Author: Bojan Paunovic, MD, Assistant Professor, Department of Internal Medicine, Section of Critical Care, University of Manitoba; Medical Director of Critical Care, Grace Hospital, Canada Bojan Paunovic is a member of the following medical societies: Canadian Medical Association, Royal College of Physicians and Surgeons of Canada, and Society of Critical Care Medicine Coauthor(s): Sat Sharma, MD, FRCP(C), FCCP, FACP, DABSM, Program Director, Associate Professor, Department of Internal Medicine, Divisions of Pulmonary and Critical Care Medicine, University of Manitoba; Site Director of Respiratory Medicine, St Boniface General Hospital Editors: Gregory Joseph Dehmer, MD, Director, Division of Cardiology, Professor, Department of Medicine, Scott & White Clinic, Texas A&M University School of Medicine; Francisco Talavera, PharmD, PhD, Senior Pharmacy Editor, eMedicine; Ronald J Oudiz, MD, Director of Pulmonary Hypertension, Associate Professor, Department of Medicine, Division of Cardiology, Harbor-UCLA Medical Center, David Geffen School of Medicine at UCLA; Amer Suleman, MD, Consultant in Electrophysiology and Cardiovascular Medicine, Department of Internal Medicine, Division of Cardiology, Medical City Dallas Hospital; Karlheinz Peter, MD, PhD, Professor of Medicine, Monash University; Head of Centre of Thrombosis and Myocardial Infarction, Head of Division of Atherothrombosis and Vascular Biology, Associate Director, Baker Heart Research Institute; Consulting Staff, Department of Cardiology, Alfred Hospital Author and Editor Disclosure Synonyms and related keywords: swan-ganz catheterization, pulmonary artery catheterization, PAC, SGC, right heart catheterization, hemodynamic monitoring, pulmonary artery catheter, PA catheter, acute myocardial infarction, AMI, flow-directed balloon-tipped pulmonary artery catheter, primary pulmonary hypertension, PPH, valvular disease, intracardiac shunts, cardiac tamponade, pulmonary embolus, PE, multiorgan system failure, severe burns, aspiration of air emboli, heart pressure, pulmonary artery pressure, pulmonary artery occlusion pressure, wedge pressure, left ventricular end-diastolic pressure, positive end-expiratory pressure hypovolemic shock, cardiogenic shock, cardiomyopathy, acute mitral regurgitation, acute ventricular septal defect, septic shock, pericardial tamponade, cardiac tamponade, constrictive pericarditis, mitral stenosis, aortic stenosis, aortic regurgitation, cardiac output INTRODUCTION
The flow-directed balloon-tipped pulmonary artery catheter (PAC) (also known as the Swan-Ganz or right heart catheter) has been in clinical use for more than 30 years. Initially developed for the management of acute myocardial infarction (AMI), it gained widespread use in the management of a variety of critical illnesses and surgical procedures. This article covers both the clinical and technical aspects of its use. History In 1929, Werner Forssmann was the first to demonstrate that a catheter could be advanced safely into the human heart (in this case, his own). His primary purpose was to develop a technique for direct delivery of drugs to the heart. With the advent of the PAC, pressure measurements and blood sampling from the cardiac chambers and PA became possible outside of the cardiac catheterization laboratory. This transformed the procedure from one that was labor and resource intensive to one that could be performed rapidly at the bedside of a critically ill patient. Although the concept of using a balloon-assisted catheter had been published 15 years earlier, a serendipitous observation by a noted cardiologist led to its further development. During a day at the beach, Dr H. J. Swan noticed a sailboat moving quickly despite the calm weather. This led to the initial idea of devising a catheter with a parachutelike or sail-like device attached. Initial testing was conducted with a balloon-tipped catheter because it was easier to fabricate. It proved so successful that the original parachute idea was abandoned. At the same time, the work of William Ganz on the thermodilution method of measuring cardiac output (CO) was incorporated into the catheter's use. This basic design remains in use today. Interestingly, despite the widespread use of their names for the flow-directed balloon-tipped PA catheter, neither the physicians nor the original manufacturer could obtain a patent. INDICATIONS AND CONTRAINDICATIONS
Despite widespread use of the PAC for more than 3 decades, no validated indications exist for its general use. Some publications present only authors' suggestions for indications and contraindications for the use of the PAC. Various association and subspecialty guidelines exist as consensus statements. For example, in 1998 the American College of Cardiology published a consensus statement regarding the use of the PAC in patients with cardiac disease. The following are general guidelines for clinical use of PAC:
TECHNICAL CONSIDERATIONS
Pulmonary artery catheter Although variations of the PAC exist, the typical version is a multilumen catheter, 110 cm long, with extra connecting tubes for attachment to the pressure transducer (see Media file 1). At the tip is the PA lumen, or distal lumen. A 1.5-cc balloon is located just proximal to the tip. Approximately 4 cm proximal to the balloon is the thermistor used to measure temperature changes for calculation of CO. Two additional lumens usually are present at 19 cm and 30 cm from the tip. Depending on the degree of right heart enlargement and the position of the catheter (ie, distance advanced into the patient), these lumina reside within the right ventricle (RV), right atrium (RA), or the superior vena cava (SVC). Some catheters are coated with heparin to reduce thrombogenicity and have connections for temporary ventricular pacing. The former is important to remember in case the patient develops heparin-induced thrombocytopenia, because only a small amount of heparin is necessary to sustain this process. Necessary equipment Proper attachment of the PAC to the monitoring equipment is essential for accurate measurements. Transmission of pressures from the body to the display system is accomplished via semirigid, noncompliant tubing filled with fluid, usually isotonic saline with a small amount of heparin. This, in turn, is connected to a fluid-filled pressure transducer. Often, a constant infusion or "interflow" device is placed into the connecting pressure line. This device does not alter the pressure and provides a small constant infusion of fluid through the catheter to prevent backup of blood. Because the fluid is incompressible and the tubing noncompliant, this system fairly accurately transmits intracardiac pressures to the transducer, causing small amounts of movement in the transducer membrane. Deformation of this membrane generates a proportional electric current that is amplified and transmitted to the monitor. Zero reference Any independent vertical movement of the transducer or the patient will affect the hydrostatic column of this fluid-filled system and thus alter the pressure measurements. At some time before or after PAC insertion, the system must therefore be zeroed to ambient air pressure. The reference point for this is the midpoint of the left atrium (LA), estimated as the fourth intercostal space in the midaxillary line with the patient in the supine position. With the transducer at this height, the membrane is exposed to atmospheric pressure, and the monitor is then adjusted to zero. Calibration Once zeroed, the monitoring system must be calibrated for accuracy. Currently, most monitors perform an automated electronic calibration. Two methods are used to manually calibrate and check the system. If the catheter has not been inserted, the distal tip of the PAC is raised to a specified height above the LA. For example, raising the tip 20 cm above the LA should produce a reading of approximately 15 mm Hg if the system is working properly (1 mm Hg equals 1.36 cm H20). Alternatively, pressure can be applied externally to the transducer and adjusted to a known level using a mercury or aneroid manometer. The monitor then is adjusted to read this pressure, and the system is calibrated. Dynamic tuning Central pressures are dynamic waveforms (ie, they vary from systole to diastole) and thus have a periodic frequency. To monitor these pressures accurately, the system requires an appropriate frequency response. A poorly responsive system produces inaccurate pressure readings, and differentiating waveforms (eg, PA from pulmonary capillary wedge pressure [PCWP]) can become difficult. When signal energy is lost, the pressure waveform is dampened. Common causes of this are air bubbles (which are compressible), long or compliant tubing, vessel wall impingement, intracatheter debris, transducer malfunction, and loose connections in the tubing. A qualitative test of the frequency response is performed by flicking the catheter and observing a brisk high-frequency response in the waveform. After insertion, the system can be checked by using the rapid flush test. When flushed, an appropriately responsive system shows an initial horizontal straight line with a high-pressure reading. Once the flushing is terminated, the pressure drops immediately, which is represented by a vertical line that plunges below the baseline. A brief and well-defined oscillation occurs, followed by return of the PA waveform. A dampened system will not overshoot or oscillate, and causes a delay in returning to the PA waveform. Insertion The PAC is inserted percutaneously into a major vein (jugular, subclavian, femoral) via an introducer sheath. The actual venous access techniques are not described here, but the following points are important. Preference considerations for cannulation of the great veins are as follows:
As with any catheterization procedure, sterile technique is essential. The total length of a PAC is approximately 150 cm; extra sterile towels around the head, shoulders, and chest ensure that aseptic technique is not compromised. While the Trendelenburg position is used for venous access (internal jugular [IJ] and subclavian routes), passage of the PAC is easier when the patient subsequently is placed flat or slightly upright. Before insertion, check the PAC for cracks and kinks. Then, check balloon function (see Media file 2), connect all lumens to stopcocks, and flush them to eliminate air bubbles. Flick the PAC tip to check frequency response. Finally, the PAC is threaded through a sterile sleeve (be sure to check orientation) to ensure sterility of the PAC after insertion and allow some adjustment of position. The packaging of the PAC causes it to have a preformed curve. This can be used to facilitate passage into the PA. The direction in which the curl is inserted into the introducer depends on which vein is cannulated. For instance, from the head of the bed using the RIJ approach, the curl should be in the direction of the patient's left shoulder (concave-cephalad). Once the PAC is in the RV, a clockwise quarter turn moves the tip anteriorly to allow easier passage into the PA. After inserting the PAC as far as the 20-cm mark (30-cm mark if the femoral route used), the balloon is inflated with air. Inflation should be slow and controlled (1 mL/s) and should not surpass the recommended volume (usually 1.5 mL). Always inflate the balloon before advancing the PAC, and always deflate the balloon before withdrawing the PAC. Always use continuous pressure monitoring from the distal lumen. Watch the monitor for changes in the waveform and abnormal cardiac rhythms. From the RIJ approach, the RA is entered at approximately 25 cm, the RV at approximately 30 cm, and the PA at approximately 40 cm; the PCWP can be identified at approximately 45 cm. If an RV waveform still present approximately 20 cm after the initial RV pattern appears, the catheter may be coiling in the RV. If withdrawal is necessary, always proceed slowly to decrease the risk of knotting the catheter upon itself. If the catheter is knotted, fluoroscopy may be necessary to visualize the catheter and remove the knot. As a last resort, slowly withdraw the PAC to the point where it catches on the introducer tip. From this point, the PAC and introducer can be removed as one unit. Apply prompt pressure for a minimum of 5 minutes. If bleeding persists, suturing the site may be necessary. Once the PCWP is obtained and the catheter sleeve secured, make sure the PCWP pattern is reproducible before removing the sterile field. Also, determine the volume of air in the balloon required to obtain a PCWP waveform. Volumes less than half the balloon maximum may indicate that the tip is too far distal. Some clinicians advocate that, after establishing that the PA diastolic pressure is equal to the PCWP pressure, further balloon inflations are unnecessary and the PA diastolic pressure should be used as the parameter to assess left ventricular (LV) filling; this relationship may not hold if the clinical situation changes. Once the procedure is complete, obtain a chest radiograph to check the position of the PAC and to assess for central venous access complications (eg, pneumothorax). Troubleshooting Passage of the PAC is difficult in certain disease states. When right-sided pressures are elevated, the air-filled balloon actually may hinder proper positioning. In such cases, filling the balloon with 1 mL of sterile saline and placing the patient in a more upright position allows gravity to cause the PAC to fall into position. Once the catheter is in position, aspirate the saline and replace it with air to ensure reproducible PCWP tracings. Insertion with a noninflated balloon may also allow passage into the PA. Neither of these techniques are advocated by PAC manufacturers and the techniques have the potential for adverse events. These techniques should only be used in extenuating circumstances and should only be attempted by experienced practitioners under the guidance of fluoroscopy. The presence of large V waves can make discriminating a PA tracing from a PCWP tracing difficult. Look for subtle signs of waveform differences, such as loss of the dicrotic notch in the PA tracing (see Media file 3). Determining the oxygen saturation of a blood sample obtained from the distal lumen while the balloon is inflated also can confirm that the waveform is a true PCWP. After aspirating enough volume (5-7 mL) to clear the blood from the PA distal to the inflated balloon, the oxygen saturation should be similar to that measured by arterial blood gas or pulse oximetry, thus confirming that the catheter is in the correct position to measure PWCP. Fluoroscopy may be required for proper placement of the PAC in difficult situations. Measurements
Other important information provided by a PAC catheter includes the CO, mixed venous oxygen saturation (SaO2), and oxygen saturations in the right heart chambers to assess for the presence of an intracardiac shunt. Using these measurements, other variables can be derived, including pulmonary or systemic vascular resistance and the difference between arterial and venous oxygen content (see Media files 4-5). Obtaining CO and PCWP measurements is the primary reason for inserting most PACs; therefore, understanding how they are obtained and what factors alter their values is of prime importance. WAVE FORM ANALYSIS IN HEALTHY STATES
Cardiac pressuresRight- and left-sided heart pressure waveforms share many physiologic similarities, but, in the healthy individual, the waves are of different magnitudes. Right-sided pressures The central venous pressure (CVP) and right atrial pressure (RAP) are nearly equal to the diastolic RV pressure in the absence of heart or lung disease (see Media file 6). The mean CVP and RAP normally range from 0-5 mm Hg, and vary as intrathoracic pressure changes with respiration (see Media file 7). RA contraction creates pressure changes, which are influenced strongly by the patient's volume status. Atrial contraction produces an increase in pressure called the A wave. The C wave is a small convexity noted on the initial descent of the A wave and is thought to be secondary to closure of the tricuspid valve. The initial descent after the A wave is called the X descent. This decline in RAP is secondary to RA relaxation and downward movement of the tricuspid valve. Following this is the V wave, which is somewhat smaller than the A wave, and reflects RA filling during ventricular systole. The Y descent occurs after the V wave and represents rapid filling of the RV after opening of the tricuspid valve (see Media file 8). CVP is most commonly elevated in the setting of biventricular heart failure. Other causes of RAP elevation are tricuspid regurgitation or stenosis, pulmonary hypertension, volume overload, constrictive pericarditis, and cardiac tamponade. Large, so-called cannon A waves occur when the RA contracts against a closed tricuspid valve. Cannon A waves are detected in certain cardiac rhythm disturbances, including junctional rhythms and ventricular tachycardia, and in some patients with ventricular pacemakers. Large V waves may occur in the presence of tricuspid regurgitation, with their magnitude affected by the size and compliance of the RA. Pulmonary arterial pressure In the pulmonary artery pressure (Ppa) tracing, an initial positive upstroke secondary to RV systole occurs, and a dicrotic notch is formed on the downstroke when the pulmonary valve closes. A normal PA systolic pressure ranges from 20 to 30 mm Hg and is equal to the RV systolic pressure. Ppa is elevated in some high-flow states (eg, hypervolemia), left ventricular failure, and high-resistance states (eg, pulmonary hypertension, mitral valve disease) (see Media file 9). Pulmonary artery occlusion pressure (wedge pressure) Understanding the theory and required assumptions behind PCWP measurement and conditions that alter it are essential for proper use of this often misunderstood measurement. When the PAC tip is positioned properly and the balloon is inflated, the PAP tracing disappears. This occurs because inflation of the balloon causes distal migration (approximately 2 cm) of the tip into a smaller branch of the PA, where it occludes blood flow. The resulting nonpulsatile pressure tracing is called the PCWP (or pulmonary artery occlusion pressure [Ppao]) (see Media file 10). Under the proper circumstances, this pressure reflects the mean left atrial pressure (LAP) (see Media file 11). The assumption is that a static column is created between the PAC tip and the LA. This assumption is correct only if the tip is in the proper lung zone and no vascular obstruction, such as pulmonary vein stenosis, occurs downstream. When the PAC catheter balloon is inflated, the balloon stops antegrade blood flow and allows an uninterrupted column of blood to exist between the catheter tip and the LA (see Media files 12-13). The PCWP waveform reflects events in the LA. The A, C, and V waves have origins similar to those that appear in the RAP waveform (see Media file 14). The waveforms can be discerned by using simultaneous ECG monitoring (see Media file 15). The 3 lung zones of West The lung can be divided into 3 vertical zones with varying pressure changes (see Media file 16). To assess proper location, a supine chest radiograph showing the tip below the level of the LA is sufficient, although occasionally a lateral chest radiograph is required. If the tip position remains questionable, blood can be aspirated from the distal port during balloon inflation. Preload (left ventricular end-diastolic pressure)PCWP is a reflection of LAP, which, in the absence of mitral valve disease, is an indication of LV diastolic pressure. Often, the inference is made that PCWP reflects left ventricular end-diastolic volume (LVEDV) or end-diastolic pressure (LVEDP). Numerous conditions in critically ill patients preclude this assumption. PCWP is the measurement by which changes in lung water (pulmonary capillary hydrostatic pressure [PCHP]) can be assessed. This concept holds true only if the resistance of the pulmonary venous system is assumed to be zero. In fact, the small pulmonary veins and capillaries account for approximately 40% of the total pulmonary vascular resistance. This value may be even higher in critically ill patients in whom pulmonary venoconstriction is common secondary to conditions such as hypoxemia and acute respiratory distress syndrome (ARDS). PCHP is always greater than PCWP. PCWP can be used to estimate the contribution of PCHP to lung edema if evidence of chronically elevated Ppv, permeability, pleural pressure, and osmotic pressure are considered. Effect of respirationThe final critical concept in PCWP interpretation is the effect of the respiratory cycle on PCWP measurements. The timing of PCWP measurement is critical because intrathoracic pressures can vary widely with inspiration and expiration and are transmitted to the pulmonary vasculature. During spontaneous inspiration, the intrathoracic pressures decrease (more negative); during expiration, intrathoracic pressures increase (more positive). Positive pressure ventilation (eg, in an intubated patient) reverses this situation. To minimize the effect of the respiratory cycle on intrathoracic pressures, measurements are obtained at end-expiration, when intrathoracic pressure is closest to zero. In patients with severe respiratory distress, end-expiration can be difficult to determine. In these situations, sedation, or even paralysis, may be necessary to remove the transmission of respiratory efforts to the pressure tracings. Positive end-expiratory pressurePEEP (intrinsic or extrinsic) also transmits pressure to the vascular space. Lung compliance is the main determinant of the amount of pressure transmission. For example, in disease states (eg, ARDS) associated with low compliance (ie, stiff lungs), pressure transmission is minimal. Debate exists over how to correct PCWP in the presence of PEEP. Although previously advocated, temporary discontinuation of PEEP may have adverse effects, such as cardiovascular collapse or hypoxemia, that are difficult to reverse. For PEEP greater than 10 cm H2O, the following general rule can be applied: Corrected PCWP equals measured PCWP minus one half the quotient of PEEP divided by 1.36. If available, an intraesophageal balloon can be used. Esophageal pressure equals pleural pressure, so corrected PCWP equals measured PCWP minus esophageal pressure. WAVE FORM ANALYSIS IN PATHOLOGIC STATES
ShockShock has been defined as inadequate perfusion to meet the metabolic demands of body tissues. The most common forms of shock are hypovolemic, cardiogenic, septic, and obstructive. PACs are used frequently in the management of various forms of shock, as described in this section (see Media file 17). Hypovolemic shock Hypovolemic shock is due to a reduction in circulating blood volume resulting from either hemorrhage or fluid depletion. Preload is markedly decreased, leading to inadequate ventricular filling. The patient with hypovolemic shock manifests hypotension and tachycardia. Systemic, venous, and intracardiac pressures are abnormally low. The overall PAC pressure tracing has a damped appearance. Cardiogenic shock Cardiogenic shock is the result of severe depression in cardiac performance. Cardiogenic shock is characterized by systolic blood pressure less than 80 mm Hg, cardiac index less than 1.8 L/min/m2, and PCWP greater than 18 mm Hg. This form of shock can occur from a direct insult to the myocardium (eg, large AMI, severe cardiomyopathy) or from a mechanical problem that overwhelms the functional capacity of the myocardium (eg, acute severe mitral regurgitation, acute ventricular septal defect). Common causes of acute mitral regurgitation in critical care units are ruptured papillary muscles secondary to AMI, myocardial ischemia leading to papillary muscle dysfunction, bacterial endocarditis, ruptured chordae, and trauma. Other causes are rheumatic fever and myxomatous degeneration of the mitral valve. With acute mitral regurgitation, large volumes of blood regurgitate into a poorly compliant LA, raising Ppv and causing pulmonary edema. Large V waves usually are observed in the PCWP pressure tracing (see Media files 18-20). The PA waveform appears falsely elevated because of the large V wave reflected back from the LA through the compliant pulmonary vasculature. The Y descent is quite rapid as the overdistended LA quickly empties. Care must be exercised to distinguish a large V wave from a systolic PA waveform. Failure to recognize a large V wave may cause the PAC to be advanced further in an attempt to record a PCWP pressure, increasing the risk of perforation. In chronic mitral regurgitation, an equivalent volume of blood may regurgitate, but this volume is better tolerated by a markedly dilated LA. Compared with acute mitral regurgitation, LA pressure may be less and large V waves may be absent. Septic shock Septic shock is the most common cause of death in intensive care units in the United States. Septic shock is an example of distributive shock, a form of shock characterized by profound peripheral vasodilation. Although the CO may be normal or even elevated in this type of shock, organ and tissue perfusion are inadequate. Other types of distributive shock include anaphylaxis, neurogenic shock, and adrenal insufficiency. Swan-Ganz catheter measurements frequently demonstrate low filling pressures. Extracardiac obstructive shock Pericardial tamponade is an example of this form of shock. Cardiac tamponade results from abnormal rapid fluid accumulation in the pericardial sac. The increased pericardial pressure impairs ventricular diastolic filling, decreasing preload, stroke volume, and CO. This may occur secondary to viral infections, malignancy, trauma, or myocardial rupture. As little as 50 mL of fluid accumulation can begin to impair cardiac filling during systole, leading to a severe reduction in CO. Ventricular filling is impaired throughout all of diastole, thereby causing equalization of all diastolic pressures. The RAP approximates the RV diastolic pressure, which approximates the PA diastolic pressure, and also approximates PCWP (see Media file 21). The RA waveform shows a minimal X and small and/or absent Y descent, and the mean RAP is elevated. Ppa loses its usual respiratory variation. In pericardial tamponade, the systemic arterial pressure shows evidence of pulsus paradoxus (see Media file 22). Other causes of extracardiac shock include massive PE and tension pneumothorax. Hemodynamics of other cardiac abnormalitiesConstrictive pericarditis Thickening of the pericardial sac creates an indolent process that may lead to constrictive pericarditis. This can occur in patients with rheumatic diseases, tuberculosis, metastatic cancer, or prior chest radiation or open-heart surgery. Idiopathic cases also occur. Early diastolic filling is normal until limited by the rigid pericardial shell. Once this occurs, ventricular filling is stopped abruptly, creating a plateau in the RV pressure, which is typical of constrictive pericarditis. This is called the "dip and plateau" pattern or square root sign; the RAP waveform has a characteristic configuration suggestive of an M or W. A and V waves are accentuated with rapid X and Y descents, in contrast to pericardial tamponade, above. PCWP may be as high as 20-25 mm Hg, and usually appears similar to the RA waveform. Pulsus paradoxus is present much less commonly with constrictive pericarditis than with pericardial tamponade (see Media files 23-24). Mitral stenosis In severe mitral stenosis, LAP, and thus PAWP, is elevated. Pulmonary hypertension also develops as the severity of the valve lesion progresses. This leads to increase in RV systolic pressure and in the RA A wave. RV diastolic pressure may increase if RV failure or important tricuspid regurgitation develops. Atrial fibrillation is a common complication in mitral stenosis and results in loss of A waves in both the RA and PCWP pressure tracings. Aortic stenosis Aortic stenosis can be supravalvular, valvular, or subvalvular in origin. The RA, RV, and PA waveforms usually are normal unless congestive heart failure is present. PCWP may show large A waves in severe cases because of poor LV compliance. Aortic regurgitation The hemodynamic abnormalities are different in acute and chronic aortic regurgitation. Acute regurgitation is observed most often in bacterial endocarditis, chest trauma, ascending aortic dissection, and degeneration of valve leaflets. The hemodynamics in acute aortic regurgitation include modestly elevated RAP and elevated RV systolic and diastolic pressures. PA systolic and diastolic pressures also are elevated, as is PCWP. A widened and elevated systemic arterial pressure without a dicrotic notch is sometimes observed. Acute and chronic aortic regurgitation often present with contrasting manifestations; a wide pulse pressure usually is not observed in acute regurgitation. OTHER HEMODYNAMIC MEASUREMENTS MADE BY THE PULMONARY ARTERY CATHETER
Cardiac output CO can be determined via the PAC by several methods. It can be determined by using the Fick principle, which is a variation of the law of conservation and states that consumption of a substance must equal the product of blood flow to the organ and the difference between the arterial and venous concentrations of the substance. In this circumstance, the substance is oxygen, and CO is determined by the following formula: CO equals oxygen consumption per minute (VO2) divided by arterial oxygen content (CaO2) minus mixed venous oxygen content (CvO2) CO is determined by using systemic arterial and PA blood samples, and by measuring or estimating VO2. The Fick method is most accurate when the CO is low and the arterial-venous oxygen difference is high. Unfortunately, in critically ill patients, establishing a steady-state and estimating or measuring VO2 is difficult; thus, the reliability of this technique is poor. The indicator-dilution technique is more accurate and reproducible. A known amount of dye (indocyanine green) is injected into the PA. Arterial blood is withdrawn from the aorta as the dye circulates, and a concentration-versus-time curve is derived. The first-pass curve is used to determine CO, which is calculated by dividing the initial mass of the injectate by the average concentration. This value then is corrected (60 s/time of the curve) to obtain CO. This procedure requires considerable blood sampling and is time consuming, because recirculation of the dye complicates the calculation. Many PACs also allow CO to be measured by using a variation of the indicator-dilution method known as the thermodilution method. This method is more efficient because the injectate does not recirculate to a significant degree and no blood sampling is necessary. A saline bolus of known volume (5-10 mL) and temperature (usually £25ºC) is injected through the proximal (RA) lumen. The thermistor at the end of the PAC monitors the change in blood temperature, and a temperature-versus-time curve is generated (see Media file 25). The change in temperature as warm venous blood dilutes the injectate is inversely proportional to the derived CO. The Stewart-Hamilton formula shows this relationship:
Understanding this formula allows discernment of artifact errors that can lead to underestimation or overestimation. Loss of injectate or inadvertent administration of a volume lower than required results in a low-amplitude temperature-versus-time curve that produces a falsely elevated CO value. Causes of this are system leak, right-to-left intracardiac shunts, inappropriately rapid injection, and a poorly positioned PAC. Conversely, too much injectate or too slow an injection leads to a falsely low CO reading. Temperature errors can occur when continuous infusions are used. Thrombus or vessel wall impingement can alter the thermistor function. Physiologic causes for CO measurement discrepancies include tricuspid and pulmonary regurgitation, which may produce recirculation peaks and thus increase the area under the curve, resulting in a falsely low CO estimate. Arrhythmias alter steadiness of PA flow and may cause difficulty in obtaining a consistent CO. CO alterations occur during the respiratory cycle and are accentuated by respiratory distress and positive pressure ventilation. Proper timing of the injection to the same phase of respiration (preferably end-expiration) provides more consistent measurements. Averaging the values of 3 injections is recommended to minimize sampling errors. While CO is one of the most important measurements that the PAC provides, the absolute value should be normalized for the size of the patient. To account for this, the cardiac index (CI), which equals CO divided by body surface area [BSA], is calculated. The physician should keep in mind that, as independent variables, CO and CI are of limited use for assessing tissue perfusion because these must be interpreted along with other clinical data. COMPLICATIONS OF PULMONARY ARTERY CATHETERIZATION
Complications associated with PAC use relate to the initial venous access, insertion of the PAC, and maintenance of the catheter in the PA. The reported incidence of complications varies on the basis of operator skill and patient status. Significant venous access complications include arterial puncture (2-16%), which may manifest immediately (eg, carotid artery hematoma if inserted via the IJ route) or insidiously (eg, hemothorax via subclavian route). Pneumothorax (incidence 2-4%) also relates to choice of access route and occurs more often in the subclavian than in the IJ. In ventilated patients, tension pneumothorax can develop rapidly from a punctured lung. Arrhythmias constitute the most common complication associated with PAC insertion. More than 80% of these are premature ventricular contractions (PVCs) or nonsustained ventricular tachycardia (VT). These resolve either with advancement of the catheter from the RV into the PA, or with prompt withdrawal of the catheter into the RA. Significant VT or ventricular fibrillation requiring treatment occurs in fewer than 1% of patients, usually those with concurrent cardiac ischemia. Right bundle-branch block (RBBB) occurs in 5% of PAC insertions and usually is transient after positioning the catheter into the PA. The presence of a preexisting left bundle-branch block (LBBB) puts the patient at risk for complete heart block should RBBB occur. In these patients, temporary pacing equipment should be kept nearby on standby. The incidence of knotting of the PAC on itself or on intracardiac structures is less than 1%. This risk is increased in patients with dilated cardiac chambers. A persistent RV tracing despite advancement of the PAC further than 20 cm into the patient should alert the physician to this possibility. Of the complications associated with maintenance of the PAC, PA rupture is most catastrophic, with a mortality rate of 50%. Fortunately, it is a rare occurrence (<1%). Patients at risk are those who have pulmonary hypertension, are older than 60 years, or are receiving anticoagulation therapy. The sudden onset of hemoptysis (especially after inflation of the PAC balloon) indicates this possibility. Immediate management includes lateral decubitus positioning (bleeding side down), intubation with a double-lumen endotracheal tube (ETT), and increasing PEEP. Embolization via bronchoscopy or angiography or lobectomy may be necessary if bleeding continues or is massive. PAC related infection is a fairly common complication. The incidence of positive catheter tip culture result is 45% in some series. Fortunately, the risk for clinical sepsis is less than 0.5% per day of catheter use. The incidence of pulmonary infarction is less than 7%. Unintentional distal migration of the PAC tip is the usual cause. Some evidence indicates that catheter-related thrombi also may be a significant cause. While postmortem studies have shown that rate of endocardial lesions (eg, thrombi, hemorrhage, vegetations) related to PAC use is significant, correlation with clinical events has not been established. RECOMMENDATIONS FOR CLINICAL PRACTICE AND FUTURE RESEARCH
Controversies Over 1 million PACs are used annually in North America. Given the frequency and duration of their use, it is surprising that only recently were quality randomized clinical trials published. Initially, observational studies from the 1980s and 1990s indicated a greater mortality rate in patients who underwent placement of a PAC than in those who did not. The major criticism of these studies is that the more acutely ill (and therefore at greatest risk of death initially) are more likely to receive a PAC. However, since then a number of randomized clinical trials were published:
Overall, the literature does not show a positive effect on patient outcome with PAC use. However, a criticism of the current available research is that patient groups potentially benefit from the use of a PAC but this effect is lost in studies that also include patient groups that gain little or no benefit. Chittock et al published an observational cohort study showing that PAC use was associated with increased mortality in less acutely ill patients but associated with decreased mortality in more acutely ill patients.10 As well, the use of a monitoring tool such as the PAC itself is unlikely to show a significant treatment effect. Some criticize the current literature because a number of studies did not use a predefined treatment protocol. The lack of defined specific treatment based on measured PAC-derived variables could contribute more to patient outcome than merely the presence or absence of a PAC. Contention also exists that PACs do not harm people; people harm people. In other words, operator competence may be the root cause of the mortality difference. Reviews have shown deficiencies in both nursing-dependent information derived from the PAC as well as physician-dependent interpretation and subsequent management. Conclusion The PAC should not be thought of as a treatment but as a tool for aiding in diagnosis and evaluation of response to treatment. Knowledge of its capabilities and limitations is essential to minimize potential deleterious effects and maximize potential benefits. However, a rigorous randomized controlled trial has yet to show a definitive indication for its widespread application. ACKNOWLEDGMENTS
The authors and editors of eMedicine gratefully acknowledge the contributions of previous author Anthony Miller to the development and writing of this article. MULTIMEDIA