding disorder or coagulopathy
- Preexisting neurological damage
- Patient uncooperativeness (Pediatric or elderly patients may need sedation.)
The 2 main classes of local anesthetics currently in use are amino esters and amino amides. Both inhibit ionic fluxes required for the initiation and conduction of nerve impulses.2 Lidocaine, the most commonly used anesthetic, has a fast onset of action and a duration of action of 30-120 minutes, which is increased to 60-400 minutes with the addition of epinephrine. The total cumulative dose of lidocaine to be infiltrated is 5 mg/kg (not to exceed 300 mg) if lidocaine without epinephrine is used, and 7 mg/kg (not to exceed 500 mg) if lidocaine with epinephrine is used. Anesthetic preparations that contain epinephrine are commonly used in the emergency department. Epinephrine induces vasoconstriction, which decreases the amount of local bleeding at the site of injection. In addition, it increases the duration of action of the anesthetic with which it is combined. Despite these advantages, the vasoconstrictive properties of epinephrine may contribute to tissue hypoxia, and its use should be avoided in areas of poor perfusion (ie, fingers, toes, penis, ears, nose). For more information, see Local Anesthetic Agents, Infiltrative Administration. Topical anesthetics may be needed in children or uncooperative adults. For more information, see Anesthesia, Topical.
- Needle, 4 cm, 25 gauge (ga)
- Needle, 18 ga
- Syringe, 10 mL
- Marking pen
- Sterile gloves
- Antiseptic solution (eg, povidone iodine [Betadine] or chlorhexidine gluconate [Hibiclens]) with skin swabs
- Alcohol swabs
- Sterile drape
- Lidocaine 1%, 10 mL
- Facial mask with eye shield
- Sterile gauze
 Equipment needed for regional block.
- Position the patient prone, with the ankle elevated by a pillow or rolled sheet, optimizing comfort.
- Alternatively, the patient may sit or lay supine with the affected leg internally rotated and the ankle elevated on pillow or rolled sheet.
- Explain the procedure, benefits, risks, and complications to the patient and/or patient’s representative, and inform the patient of the possibility of paresthesia during the procedure.
- Obtain informed consent in accordance with hospital protocol.
- Perform and document neurovascular and musculoskeletal examinations prior to the procedure. Testing the sural nerve prior to block includes the following:
- Sensation of posterolateral calf
 Sural nerve dermatome at the level of the posterior calf. - Sensation of the lateral fifth digit
 Dermatome map of sural nerve contributions at the level of the anterior lower leg and anterior foot.
- Using nonsterile gloves, expose the area of injection and identify the landmarks. Locate the posterior border of the lateral malleolus and the Achilles tendon.
 Site adjacent to the Achilles tendon. - Mark the site just lateral to the Achilles tendon, between the 2 landmarks.
 Mark the site adjacent to the Achilles tendon.
 Marked site adjacent to the Achilles tendon. - Wipe the area with an alcohol pad and clean the site thoroughly with antiseptic solution, moving outward in a circular fashion.
 Betadine preparation. - Open a sterile drape and place the syringe, needle, and gauze on the tray, maintaining sterility.
- Put on sterile gloves. Attach the 18-ga needle to the 10-mL syringe and draw up the lidocaine. Then, change to the 25-ga needle.
- Place a skin wheal at the site marked. Advance the needle through the skin wheal, angling toward the lateral malleolus. Inject 5-7 mL of lidocaine in a transverse line until the lateral malleolus is reached.
 Advance the needle toward the lateral malleolus. - Calor and rubor of the foot from loss of sympathetic tone may be noted initially.
- Successful anesthesia of the areas noted heralds a successful sural nerve block.
- Equipment preparation and proper patient positioning may make the difference between success and failure.
- Be sure to not inject within the Achilles tendon.
- In children or noncompliant adults, consider the use of LET (lidocaine, epinephrine, tetracaine) topical preparation or EMLA cream.
- Pediatric or elderly patients may require additional sedation for compliance.
- Consider a hematoma block or bier block to attain more effective analgesia when a fracture exists or when more extensive manipulation of the foot is expected.
- Adding a buffering solution, like sodium bicarbonate, can significantly decrease the pain of the injection when performing a nerve block.3, 4 Add 1 mL of sodium bicarbonate (44 meQ/50 mL) to 9 mL of lidocaine.
- Warming the anesthetic solution to body temperature can significantly decrease the pain of the injection.4
- When unassisted, tape a bottle of lidocaine upside down to the wall before starting the procedure. If more anesthetic is needed during the procedure, it can be obtained from this bottle without compromising the sterility of gloves and equipment.
- Infection: Infection occurs when the puncture site is not clean. Avoid puncture through infected skin or skin lesion. Be sure to use sterile technique during the procedure, as the risk of infection is insignificant when sterility is properly maintained.
- Intravascular injection: Intra-arterial injection may result in vasospasm and lead to ischemia of the limb tissue. Intravenous injection can lead to systemic toxicity in high doses. Tissue texture changes revealing pallor, bogginess, and cool temperature may indicate that either intravascular injection or vascular compression has occurred. Always draw back on the plunger of the syringe prior to injection to ensure that intravascular placement has not taken place. Alpha-adrenergic antagonists (eg, phentolamine 0.5-5 mg diluted 1:1 with NaCl 0.9%) can be administered by local infiltration to relieve arterial vasospasm secondary to intra-arterial injection.4
- Nerve injury: Patients may develop paresthesia, sensory deficits, or motor deficits secondary to inflammation of the nerve. Most often, this type of neuritis is transient and resolves completely. During the procedure, pull back gently after inducing a paresthesia so as to not inject the nerve directly. Make sure to document a complete neuromuscular examination both before and after the procedure.
- Hemorrhage: Reports of significant hemorrhage during regional anesthesia are rare, even in patients with blood coagulopathies.3 A hematoma may develop with intravascular puncture. If prolonged bleeding occurs, apply direct pressure and elevate the limb.
- Allergic reaction: Allergic reactions to local anesthetics occur at a rate of 1%.5, 4 Reactions range from delayed hypersensitivity (type IV) to anaphylactic (type I). Although rare, the most common cause of such allergic reaction is the preservative in the local anesthetic solution. Preservative-free lidocaine (found in cardiac preparations) is an alternative, since it does not contain the preservative. Alternatively, a 1-2% diphenhydramine solution can be used as a local anesthetic.4
- Exceeding total volume of anesthesia: The dose of 1% lidocaine without epinephrine should not exceed 5 mg/kg. If lidocaine with epinephrine is used, total dose should not exceed 7 mg/kg. Systemic toxicity manifests in the central nervous and cardiovascular systems.5 Signs such as tremors, convulsions, tachycardia, or respiratory compromise should alert the physician to stop the procedure and reassess the patient.
New York School of Regional Anesthesia
American Society of Regional Anesthesia and Pain Medicine
| Media file 1:
Areas of anesthetization to complete an ankle block. This block requires anesthetization of 5 nerves for complete sensory block below the ankle. The areas to anesthetize include a line along the anterior ankle for the superficial peroneal nerve (blue line), the deep peroneal nerve (red star), the saphenous nerve (pink star), the sural nerve (green arrow), and the posterior tibial nerve (orange arrow). |
 | View Full Size Image | |
Media type: Photo
|
| Media file 3:
The sural nerve as it travels down the posterior calf. |
 | View Full Size Image | |
Media type: Illustration
|
| Media file 4:
Sural nerve dermatome at the level of the posterior calf. |
 | View Full Size Image | |
Media type: Illustration
|
| Media file 5:
Sural nerve dermatome at the level of the sole of the foot. |
 | View Full Size Image | |
Media type: Illustration
|
| Media file 6:
Dermatome map of sural nerve contributions at the level of the anterior lower leg and anterior foot. |
 | View Full Size Image | |
Media type: Illustration
|
| Media file 11:
Advance the needle toward the lateral malleolus. |
 | View Full Size Image | |
Media type: Photo
|
- Myerson MS, Ruland CM, Allon SM. Regional anesthesia for foot and ankle surgery. Foot Ankle. Jun 1992;13(5):282-8. [Medline].
- Norris RL Jr. Local anesthetics. Emerg Med Clin North Am. Nov 1992;10(4):707-18. [Medline].
- Reichman EF, Tolson DR. Regional nerve blocks (regional anesthesia). In: Reichman EF, Simon RR, eds. Emergency Medicine Procedures. New York, NY: McGraw-Hill; 2004:965-81.
- McGee D. Local and topical anesthesia. In: Roberts JR, Hedges JR, eds. Clinical Procedures in Emergency Medicine. 4th ed. Philadelphia, Pa: WB Saunders; 2004:541-5.
- Gmyrek R. Local anesthesia and regional nerve block anesthesia. eMedicine from WebMD [serial online]. February 7, 2007;Accessed September 6, 2006. Available at www.emedicine.com/derm/topic824.htm.
- Kelly JJ, Spektor M. Nerve blocks of the thorax and extremities. In: Roberts JR, Hedges JR, eds. Clinical Procedures in Emergency Medicine. 4th ed. Philadelphia, Pa: WB Saunders; 2004:584-9.
- Moore K. The lower limb. In: Clinically Oriented Anatomy. 3rd ed. Baltimore, Md: Williams and Wilkins; 1992:468-70.
- Netter FH. Lower limb. In: Atlas of Human Anatomy. 2nd ed. Teterboro, NJ: ICON Learning Systems; 1997:502-9.
- Paris PM, Yearly DM. Pain management. In: Marx JA, ed. Rosen's Emergency Medicine Concepts and Clinical Practice. 5th ed. St Louis, Mo: Mosby; 2002:2571-3.
- Raj PP. Practical Management of Pain. 3rd ed. St Louis, Mo: Mosby; 2000.
- Richardson EG. Surgical techniques. In: Canale ST, ed. Campbell's Operative Orthopaedics. 10th ed. St. Louis, Mo: Mosby; 2003:3911-4.
- Younger DS. Peripheral nerve disorders. Prim Care. Mar 2004;31(1):67-83. [Medline].
Nerve Block, Sural excerpt Article Last Updated: Feb 14, 2008 Topic originally published: Feb 14, 2008
|