AUTHOR AND EDITOR INFORMATION
Section 1 of 11
Author: P Syamasundar Rao, MD, Professor of Pediatrics and Medicine, University of Texas-Houston Medical School; Director, Division of Pediatric Cardiology, Children's Memorial Hermann Hospital; Professor of Pediatrics, MD Anderson Cancer Center, University of Texas
P Syamasundar Rao is a member of the following medical societies: American Academy of Pediatrics, American College of Cardiology, American Heart Association, American Medical Association, American Pediatric Society, Medical Association of Georgia, Society for Cardiac Angiography and Interventions, Society for Pediatric Research, Southern Society for Pediatric Research, and Western Society for Pediatric Research
Coauthor(s):
Kurt Pflieger, MD, Active Staff, Department of Pediatrics, Lake Pointe Medical Center
Editors: Jeffrey Towbin, MD, Associate Chair of Pediatric/Cardiology, Departments of Pediatrics, Molecular and Human Genetics, Cardiovascular, Professor, Baylor College of Medicine and Texas Children's Hospital; Mary L Windle, PharmD, Adjunct Assistant Professor, University of Nebraska Medical Center College of Pharmacy, Pharmacy Editor, eMedicine.com, Inc; John W Moore, MD, MPH, Professor of Clinical Pediatrics, Division of Pediatric Cardiology, Mattel Children's Hospital of University of California at Los Angeles; Gilbert Herzberg, MD, Assistant Professor, Department of Pediatrics, Section of Pediatric Cardiology, New York Medical College; Stuart Berger, MD, Professor of Pediatrics, Division of Cardiology, Medical College of Wisconsin; Chief of Pediatric Cardiology, Medical Director of Pediatric Heart Transplant Program, Medical Director of The Heart Center, Children's Hospital of Wisconsin
Author and Editor Disclosure
Synonyms and related keywords:
valvar pulmonary stenosis, pulmonary artery, right ventricle, valvular pulmonary stenosis, valvate pulmonary stenosis, pulmonary stenosis with a normal aortic root, pulmonary stenosis with an intact ventricular septum, isolated pulmonary valve stenosis, ventricular septal defect
Background
Diseases of the pulmonary valve are most often congenital, and only rarely do acquired disorders such as carcinoid and rheumatic fever affect the pulmonary valve. The pulmonary valve may be stenotic or atretic, or the leaflets of the valve may be absent. Pulmonary stenosis may be valvar, supravalvar, or subvalvar (infundibular or mid cavity, ie, double-chamber right ventricle); it may also be in the branch pulmonary arteries. These lesions are collectively associated with obstruction of the right ventricular outflow tract. In this article, only valvar pulmonary stenosis is reviewed.
A stenotic pulmonary valve may occur without associated congenital abnormalities, though it is most often associated with other structural abnormalities of the heart. To distinguish the former from the latter, terms such as pulmonary stenosis with a normal aortic root or pulmonary stenosis with an intact ventricular septum have been used. However, the term isolated pulmonary valve stenosis is preferred. The term isolated may be used even when a patent foramen ovale or a small atrial septal defect is present. The term congenital need not be used because most are congenital.
Pathologic Anatomy: Pathologic features of the stenotic pulmonary valve vary. The most common pathology is a dome-shaped pulmonary valve. The fused leaflets of the pulmonary valve protrude from their attachment into the pulmonary artery as a conical, windsock-like structure. The size of the pulmonary valve orifice varies from a pinhole to several millimeters. The orifice is most usually central, but can be eccentric. Raphe, presumably fused commissures of the valve, extend from the stenotic orifice a variable distance down into the base of the dome-shaped valve. The number of the raphe may vary from 0-7. Relatively uncommon variants are unicommissural, bicuspid, and tricuspid valves. The valve annulus is abnormal in most cases, and the fibrous back bone is partial or complete lacking; therefore, a true annulus may not be present.
Hypoplasia of the pulmonary valve ring and dysplastic pulmonary valves may be present in a few of patients. Pulmonary valve dysplasia is characterized by thickened, nodular, and redundant valvular leaflets with minimal or no commissural fusion; hypoplasia of the valvular ring; and lack of poststenotic dilatation of the pulmonary artery. The obstruction is mainly related to thickened, myxomatous, immobile pulmonary valve cusps and hypoplasia of the valve ring.
Changes secondary to pulmonary valve obstruction occur in the right ventricle and pulmonary artery. Hypertrophy of the right ventricular muscle is proportional to the degree (and perhaps the duration) of obstruction. The muscle hypertrophy is particularly prominent in the infundibular region and may become physiologically important; this appears to be related to the degree and duration of obstruction. There is mild dilatation of the right ventricular cavity. In extremely severe or critical obstruction, the right ventricular cavity may be markedly dilated. In rare cases, the right ventricle may be hypoplastic.
The main pulmonary artery is dilated in almost all cases. This dilatation is independent of the severity of the pulmonary valve obstruction and presumably related to a high-velocity jet across the stenotic valve. As noted above, such poststenotic dilatation is remarkably absent in patients with dysplastic pulmonary valves.
An interatrial communication, a patent foramen ovale or an atrial septal defect may be present and may be the seat for right-to-left shunt in patients with severe or long-standing pulmonary stenosis.
Pulmonary valve stenosis is the most common cardiac lesion in Noonan syndrome (phenotypically Turner syndrome and genotypically normal [XX or XY]).
Supravalvar pulmonary stenosis is often associated with rubella syndrome and Williams syndrome (ie, elfin facies, supravalvar aortic stenosis, and hypercalcemia with or without mental retardation).
Isolated infundibular or subvalvar pulmonary stenosis is uncommon and usually associated with a ventricular septal defect (VSD), such as in tetralogy of Fallot.
Peripheral pulmonary stenosis is frequently observed in newborns. It is related to relative narrowing of the branch pulmonary arteries and the acute angle of the origin of the branch pulmonary arteries at this age.
Pathophysiology
Clinically significant narrowing of a valve or a blood vessel increases pressure proximal to the obstruction. This pressure gradient is necessary to maintain flow across the stenotic site. In pulmonic stenosis, hypertrophy of the right ventricle ensues and maintains this forward flow. The magnitude of right ventricular pressure and the pressure gradient across the pulmonary valve are generally proportional to the degree of obstruction. Under usual circumstances, proportional right ventricular hypertrophy maintains normal pulmonary blood flow. If the normal output is not maintained, right-sided heart failure ensues. This occurs in neonates with critical pulmonary stenosis and in patients with severe obstruction that occurs in childhood or adulthood.
Changes in the geometry of the left ventricle and decreased left ventricular function can also occur. The changes are proportional to the degree of right ventricular hypertrophy; however, they revert to normal after obstruction of the right ventricular outflow tract is relieved.
With increasing right ventricular hypertrophy, right ventricular compliance decreases with a resultant increase in end-diastolic pressure and with prominent a waves in the right atrium. As right atrial pressure rises, a right-to-left shunt may occur if the foramen ovale is patent or if an atrial septal defect is present; this change results in systemic arterial desaturation and clinically discernible cyanosis. Such a right-to-left shunt can also occur in patients with an underdeveloped (hypoplastic) right ventricle.
Frequency
United States
Pulmonary stenosis represents 8-12% of all congenital heart defects in children. In adults, the prevalence about 15% of all congenital heart defects. Isolated valvar pulmonary stenosis with an intact ventricular septum is the second most common congenital cardiac defect in children. It may occur in as many as 50% of all patients with congenital heart disease associated with other congenital cardiac lesions.
Mortality/Morbidity
The severity of the valvar dysplasia determines its related morbidity and mortality.
- Mild-to-moderate valvar pulmonary stenosis is extremely well tolerated.
- Severe pulmonary stenosis can be associated with decreased cardiac output, right ventricular hypertrophy, early congestive heart failure (CHF), and cyanosis. The morbidity formerly associated with surgery is lessened with balloon pulmonary valvuloplasty.
Sex
The male-to-female ratio is 1:1.
Age
The patient's age at presentation is related to the severity of the obstruction. If the stenosis is severe, patients may present in the neonatal period or in infancy. Patients with mild obstruction may present in childhood with asymptomatic murmurs.
History
- Most children with pulmonic stenosis, particularly those with trivial and mild pulmonary stenosis, present with asymptomatic cardiac murmurs that are detected on routine examination.
- Patients with moderate or severe pulmonary stenosis may have mild exertional dyspnea. Adults may be asymptomatic irrespective of the severity of their obstruction.
- Patients with severe or critical obstruction may present with signs of systemic venous congestion, which are usually interpreted as signs of CHF. The signs are due to severe right ventricular dysfunction or to cyanosis secondary to a right-to-left shunt across a patent foramen ovale or an atrial septal defect.
- Lightheadedness, syncope, and chest pain that resembles angina pectoris are rare, even in patients with severe obstruction.
- Of note, many patients with moderate or severe pulmonary stenosis remain asymptomatic.
Physical
- Physical findings depend on the degree of obstruction.
- Most patients with pulmonary stenosis appear healthy and are well developed. Indeed, the chubby and rounded faces, described as moon facies, were initially thought to be characteristic for this anomaly, but this facial appearance is not a helpful diagnostic tool.
- Most patients with trivial, mild, or moderate stenosis and many with severe stenosis are acyanotic. However, some may have cyanosis secondary to an interatrial right-to-left shunt.
- The jugular venous pulse is normal. However, in patients with decreased right ventricular compliance (ie, severe stenosis), a prominent a wave may be visualized in the neck pulsation. Patients may have a concomitant presystolic pulsation in the liver as well.
- In patients with trivial or mild obstruction, the right ventricular impulse is normal. When the pulmonary stenosis is moderate to severe, a sustained and forceful right ventricular impulse and a right ventricular heave are felt.
- A thrill may be felt in the suprasternal notch and at the left upper sternal border (pulmonic area). The precordial thrill is most likely to be associated with severe obstruction, though no consistent relationship is observed between the thrill and the degree of obstruction.
- On auscultation, the first heart sound may be normal in intensity, or it may be loud. The second heart sound is split widely. The width of the split increases with worsening stenosis. The intensity of pulmonary component of the second heart sound may be loud (in mild stenosis), or it may be soft, diminished or absent depending on the severity of obstruction. A fourth heart sound may be heard at the left lower sternal border in patients with severe obstruction and is usually associated with prominent a wave in the jugular pulse.
- An ejection systolic click is heard along the left sternal border and varies with respiration (decreases or disappears during inspiration). With increasing severity, the click comes closer to the first heart sound.
- An ejection systolic murmur of grade II-VI to V-VI is heard best at the left upper sternal border with radiation into infraclavicular regions, axillae, or back. The intensity of the murmur is not necessarily related to the severity of pulmonary valve obstruction, but the duration and timing of peaking of the murmur are related to the severity of stenosis.
- An early diastolic decrescendo murmur of pulmonary regurgitation is not usually heard in the typical case of pulmonary stenosis. Previous surgical or balloon intervention or valvar calcification may result in such a murmur.
- A holosystolic murmur at the left lower sternal border, which indicates tricuspid regurgitation, may be audible in some patients with extremely severe pulmonary stenosis.
- Hepatosplenomegaly may develop in cases of CHF.
- Peripheral pulmonary stenosis (commonly encountered in the neonate) is usually associated with a grade II-VI systolic murmur that radiates into the posterior lung fields and axillae. The pathology of peripheral pulmonary stenosis is related to branch pulmonary arteries that are relatively small compared with the large main pulmonary artery, as well as to the acute angular takeoff of the branch pulmonary arteries from the main pulmonary artery specific to a neonate's anatomy. This condition and the associated murmur usually resolve spontaneously in the first month of life.
- Clinical assessment of severity
- The severity of the obstruction of the pulmonary valve can often be estimated by carefully analyzing the ausculatory findings. The timing of the ejection click, the extent of splitting of the second heart sound, the intensity of the pulmonary component of the second sound, the duration of the systolic murmur, and the timing of the peaking of the ejection murmur usually indicate the severity of pulmonary valve stenosis (see Image 1).
- With trivial and mild cases of pulmonary valve obstruction, the click is clearly separated from the first heart sound. Almost normal splitting of the second heart sound with normal or slightly increased pulmonary component of the second heart sound is heard. An ejection systolic, diamond-shaped murmur that peaks early in systole and that ends much before the aortic component of the second heart sound is appreciated.
- Findings in moderate pulmonary valve stenosis are an ejection systolic click that is closer to the first heart sound than it is in mild forms, a widely split second sound with a diminished pulmonary component, and an ejection systolic murmur that peaks in mid-to-late systole and that ends just before the aortic component of the second heart sound.
- In severe narrowing of the pulmonary valve, ausculatory features are an ejection systolic click that is absent or that occurs so close to the first heart sound that it becomes inseparable from it, markedly increased splitting with a soft or inaudible pulmonary component of the second heart sound, and a long ejection systolic murmur that peaks late in systole and that extends beyond the aortic component of the second heart sound so that the latter cannot be heard.
- The duration and time of peaking of the ejection systolic murmur, and not its intensity, indicate the severity of the pulmonary valve obstruction. The longer the murmur and the later it peaks, the more severe the obstruction. Likewise, the shorter the interval between the first heart sound and ejection click, the wider the splitting of the second heart sound. The softer the pulmonary component, the more severe the stenosis of the pulmonary valve.
Causes
Pulmonary valve stenosis is primarily due to maldevelopment of the pulmonary valve tissue and the distal portion of the bulbus cordis, which is characterized by fusion of leaflet commissures, resulting in a thickened and domed appearance of the valve.
Although familial forms of pulmonary stenosis are described, it is generally considered to be multifactorial in origin. Rates of recurrence in siblings are on the order of 2-3%. The prevalence of pulmonary stenosis in the offspring of a parent with pulmonary stenosis is 3.6%.
Aberrant flow patterns in utero also may be partly associated with maldevelopment of the pulmonary valve.
Aortic Stenosis, Valvar
Atrial Septal Defect, Ostium Secundum
Bundle Branch Block, Right
Double Outlet Right Ventricle, Normally Related Great Arteries
Holt-Oram Syndrome
Partial Anomalous Pulmonary Venous Connection
Pulmonary Stenosis, Infundibular
Tetralogy of Fallot With Pulmonary Atresia
Tetralogy of Fallot: Surgical Perspective
Ventricular Septal Defect, Supracristal
Ventricular Septal Defect: Surgical Perspective
Other Problems to be Considered
Complex congenital heart disease associated with findings of pulmonary stenosis Infundibular and/or subpulmonary stenosis Supravalvar pulmonary stenosis Double-chambered right ventricle Syndrome of absent pulmonary valves
Lab Studies
- Laboratory evaluation usually is not helpful.
- Hemoglobin and hematocrit measurements in cyanotic patients may be helpful in that they are increased in patients with right-to-left shunt.
- The degree of polycythemia is proportional to the degree and duration of right to left shunt.
- Microcytosis and hypochromia suggest iron deficiency and warrant treatment with iron supplement.
- Oximetry provides information of potential right-to-left shunting in borderline cyanotic lesions or in patients with anemia, but it does not help in identifying the cause of the shunt (pulmonary, interatrial, interventricular, great arterial).
- Although ABG analysis is usually not needed, one notable exception is the hyperoxia test in the newborn with cyanosis of undetermined origin.
- The fraction of inspired oxygen FIO2 with 100% oxygenation generally does not increase the partial pressure of oxygen to >100 mm Hg in patients with a cyanotic congenital heart defect (right-to-left intracardiac shunt).
Imaging Studies
- Chest radiography
- Chest radiographs demonstrate a prominent main pulmonary artery segment, but the size of the heart is usually normal (see Image 2).
- Pulmonary vascular markings are usually normal, but they may be decreased in severe pulmonary stenosis with associated right-to-left shunt.
- CHF appears as cardiomegaly with right ventricular and right atrial enlargement in severe valvar pulmonary stenosis, with or without tricuspid insufficiency.
- Echocardiography
- The sine qua non of diagnosis is 2-dimensional and Doppler echocardiography.
- A thickened pulmonary valve with restricted systolic motion (doming) in the parasternal short axis view is demonstrated.
- Multiple views are used to confirm the absence of coexistent congenital cardiac disease.
- Dilatation of the main pulmonary artery distal to the stenotic orifice commonly occurs.
- Markedly thickened, nodular, and immobile pulmonary valve leaflets may be recognized easily and suggest dysplastic pulmonary valves.
- The pulmonary valve annulus can also be visualized and measured. Measurements can be compared with normal values to determine if the annulus is hypoplastic. Such measurements are also useful in selecting the diameter of the balloon to used during balloon valvuloplasty.
- Pulsed, continuous-wave (or high-frequency pulsed), and color Doppler evaluation, in conjunction with 2-dimensional echocardiography, is most useful in confirming the clinical diagnosis and in quantitating the degree of obstruction.
- Pulsed Doppler interrogation of the right ventricular outflow tract with sample volume moved across the pulmonary valve demonstrates an abrupt increase in peak Doppler flow velocity, which suggests pulmonary valve obstruction. In addition, the flow pattern in the main pulmonary artery is turbulent instead of laminar. Color Doppler imaging also shows smooth, laminar subpulmonary flow (blue) and some flow acceleration (red) immediately beneath the pulmonary valve, with turbulent (mosaic) flow beginning immediately distal to the pulmonary valve leaflets.
- Doppler studies can be used to accurately determine the velocity of flow at single or multiple levels, which then can be converted to reproducible pressure gradients by applying the modified Bernoulli equation, as follows: pressure gradient (in millimeters of mercury) = 4 X (velocity in meters/second)2.
- The use of several views and measurements increases the accuracy of the predicted gradient of peak systolic pressure.
- Doppler study should be performed when the patient is quiet and in a resting state. Young children and patients who are extremely anxious may have to be mildly sedated.
- Severe pulmonary stenosis with gradients >50 mm Hg, as diagnosed by using a continuous-wave Doppler recording through the pulmonary valve, should be treated with balloon valvuloplasty or surgery. However, Doppler measurements represent peak instantaneous gradients, whereas catheterization gradients are peak-to-peak gradients; recognition of this concept is more important with aortic-valve gradients than with pulmonary-valve gradients. The peak instantaneous gradient was initially though to reflect the peak-to-peak systolic gradient measured during cardiac catheterization. However, this instantaneous gradient overestimates the peak-to-peak gradient, presumably because of a pressure-recovery phenomenon. In the authors' experience, the catheter peak-to-peak gradient is somewhere in between the Doppler peak instantaneous and mean gradients.
- Infundibular gradients secondary to severe right ventricular hypertrophy may be present, but these are not usually observed because severe obstruction of the distal pulmonary valve masks infundibular gradients. However, the infundibular gradients do appear after balloon pulmonary valvuloplasty. A triangular pattern of Doppler signal, similar to that described in subaortic obstruction, is characteristic of infundibular obstruction (see Image 3).
- Pulmonary insufficiency is easily seen during pulsed, continuous-wave, or color Doppler imaging, but is unlikely to be present without previous surgical or balloon pulmonary valvuloplasty.
- Color Doppler and pulsed Doppler interrogation of the atrial septum is useful and may reveal a left-to-right or right-to-left shunt. Because of high sensitivity of color Doppler imaging, contrast echocardiography is not routinely used to document right-to-left shunt.
- Most children with pulmonary stenosis do not require evaluation beyond echocardiography.
Other Tests
- Electrocardiography
- Results are usually normal in mild pulmonary stenosis.
- Right-axis deviation and right ventricular hypertrophy occur in moderate valvar pulmonary stenosis.
- The degree of right ventricular hypertrophy is well correlated with the severity of pulmonary stenosis.
- Right atrial hypertrophy and right ventricular hypertrophy with strain pattern are observed when pulmonary stenosis is severe.
- A superior QRS axis (left-axis deviation) is seen with dysplastic pulmonary valves and Noonan syndrome.
- Other noninvasive studies
- CT and MRImay demonstrate pulmonary valve stenosis, but the state-of-the-art echocardiographic Doppler studies are more useful that CT or MRI in diagnosing and quantitating pulmonary valve obstruction.
- Myocardial energy demands and perfusion may be evaluated by performing magnetic resonance spectroscopy and positron emission tomography (PET), respectively. However, the clinical utility of these techniques in the management of pulmonic stenosis has not been established.
Procedures
- Cardiac catheterization
- Catheterization is not indicated in mild pulmonary stenosis, but it is essential in severe stenosis. Catheterization is an integral part of balloon pulmonary valvuloplasty.
- This procedure is used to confirm the diagnosis; to discern the degree of obstruction; to assess the morphology of the right ventricle, pulmonary outflow tract, and pulmonary arteries; and to exclude other associated cardiac abnormalities.
- Patients with echocardiographic evidence of clinically significant pulmonary stenosis (50-60 mm Hg) should undergo diagnostic and therapeutic cardiac catheterization with preparation for balloon dilatation of the pulmonary valve.
- Monitoring of oxygen saturation
- Oxygen-saturation data usually do not show evidence of left-to-right shunts. A right-to-left shunt across the patent foramen ovale (or an atrial defect) may be present in moderate-to-severe obstruction of the pulmonary valve.
- Monitoring of pressures
- Right atrial pressure (the particularly a wave) may be increased.
- Right-ventricular peak systolic pressure is increased; the magnitude of the pressure is proportional to the degree of obstruction.
- The transpulmonary valve peak-to-peak gradient also indicate the severity of obstruction. A peak-to-peak gradient in excess of 50 mm Hg is usually considered an indication for therapeutic intervention.
- Angiography
- Right ventricular angiography usually reveals thickened and domed leaflets of the pulmonary valve with a thin jet of contrast material across the pulmonary valve (see Image 4).
- Enlargement of the right ventricle and a dilated segment of the main pulmonary artery are also seen.
- In patients with severe or long-standing pulmonary valve obstruction, infundibular constriction may be seen.
- Additional cineangiograms obtained at other locations are not necessary unless the echocardiographic and hemodynamic data suggest other abnormalities.
Medical Care
- Prehospital care
- Collect essential information about vital signs, including the patient's pulse, respiratory rate, work of breathing, and BP in the upper and lower extremities. Note the presence or absence of cyanosis.
- Associated congenital cardiac anomalies should be anticipated until proven otherwise.
- If the patient has a known large left-to-right shunt, such as patent ductus arteriosus (PDA) or VSD, and if the patient is in respiratory distress, diuresis should be attempted. Diuresis is effective in reducing the cyanosis secondary to pulmonary edema.
- Use of oxygen may reduce pulmonary artery pressure in patients with a reactive pulmonary vasculature, increasing pulmonary blood flow.
- Administer oxygen to any patient with cyanosis and respiratory distress.
- Emergency department care
- Limited diagnostic tests are needed after the structural diagnosis is made.
- Workup performed for the cyanotic infant with respiratory distress and hypotension or shock is often the same as that performed in a septic patient.
- If a neonate or young infant presents to the emergency room with severe cyanosis, ductal dependent lesions should be considered. Prostaglandin infusion may open the ductus and improve oxygen saturation.
- Therapeutic approach
- Patients with trivial (gradient <25 mm Hg) or mild (gradient <50 mm Hg) pulmonary stenosis do not need intervention to relieve the obstruction of the pulmonary valve. They should be clinically followed up at periodic intervals, perhaps on a yearly basis. During the period of rapid growth (infancy and adolescence), follow-up more frequent than this may be indicated. Routine well-child care, including immunizations by the primary physician, should be provided. Patients with pulmonary stenosis are candidates for infective endocarditis prophylaxis before they undergo any bacteremia-producing procedures and surgery, as indicated in the recommendations of the American Heart Association. Limitations in exercise or activity levels are not needed.
- Patients with moderate (gradient 50-79 mm Hg) and severe (gradient >80 mm Hg) obstruction should undergo intervention to relieve the stenosis of the pulmonary valve. After the obstruction is relieved, recommended routine care, endocarditis prophylaxis, and exercise limitations are the same as those described for trivial and mild stenosis.
- Patients with signs of right ventricular failure should be promptly treated with anticongestive measures, including digitalis and diuretics. However, the problem does not resolve until the obstruction is relieved. Therefore, prompt balloon or surgical intervention should be undertaken. Right ventricular function may not recover completely if intervention is withheld for too long and if myocardial damage sets in.
- A fetus with critical pulmonary stenosis or atresia with intact ventricular septum may benefit from pulmonary balloon valvuloplasty in utero, which promote growth of the right ventricle.
- Although the consensus is to offer relief of pulmonary valve obstruction in children with moderate or severe stenoses, this approach is somewhat controversial in adults because of reported lack of progression and the lack of complications in 1 group of adults followed up for 5-24 years in the 1970s. However, a prudent strategy may be to relieve pulmonary valve obstruction in adults with moderate-to-severe pulmonary stenosis, irrespective of their symptoms, because of the potential for myocardial damage associated with long-term pressure overload of the right ventricle, for generally lowered cardiac indices both before and after exercise in adults compared with children, and for exercise-induced hemodynamic abnormalities in adults.
Surgical Care
- Balloon pulmonary valvuloplasty
- Introduction: Rubio-Alverez et al first attempted to relieve pulmonary valve obstruction with transcatheter methods in the early 1950s. They used a ureteral catheter with a wire to cut open the stenotic pulmonary valve.
- In 1979, Semb et al used a balloon-tipped angiographic (Berman) catheter to rupture pulmonary valve commissures by rapidly withdrawing the inflated balloon across the valve.
- More recently, Kan et al applied the techniques of Dotter and Judkins and Gruntzig et al to relieve pulmonary valve obstruction by utilizing the radial forces of balloon inflation of a balloon catheter positioned across the pulmonic valve. This static balloon-dilation technique is currently performed worldwide to relieve.
- The general consensus based on current data is that balloon valvuloplasty is the treatment of choice for managing isolated pulmonary valve stenosis.
- Indications: In general, indications for balloon pulmonary valvuloplasty are similar to those used of surgical pulmonary valvotomy, ie, moderate pulmonary valve stenosis with a peak-to-peak gradient >50 mm Hg with a normal cardiac index. Some change this criterion to a gradient of 40 mm Hg or right ventricular pressure of 50 mm Hg. Careful evaluation of the available data suggests that (1) right ventricular pressure is only marginally reduced if mildly stenotic valves are dilated, (2) trivial and mild stenoses (gradient <50 mm Hg) are likely to remain mild at follow-up (as shown in natural-history studies), and (3) an increase in gradient is easily quantitated on follow-up Doppler echocardiography. If an increased gradient is documented, the patient can then undergo balloon dilatation. Given these observations, balloon dilation should be performed only in patients with peak-to-peak gradient >50 mm Hg.
- Technique: The technique of balloon pulmonary valvuloplasty involves positioning a balloon catheter (see Image 5) across the stenotic valve, usually over an extra-stiff exchange-length guide wire and inflating the balloon with diluted contrast material to accomplish valvotomy. The initially recommended balloon-to-annulus ratio was 1.2-1.4; subsequent data suggested a ratio of 1.2-1.25. When the pulmonary valve annulus is too large to dilate with a single balloon (about 20 mm), valvuloplasty with simultaneous inflation of 2 balloons across the pulmonary valve may be performed, though the current availability of large-diameter balloons make this technique unnecessary. The Inoue balloon has been used in adults with success. The major advantage of the Inoue balloon over conventional balloons its adjustable diameter that makes stepwise dilation possible.
- Mechanism of valvuloplasty: Inflation of a balloon placed across an obstructive lesion exerts radial forces on the stenotic lesion without any axial component.
- The mechanism of valvuloplasty is assessed by directly visualizing the valvar mechanism during surgery and postmortem examination and by indirectly means such as angiography and echocardiography. Splitting of the valve commissures and tearing and avulsion of the valve leaflets have been observed and are conceivably the mechanism by which balloon dilation relieves pulmonary valve obstruction. The circumferential dilating force that balloon inflation exerts is likely to rupture (tear) the weakest part of the valve mechanism. The fused commissures are the likely weakest links that can be broken with balloon dilation. However, when fused commissures are strong and cannot be torn, the valve cusps can be torn or the valve leaflet avulse. These events are likely to worsen pulmonary insufficiency.
- Pulmonary valve dysplasia, if severe, may preclude successful balloon valvuloplasty unless an associated commissural fusion is present.
- Immediate results: Results observed immediately after balloon valvuloplasty include reduced peak-to-peak gradients and right ventricular–to–left ventricular pressure ratios and an increased pulmonary artery pressures, jet widths and free motion of the pulmonary valve leaflets with decreased doming. Improvement of right ventricular function, tricuspid insufficiency, and right-to-left shunt, if present before dilation, are also documented.
- Infundibular stenosis: Infundibular gradients occur in nearly 30% patients. The older the patient's age and the greater the severity of obstruction, the greater the prevalence of infundibular reaction. When residual infundibular gradient is >50 mm Hg beta-blockade is generally recommended. Infundibular obstruction greatly regresses at follow-up (see Image 6), as demonstrated for infundibular reactions after surgical valvotomy. Rare patients require surgical intervention.
- Follow-up evaluation: Clinical, ECG, and Doppler echocardiographic evaluations are generally recommended at 1, 6 and 12 months after the procedure and yearly thereafter. Regression of right ventricular hypertrophy, as shown on ECG after balloon dilatation, is well documented. ECG is a useful adjunct in the evaluation of follow-up results. However, ECG evidence of hemodynamic improvement does not become apparent until 6 months after valvuloplasty. The Doppler gradient generally reflects residual obstruction and is a useful and reliable noninvasive monitoring tool.
- Intermediate-term results: At intermediate-term follow-up (usually <2 y), both catheterization-measured peak-to-peak and Doppler-measured peak instantaneous gradients remain improved as a whole. However, restenosis (gradient >50 mm Hg) is observed in nearly 10% patients.
- Predictors of restenosis include a balloon-to-annulus ratio of <1.2 and a gradient of >30 mm Hg after immediate valvuloplasty. In addition, early in one study, a small valve annulus, and a postsurgical or complex pulmonary stenosis were predictive of restenosis (McCrindle, 1994).
- Patients with restenosis have been successfully treated with re-dilatation with balloons larger than those used for initial balloon valvuloplasty. Re-dilatation is the procedure of choice for managing restenosis after previous balloon dilatation. However, if the pulmonary valve annulus is hypoplastic, if the pulmonary valve leaflets are dysplastic, or if clinically significant supravalvar pulmonary artery stenosis is the major reason for the restenosis, surgery is recommended.
- Long-term follow-up results: Although immediate and short-term results are been documented, data on long-term results are scarce. Published studies reveal generally low residual peak instantaneous Doppler gradients with minimal (1-2%) late recurrence of pulmonary stenosis (beyond that seen at intermediate follow-up) (Rao, 1998; Rao, 1999).
- In 1 study, approximately 5% of patients needed surgical intervention to relieve fixed subvalvar or supravalvar stenosis. Actuarial freedom for reintervention was 88% and 84%, respectively, at 5 and 10 years. Pulmonary valve insufficiency was noted in 80-90% patients but, right ventricular volume overloading did not develop, and none of the patients required surgical intervention because of pulmonary insufficiency. The authors concluded that balloon pulmonary valvuloplasty may continue to be the treatment of choice for moderate-to-severe valvar pulmonary stenosis.
- Ten- to 20-year follow-up studies to evaluate the clinical significance of residual pulmonary insufficiency should be undertaken.
- One report documented the development of clinically significant pulmonary insufficiency in 6 (6%) of 107 patients at late follow-up. Some of these patients required pulmonary valve replacement (Berman, 1999).
- Comparison with surgical valvotomy: Comparison of balloon therapy with surgical valvotomy has limitations, but the mortality and morbidity rates are generally increased after surgery. Greatest reduction of the gradient is observed after surgery, but the degree and frequency of pulmonary insufficiency may be higher after surgery than after balloon therapy.
- Critical pulmonary stenosis in the neonate: The term critical pulmonary stenosis with intact ventricular septum is applied to severe pulmonary valvar obstruction resulting in suprasystemic right ventricular systolic pressure with resultant tricuspid insufficiency, a right-to-left shunt across the atrial septum, and often a ductal-dependent pulmonary circulation.
- Although the surgical approach to relieve the obstruction with or without aorta-pulmonary shunt was standard treatment in the past, transcatheter treatment is now first-line therapy.
- Prostaglandin E1 is infused to augment pulmonary blood flow and improve systemic arterial desaturation, followed by cardiac catheterization and biplanar (sitting-up and lateral) right ventricular cineangiography. A right coronary artery, angled glide catheter, or cobra catheter (according to the operator's preference) is placed in the right ventricular outflow tract, and a floppy-tipped coronary guidewire is advanced across the pulmonary valve and into the descending aorta through the ductus. The catheter is advanced across the pulmonary valve into the descending aorta. The guidewire is then exchanged for a guidewire that is suited to position the balloon-dilation catheter. Then, balloon pulmonary valvuloplasty is performed in manner described earlier.
- In some cases, the balloon catheter cannot be advanced across a severely stenotic pulmonary valve, and balloon catheters 3-6 mm in diameter may be used for initial predilation and then replaced with larger, more appropriately sized balloon catheters.
- Although the results of this approach are reasonably good, the need for reintervention to address the complications associated with the procedure, residual obstruction, or associated defects is 25% in neonates compared with older children 8-10%.
- Other catheter interventions: A number other catheter interventions may become necessary in patients with pulmonary stenosis.
- Transcatheter occlusion of a PDA: Some patients with pulmonary stenosis may have a PDA of significant size. In such patients, transcatheter occlusion of the PDA performed with a coil (for small PDAs) or with an Amplatzer duct occluder (for medium or large PDAs) is recommended immediately after balloon pulmonary valvuloplasty. (See also the eMedicine article Patent Ductus Arteriosus.)
- Occlusion of a patent foramen ovale or atrial septal defect: A patent foramen ovale or atrial septal defect may occur in association with pulmonary stenosis. If these atrial defects do not spontaneously close during follow-up after balloon valvuloplasty, they may be closed with an Amplatzer septal occluder, if the criteria for their closure are met. Sometimes, the defects may need to be closed to prevent recurrent paradoxical embolisms. (See also the eMedicine article Atrial Septal Defect, General Concepts.)
- Balloon atrial septostomy: In neonates with a severely hypoplastic right ventricle, balloon atrial septostomy may be necessary to provide adequate egress to the systemic venous return. Indications are clinical signs of systemic venous congestion, restrictive patent foramen ovale on Doppler echocardiography, markedly elevated right atrial pressure with tall a waves, and/or a mean atrial pressure difference of >5 mm Hg.
- In neonates, Rashkind balloon septostomy is effective.
- Beyond the neonatal period, Park blade septostomy or surgical septostomy may be necessary. In patients with only mild or moderate right ventricular hypoplasia, balloon septostomy should be avoided so that forward flow through the right ventricle is encouraged with a consequent opportunity for its growth.
- Cardiac catheterization with balloon valvuloplasty: This is the preferred therapy for severe or critical valvar pulmonary stenosis. In neonates with critical valvar pulmonary stenosis, mortality rates related to balloon dilation are lower than those related to surgery mortality, and balloon dilation is the treatment of choice.
- Other surgical techniques: Since the first description of surgical relief of pulmonary stenosis by closed pulmonary valvotomy in the late 1940s, a variety of techniques were developed and include valvotomy with inflow occlusion, hypothermia and cardiopulmonary bypass. The currently preferred approach is transpulmonary arterial valvotomy under cardiopulmonary bypass.
- Results: Results of surgery are generally good with low mortality rate (3-7%) and decreased right ventricular pressures and pulmonary valve gradients. In a natural-history study in the United States, only 3% of 294 operated patients had gradients >50 mm Hg at 4-8 years after surgery (Nugent, 1977). The incidence of pulmonary insufficiency at follow-up was 90-90%. Despite these good results, transluminal balloon pulmonary valvuloplasty (discussed above) has replaced surgical pulmonary valvotomy.
- Indications: Surgery is reserved for cases in which balloon valvuloplasty is not feasible or not successful. One example is dysplastic pulmonary valve with valve ring hypoplasia. The treatment of this disorder is to excise the obstructive valve leaflets and enlarge the annulus by a transannular patch. Other examples are fixed infundibular and supravalvar stenosis after successful balloon valvuloplasty.
- Blalock-Taussig shunting: Patients with critical pulmonary stenosis and marked hypoplasia of the right ventricle may need a Blalock-Taussig shunt in addition to or instead of balloon or surgical valvotomy.
- Right ventricular repair: Right ventricular hypoplasia, as alluded to above, occurs in some patients with pulmonary stenosis, though this hypoplasia is most common in patients with pulmonary atresia and an intact ventricular septum. The right ventricle may enlarge after right ventricular outflow obstruction is relieved, particularly in tripartite right ventricles. If the right ventricle does not grow adequately to support the pulmonary circulation, 1.5 or single ventricular repair should be considered. In 1.5-ventricular repair, a bidirectional Glenn procedure (superior vena cava–to–right pulmonary artery anastomosis, end to side) is performed and the atrial septal defect closed to divert the blood from the upper part of the body directly into the pulmonary artery. The right ventricle is bypassed, and the blood in the inferior vena cava from the lower part of the body is allowed to flow to the lungs through the right ventricle. In the Fontan procedure, stagedtotalcavopulmonaryconnection is
performed. - See the eMedicine article Tricuspid Atresia.
Consultations
- Consultation with a pediatric cardiologist precedes consultation with a cardiothoracic surgeon.
- Patients with pulmonary valve atresia or a critical pulmonary stenosis with an inadequate right ventricle require a shunt (usually a modified Blalock-Taussig or central shunt) after the ductus arteriosus is pharmacologically kept patent with prostaglandin E1.
- Definitive repair may not be possible if the right ventricle is hypoplastic. Single ventricular palliation, such as the Fontan procedure, or variation, such as staged total cavopulmonary connection, may be required.
Activity
A prudent philosophy is to allow patients to limit their own activity according to personal tolerance. Restriction from highly exertional and competitive sports is recommended only for patients with severe pulmonary stenosis. After successful relief of the obstruction, normal activity may be resumed.
No medications are useful in isolated valvar pulmonary stenosis. Patients with CHF may benefit from anticongestive therapy. Patients with cyanosis may benefit from oxygen and prostaglandin E1. Patients with cyanosis due to a large right-to-left shunt require a definitive surgical procedure.
Patients with certain cardiac conditions, such as pulmonary stenosis, typically require antibiotic prophylaxis of endocarditis before they undergo procedures that may cause bacteremia. For more information, see Antibiotic Prophylactic Regimens for Endocarditis.
Drug Category: Prostaglandins
Alprostadil (Prostaglandin E1, PGE1) is used to treat ductal-dependent cyanotic congenital heart disease, which is caused by decreased pulmonary blood flow. Alprostadil acts as a smooth muscle relaxer and maintains patency of the ductus arteriosus when a cyanotic lesion (eg, critical pulmonary stenosis or atresia) or when an interrupted aortic arch occurs in a newborn. This drug is effective only in the neonatal period.
| Drug Name | Alprostadil (Prostin VR) |
| Description | First-line drug used as palliative therapy to temporarily maintain patency of ductus arteriosus before surgery. Produces vasodilation and increases cardiac output. Also inhibits platelet aggregation and stimulates intestinal and uterine smooth muscle. Used in suspected critical pulmonary stenosis when presentation includes cyanosis. Also used in ductal-dependent lesion (eg, pulmonary atresia variants, coarctation of aorta, interrupted aortic arch). Each 1-mL ampule contains 500 mcg/mL. |
| Pediatric Dose | 0.05-0.1 mcg/kg/min IV initially; after ductal dilatation achieved and O2 saturation improved, may gradually decrease to 0.02-0.025 mcg/kg/min if O2 saturations do not fall |
| Contraindications | Documented hypersensitivity; hyaline membrane disease, respiratory distress syndrome |
| Interactions | Data limited; caution with concurrent use of antiplatelet drugs or anticoagulants |
| Pregnancy | C - Safety for use during pregnancy has not been established.
|
| Precautions | Adverse effects and toxicity include apnea, seizures, fever, hypotension, leukocytosis, fever, and pulmonary overcirculation; neonates usually intubated prophylactically because of potential risk of apnea (10-12%); prolonged use occasionally necessary (in candidates for hypoplastic left heart syndrome transplantation) and may be associated with third spacing of fluid; monitor blood oxygenation and arterial pressure |
Drug Category: Beta-blockers
These drugs inhibit chronotropic, inotropic, and vasodilatory responses to beta-adrenergic stimulation.
| Drug Name | Atenolol (Tenormin) |
| Description | Used to treat hypertension. Selectively blocks beta1-receptors with little or no affect on beta2 types. Beta-blockers affect BP by means of several mechanisms, including negative chronotropic effect that decreases heart rate at rest and after exercise, negative inotropic effect that decreases cardiac output, reduction of sympathetic outflow from the CNS, and suppression of renin release from the kidneys. Used to improve and preserve hemodynamic status by acting on myocardial contractility, reducing congestion, and decreasing myocardial energy expenditure. Beta-blockers reduce inotropic state of left ventricle, decrease diastolic dysfunction, and increase LV compliance, reducing pressure gradient across LV outflow tract. Decreases myocardial oxygen consumption, reducing myocardial ischemia potential. Decreases heart rate, reducing myocardial oxygen consumption and reducing potential for myocardial ischemia. During IV administration, carefully monitor BP, heart rate, and ECG. |
| Adult Dose | 25-50 mg PO qd initially; may increase to 100 mg/d prn |
| Pediatric Dose | 0.8-1 mg/kg/d PO initially; may uptitrate according to symptoms; not to exceed 2 mg/kg/d |
| Contraindications | Documented hypersensitivity; CHF, pulmonary edema, cardiogenic shock, atrioventricular (AV) conduction abnormalities, and heart block (without pacemaker) |
| Interactions | Coadministration with aluminum salts, barbiturates, calcium salts, cholestyramine, nonsteroidal anti-inflammatory drugs (NSAIDs), penicillins, and rifampin may decrease effects; haloperidol, hydralazine, loop diuretics, and monoamine oxidase (MAO) inhibitors may increase toxicity |
| Pregnancy | D - Unsafe in pregnancy
|
| Precautions | Beta-blockade may reduce symptoms of acute hypoglycemia and mask signs of hyperthyroidism; abrupt withdrawal may exacerbate symptoms of hyperthyroidism and cause thyroid storm; monitor patients closely and withdraw slowly; during an IV, carefully monitor BP, heart rate, and ECG |
| Drug Name | Esmolol (Brevibloc) |
| Description | Ultra–short-acting that selectively blocks beta1-receptors with little or no effect on beta2-receptors. Particularly useful in elevated arterial pressure, especially if surgery planned. Reduced episodes of chest pain and clinical cardiac events compared with placebo. Can discontinue abruptly if necessary. Useful in patients at risk for complications from beta-blockade, particularly those with reactive airway disease, mild-to-moderate LV dysfunction, and/or peripheral vascular disease. Short half-life of 8 min allows for titration to desired effect and quick discontinuation if needed. |
| Adult Dose | Loading dose: 250-500 mcg/kg/min IV for 1 min, followed by maintenance infusion of 50 mcg/kg/min for 4 min; if adequate therapeutic effect (decreased HR and BP) not observed within 5 min, repeat loading dose and follow with maintenance infusion of 100 mcg/kg/min for 4 min; may repeat sequence q5-10min, increasing maintenance infusion by 50 mcg/kg/min with each sequence; not to exceed 200 mcg/kg/min |
| Pediatric Dose | 100-500 mcg/kg IV administered over 1 min initially; followed by continuous IV infusion titrated to 25-100 mcg/kg/min; continue titration according to BP or HR response; doses exceeding 200 mcg/kg/min rarely necessary |
| Contraindications | Documented hypersensitivity; uncompensated CHF, bradycardia, cardiogenic shock, and AV conduction abnormalities |
| Interactions | Aluminum salts, barbiturates, NSAIDs, penicillins, calcium salts, cholestyramine, and rifampin may decrease bioavailability and plasma levels of esmolol, possibly decreasing pharmacologic effect; cardiotoxicity may increase with concurrent sparfloxacin, astemizole, calcium channel blockers, quinidine, flecainide, or contraceptives; toxicity increases with concurrent digoxin, flecainide, acetaminophen, clonidine, epinephrine, nifedipine, prazosin, haloperidol, phenothiazines, or catecholamine-depleting agents |
| Pregnancy | C - Safety for use during pregnancy has not been established.
|
| Precautions | Beta-blockers may mask signs and symptoms of acute hypoglycemia and clinical signs of hyperthyroidism; symptoms of hyperthyroidism, including thyroid storm may worsen, when abruptly withdrawn; withdraw slowly and monitor patient closely |
| Drug Name | Propranolol (Inderal) |
| Description | Nonselective beta-blocker. Lipophilic (penetrates CNS). Has membrane-stabilizing activity and decreases automaticity of contractions. Also has class II antiarrhythmic properties. |
| Adult Dose | 40-80 mg PO bid initially; increase to 160-320 mg/d (<640 mg/d in some patients) |
| Pediatric Dose | 1-4 mg/kg/d PO divided q6-8h; not to exceed 16 mg/kg/d; avoid IV use; however, if necessary (eg, hypercyanotic spell in tetralogy of Fallot or severe infundibular reaction after balloon pulmonary valvuloplasty), 0.1-0.25 mg/kg/dose IV over >10 min; not to exceed 1 mg/dose (infants) or 3 mg/dose (children) |
| Contraindications | Documented hypersensitivity; uncompensated CHF; bradycardia, cardiogenic shock; A-V conduction abnormalities |
| Interactions | Coadministration with aluminum salts, barbiturates, NSAIDs, penicillins, calcium salts, cholestyramine, and rifampin may decrease effects; calcium channel blockers, cimetidine, loop diuretics, and MAOIs may increase toxicity; may increase toxicity of hydralazine, haloperidol, benzodiazepines, and phenothiazines |
| Pregnancy | C - Fetal risk revealed in studies in animals but not established or not studied in humans; may use if benefits outweigh risk to fetus
|
| Precautions | Beta-blockade may decrease signs of acute hypoglycemia and hyperthyroidism; abrupt withdrawal may exacerbate symptoms of hyperthyroidism, including thyroid storm; withdraw slowly and monitor closely |
Further Inpatient Care
- The neonate with critical pulmonary stenosis requires special consideration.
- Patients with critical pulmonary stenosis may present with near–pulmonary atresia (cyanotic lesion) with a small and often inadequate right ventricle. These patients survive because of a PDA.
- Although balloon pulmonary valvuloplasty produces good results, nearly 25% patients are rehospitalized to address related complications, restenosis, and associated defects.
- Patients with associated severe infundibular or supravalvar pulmonary stenosis require surgical intervention.
- Definitive repair may not be possible if the right ventricle is hypoplastic or if single ventricular palliation (eg, as with the Fontan procedure or a variation of this) is needed. The modified Fontan procedure currently used is staged cavopulmonary connection.
Further Outpatient Care
- Clinical, ECG, and Doppler echocardiographic evaluation are recommended at 1, 6 and 12 months after balloon pulmonary valvuloplasty and yearly thereafter.
- Patients with trivial and mild pulmonary stenosis do not need intervention to relieve the pulmonary valve obstruction. However, they should be clinically followed up at periodic intervals, eg, on a yearly basis.
- Routine well-child care, including immunizations, as per the primary physician, is suggested.
- Physical activity should be normal.
- Most patients with PS are given prophylaxis for subacute bacterial endocarditis (SBE).
- Opinions differ about the need for SBE prophylaxis in patients with valvar pulmonary stenosis because of the extremely low incidence of pulmonary valve endocarditis in this relatively large subpopulation. The first author recommends SBE prophylaxis for all patients with valvar pulmonary stenosis.
Transfer
- Transfer patients with symptomatic pulmonary stenosis to a tertiary care center offering pediatric cardiology and pediatric cardiothoracic surgery services.
Complications
- Hypercontractile residual obstructing muscular hypertrophy in the infundibulum: This complication is associated with acute palliation for severe pulmonary stenosis.
- Infundibular obstruction: This phenomenon, which occurs after valvar stenosis repair by means of surgery or valvuloplasty, has led to the designation of a suicide right ventricle. Beta-blockers and volume replacement are used to treat this condition, which occurs more frequently in relatively old patients with long-standing pulmonary stenosis.
- Late atrial arrhythmias
- Persistent repolarization abnormalities
Prognosis
- Mild valvar pulmonary stenosis usually does not progress, but moderate-to-severe disease does tend to progress.
- After the stenosis is relieved by means of balloon valvuloplasty, 8-10% patients require repeat intervention. The condition does not recur in the remaining patients, and right ventricular hypertrophy regresses.
- The outcome after balloon or surgical valvulotomy is generally excellent. The probability of survival is similar to that of the general population, and the vast majority of patients are asymptomatic.
Patient Education
- Reassure patients and parents of children with mild valvar pulmonary stenosis that this condition is not related to or associated with coronary artery disease, dysrhythmia, or sudden death.
- Insurability may become a factor in obtaining further care. Patients are no more at risk for disastrous health consequences than is the usual population.
- Provided the patient is asymptomatic and acyanotic and provided that initial Doppler echocardiograms show only mild valvar pulmonary stenosis, yearly screening examination and ECG are prudent follow-up care.
- If evaluations performed a few years after the initial evaluation reveal no clinically significant change, the patient may be followed up once every 3-5 years.
- For excellent patient education resources, visit eMedicine's Heart Center. Also, see eMedicine's patient education article Tetralogy of Fallot.
Medical/Legal Pitfalls
- Failure to exclude associated congenital anomalies and to detect the presence of cyanosis or a ductal-dependent lesion is a major error.
- Failure to diagnose a serious congenital heart defect, such as tetralogy of Fallot is problematic.
- Acyanotic patients with tetralogy of Fallot and those with a mild obstruction of the right ventricular outflow tract may have similar presentations and physical findings.
- Tetralogy of Fallot is a lesion that is surgically correctable and that can be corrected safely, even in the neonatal period.
- A tet spell, or hypercyanotic spell, is potentially lethal but frequently aborted with simple skills. Such spells and can occur in previously acyanotic patients with tetralogy of Fallot ("pink tets").
- Echocardiography can reliably confirm the precise diagnosis and help in differentiating valvar pulmonary stenosis from tetralogy of Fallot.
- Echocardiography should not be withheld if complex anatomy is suspected.
- Complications associated with balloon pulmonary valvuloplasty, such a rupture or perforation of the pulmonary artery or right ventricular outflow tract, are uncommon but can occur. This possibility should be explained to all parents.
| Media file 1:
In valvar pulmonic stenosis, the severity of obstruction may be judged by auscultatory findings. In mild stenosis, the ejection click (EC) is clearly separated from the first heart sound (S1). The murmur starts with the click, peaks in early systole, and ends way before the aortic component of the second heart sound (A2) The pulmonary component of the second heart sound (P2) is normal to increased in intensity. In moderate pulmonic stenosis, the click is closer to the first heart sound, the ejection murmur peaks later in the systole and the murmur reaches the A2, and the second heart sound is widely split with soft pulmonary component. In severe valvar obstruction, the click is either absent or occurs so close to S1 that it cannot be heard separately, and the murmur peaks late in systole and extends beyond the A2. The second heart sound is widely split with an extremely soft or inaudible P2. Reproduced from Rao PS: Evaluation of cardiac murmur in children. Indian J Pediatr 1991 Jul-Aug; 58(4): 471-91. |
 | View Full Size Image | |
Media type: Graph
|
| Media file 2:
Posteroanterior chest roentgenogram in a patient with valvar pulmonic stenosis showing normal-sized heart with normal pulmonary vascular markings. Note prominent main pulmonary artery (arrow). Reproduced with permission from Rao PS: Diagnosis and management of acyanotic heart disease: Part I – Obstructive lesions. Indian Journal of Pediatrics 2005; 72: 495-502. |
 | View Full Size Image | |
Media type: X-RAY
|
| Media file 3:
Doppler flow velocity recordings from the main pulmonary artery prior to (left) and 1 day (center) and 10 months (right) after successful balloon pulmonary valvuloplasty. Note that no significant fall in the peak flow velocity is present on the day after balloon procedure, but a characteristic triangular pattern is present, indicative of infundibular obstruction. At 10-month follow-up, the flow velocity decreased, suggesting resolution of infundibular obstruction. Reproduced with permission from Thapar MK: Significance of infundibular obstruction following balloon valvuloplasty for valvar pulmonic stenosis. Am Heart J 1989; Jul; 118(1): 99-103. |
 | View Full Size Image | |
Media type: CT
|
| Media file 4:
Right ventricular (RV) cineangiogram in lateral view in a child with valvar pulmonary stenosis demonstrating thickened and domed pulmonary valve leaflets and poststenotic dilatation of the pulmonary artery (PA). Reproduced with permission from Rao PS: Diagnosis and management of acyanotic heart disease: Part I – Obstructive lesions. Indian Journal of Pediatrics 2005; 72: 495-502. |
 | View Full Size Image | |
Media type: Image
|
| Media file 5:
Selected cineradiographic frames of a balloon dilatation catheter placed across a stenotic pulmonary valve. Note "waisting" of the balloon during the initial phases of the balloon inflation (A), which was almost completely abolished during the later phases of balloon inflation (B). Reproduced from Rao PS: Balloon pulmonary valvuloplasty for isolated pulmonic stenosis. In: Rao PS, ed: Transcatheter Therapy in Pediatric Cardiology New York, NY: Wiley-Liss; 1993: 59-104. |
 | View Full Size Image | |
Media type: Image
|
| Media file 6:
Selected frames from lateral view of the right ventricular (RV) cineangiogram showing severe infundibular stenosis (A) immediately following balloon valvuloplasty (corresponding Image 3, center). At 10 months after balloon valvuloplasty, the right ventricular outflow tract (B) is wide open and corresponds to Image 3, right. Peak-to-peak pulmonary valve gradient was 20 mm Hg and no infundibular gradient was present. PA = Pulmonary artery. Reproduced with permission from Thapar MK: Significance of infundibular obstruction following balloon valvuloplasty for valvar pulmonic stenosis. Am Heart J 1989; Jul; 118(1): 99-103. |
 | View Full Size Image | |
Media type: Image
|
- Anand R, Mehta AV. Natural history of asymptomatic valvar pulmonary stenosis diagnosed in infancy. Clin Cardiol. Apr 1997;20(4):377-80. [Medline].
- Ardura J, Gonzalez C, Andres J. Does mild pulmonary stenosis progress during childhood? A study of its natural course. Clin Cardiol. Sep 2004;27(9):519-22. [Medline].
- Bahl VK, Chandra S, Goel A, et al. Versatility of Inoue balloon catheter. Int J Cardiol. Mar 1997;59(1):75-83. [Medline].
- Balfour IC, Rao PS. Pulmonary Stenosis. Curr Treat Options Cardiovasc Med. Dec 2000;2(6):489-498. [Medline].
- Berman W, Fripp RR, Raisher BD, Yabek SM. Significant pulmonary valve incompetence following oversize balloon pulmonary valveplasty in small infants: A long-term follow-up study. Catheter Cardiovasc Interv. Sep 1999;48(1):61-5; discussion 66. [Medline].
- Castenada AR, Jonas RA, Meyer JE. Surgery for infants with congenital heart defects In: Cardiac Surgery of the Neonate and Infant. 1st ed. 1993:1013-1035.
- Fontes VF, Esteves CA, Sousa JE, et al. Regression of infundibular hypertrophy after pulmonary valvuloplasty for pulmonic stenosis. Am J Cardiol. Nov 1 1988;62(13):977-9. [Medline].
- Galindo A, Gutierrez-Larraya F, Velasco JM, de la Fuente P. Pulmonary balloon valvuloplasty in a fetus with critical pulmonary stenosis/atresia with intact ventricular septum and heart failure. Fetal Diagn Ther. 2006;21(1):100-4. [Medline].
- Gielen H, Daniels O, van Lier H. Natural history of congenital pulmonary valvar stenosis: an echo and Doppler cardiographic study. Cardiol Young. Mar 1999;9(2):129-35. [Medline].
- Gikonyo BM, Lucas RV, Edwards JE. Anatomic features of congenital pulmonary valvar stenosis. Pediatr Cardiol. 1987;8(2):109-16. [Medline].
- Hayes CJ, Gersony WM, Driscoll DJ, et al. Second natural history study of congenital heart defects. Results of treatment of patients with pulmonary valvar stenosis. Circulation. Feb 1993;87(2 Suppl):I28-37. [Medline].
- Jarrar M, Betbout F, Farhat MB, et al. Long-term invasive and noninvasive results of percutaneous balloon pulmonary valvuloplasty in children, adolescents, and adults. Am Heart J. Nov 1999;138(5 Pt 1):950-4. [Medline].
- Johnson LW, Grossman W, Dalen JE, Dexter L. Pulmonic stenosis in the adult. Long-term follow-up results. N Engl J Med. Dec 7 1972;287(23):1159-63. [Medline].
- Jureidini SB, Rao PS. Critical Pulmonary Stenosis in the Neonate: Role of Transcatheter Management. J Invasive Cardiol. Sep 1996;8(7):326-331. [Medline].
- Kan JS, White RI, Mitchell SE, Gardner TJ. Percutaneous balloon valvuloplasty: a new method for treating congenital pulmonary-valve stenosis. N Engl J Med. Aug 26 1982;307(9):540-2. [Medline].
- Koretzky ED, Moller JH, Korns ME, et al. Congenital pulmonary stenosis resulting from dysplasia of valve. Circulation. Jul 1969;40(1):43-53. [Medline].
- Lima CO, Sahn DJ, Valdes-Cruz LM, et al. Noninvasive prediction of transvalvular pressure gradient in patients with pulmonary stenosis by quantitative two-dimensional echocardiographic Doppler studies. Circulation. Apr 1983;67(4):866-71. [Medline].
- Massoud I, Imam A, Mabrouk A, et al. Palliative balloon valvoplasty of the pulmonary valve in tetralogy of Fallot. Cardiol Young. Jan 1999;9(1):24-36. [Medline].
- McCrindle BW. Independent predictors of long-term results after balloon pulmonary valvuloplasty. Valvuloplasty and Angioplasty of Congenital Anomalies (VACA) Registry Investigators. Circulation. Apr 1994;89(4):1751-9. [Medline].
- Milazzo AS, Golioto A, Camitta MG, O''Laughlin MP. An infant with subvalvar and valvar aortic stenosis, subvalvar and valvar pulmonary stenosis, severe biventricular hypertrophy and pulmonary hemorrhage: a case report. Pediatr Cardiol. Mar-Apr 2003;24(2):169-71. [Medline].
- Milo S, Bar-el Y, Adler Z, et al. Punch pulmonary valvectomy: another alternative to treat critical pulmonary stenosis. Ann Thorac Surg. Jul 1997;64(1):261-4. [Medline].
- Nadas A. Pulmonary stenosis. In: Fyler DC, ed. Nadas' Pediatric Cardiology. Hanley & Belfus;1992:459-470.