You are in: eMedicine Specialties > Cardiology > Electrophysiology Procedures Pacemaker CrosstalkArticle Last Updated: Nov 1, 2005AUTHOR AND EDITOR INFORMATIONAuthor: Vijai V Chauhan, MD, Assistant Professor, Department of Surgery, Division of Emergency Medicine, St Louis University School of Medicine Vijai V Chauhan is a member of the following medical societies: Society for Academic Emergency Medicine Coauthor(s): Antonella Quattromani, MD, Medical Director of Electrophysiology, Associate Professor of Medicine, Department of Medicine, Division of Cardiology, St Louis University Hospital Editors: Justin D Pearlman, MD, PhD, ME, MA, Director of Dartmouth Advanced Imaging Center, Professor of Medicine, Professor of Radiology, Adjunct Professor, Thayer Bioengineering and Computer Science, Dartmouth-Hitchcock Medical Center; Francisco Talavera, PharmD, PhD, Senior Pharmacy Editor, eMedicine; Brian Olshansky, MD, Professor of Medicine, Department of Internal Medicine, University of Iowa College of Medicine; Amer Suleman, MD, Consultant in Electrophysiology and Cardiovascular Medicine, Department of Internal Medicine, Division of Cardiology, Medical City Dallas Hospital; Leonard Ganz, MD, Associate Professor of Medicine, Temple University School of Medicine; Cardiac Electrophysiologist, Children's Hospital of Pittsburgh, University of Pittsburgh Medical Cent, West Penn Hospital Author and Editor Disclosure Synonyms and related keywords: pacemaker crosstalk, self-inhibition, far-field sensing, myopotential sensing, pacemaker lead fractures, electromagnetic interference, dual-mode, dual-pacing, dual-sensing pacemakers, DDD pacemakers, dual-chamber pacemakers, ventricular safety pacing, VSP INTRODUCTIONBackgroundCrosstalk is defined as the inappropriate detection of a pacemaker-generated event in one channel by the sensing amplifier of the other channel that causes inhibition of the second channel's output. This is a complication of dual-chamber pacemakers. Far-field sensing is a generic term used to refer to the oversensing of any electrical event that is not meant to be sensed. Oversensing in this case refers to the unintended detection by the pacemaker of any electrical complex including myopotential sensing, pacemaker lead fractures, and electromagnetic interference from appliances such as MRIs or microwave ovens. Pacemakers are designed to recognize electrical activity at a preprogrammed amplitude, measured in millivolts, at which any electrical event will be sensed. The dual-chamber pacemaker equipped with a unipolar atrial lead will have a larger pacing spike than those with a bipolar lead; therefore, they will have more crosstalk events. At present, most dual-mode, dual-pacing, dual-sensing (DDD) pacemakers have 2 functions designed to prevent measures against crosstalk.
PathophysiologyCrosstalk depends on the amplitude of the atrial electrical stimulus and the sensitivity of the ventricular channel. This occurs only in dual pacemakers that are programmed for synchronous AV pacing. The atrial pacing spike is sensed by the ventricular channel as a ventricular event and causes inhibition of the ventricular pacing. This results in loss of left ventricular output, and in the case of a pacer-dependent patient, in syncope or asystole. In patients with intrinsic AV conduction, crosstalk inhibition of ventricular pacing output rarely causes severe symptoms. FrequencyUnited StatesIn 166 clinical trials, only 106 patients with the same model pacemaker (Medtronic model 7006 DDD), were found to have a crosstalk rate of 1 per 59 (0.8%) at 110 beats per minute (bpm). At 130 bpm, the rate was 1 per 33 (3%), and no crosstalk was found at higher rates. Crosstalk inhibition is less common with bipolar atrial pacing than with unipolar atrial pacing. Crosstalk is a rare complication with the current DDD pacemakers. CLINICALHistoryPatients can present with dizziness, lightheadedness, palpitations, fatigue, shortness of breath, and near-syncope or syncope due to crosstalk. The patient pacemaker card should be reviewed in order to identify the manufacturer and to proceed with interrogation. Relevant pacemaker information includes the indication for the placement, date of placement, chamber type (single- or dual-chamber pacemaker), lead type (unipolar or bipolar), and current pacemaker programming. PhysicalThe vital signs should be reviewed for signs of bradycardia, hypotension, and tachypnea. The lower heart rates for pacing are at 50-60 bpm and the high are generally at 100-140 bpm. In crosstalk, where there is inappropriate pacemaker inhibition, the rate will be below 50-60 bpm. The patient may have signs of congestive heart failure with elevated jugular venous pressure, crackles in the lung fields, a third heart sound, and peripheral edema. DIFFERENTIALSSecond-Degree Atrioventricular Block Sinus Node Dysfunction Sudden Cardiac Death Syncope Third-Degree Atrioventricular Block WORKUPImaging Studies
Other Tests
TREATMENTMedical CareChanging the refractory period and/or ventricular sensitivity is one approach to restoring proper function to a pacemaker that has developed crosstalk. A change in atrial output or pulse width may alleviate the malfunction, as well. Reprogramming atrial pacing to the bipolar mode, if possible, may alleviate the problem. Any changes in pacemaker settings should be made by a qualified individual.
ConsultationsA cardiologist consultant is necessary in any suspected pacemaker malfunction. FOLLOW-UPFurther Inpatient Care
MISCELLANEOUSMedical/Legal Pitfalls
REFERENCES
Article Last Updated: Nov 1, 2005 |