Ventilator Graphics
Article Last Updated: Jun 11, 2008
AUTHOR AND EDITOR INFORMATION
Section 1 of 9
Author: Shakeel Amanullah, MD, Consulting Staff, Pulmonary, Critical Care, and Sleep Medicine, Clarian Arnett Health
Shakeel Amanullah is a member of the following medical societies: American College of Chest Physicians, American Thoracic Society, and Society of Critical Care Medicine
Coauthor(s):
Keith A Beaulieu, MD, Fellow in Pulmonary/Critical Care, Division of Pulmonary, Allergy and Critical Care Medicine, University of Pittsburgh Medical College
Editors: Francisco Talavera, PharmD, PhD, Senior Pharmacy Editor, eMedicine; Om Prakash Sharma, MD, FRCP, FCCP, DTM&H, Professor, Department of Medicine, Division of Pulmonary and Critical Care Medicine, University of Southern California Keck School of Medicine; Timothy D Rice, MD, Associate Professor, Departments of Internal Medicine and Pediatrics and Adolescent Medicine, Saint Louis University School of Medicine; Zab Mosenifar, MD, Director, Division of Pulmonary and Critical Care Medicine, Director, Women's Guild Pulmonary Disease Institute, Executive Vice Chair, Department of Medicine, Cedars Sinai Medical Center; Professor of Medicine, David Geffen School of Medicine at UCLA
Author and Editor Disclosure
Synonyms and related keywords:
ventilator waveforms, mechanical ventilation, pressure-control ventilation, volume-control ventilation, mechanical ventilation, ventilator asynchrony, triggering, ventilator trigger, ventilator flow patterns
Understanding ventilator graphics is an integral part of adequately treating patients on mechanical ventilators.
Just as pulmonary functions tests are used to better understand the lung pathophysiology in nonmechanically ventilated patients, ventilator graphics are an important part of understanding the pathophysiology in mechanically ventilated patients. Ventilator graphics have the added advantage of not producing the noise encountered from the oropharynx that occurs with routine pulmonary functions tests as the endotracheal tube bypasses the oropharynx.
Patient-ventilator asynchrony is common and may be seen in up to 25% of patients within 24 hours of initiating mechanical ventilation.1 This article will enable the reader to understand ventilator waveforms and to identify and correct patient-ventilator asynchrony.
The following eMedicine articles and the Medscape CME course may be of interest:
Graphical displays are common in the intensive care setting. The following are examples of graphical displays:
- Ventilator waveforms
- Arterial waveforms
- Venous waveforms (central venous pressure)
- Intracranial pressure waveforms
- Intra-aortic balloon pressure waveforms
One of the many advantages of using graphical displays is that they enable the analyst to better comprehend the behavior of the system being monitored. This is extremely important, especially during mechanical ventilation, because patient-ventilator asynchrony can be better understood and treated. Also important is understanding that more than one asynchrony can be in play and that one form of asynchrony (primary) may subsequently lead to multiple other (secondary) asynchronies. Thus, carefully analyzing the waveforms in a systematic fashion, akin to reading ECG tracings, is critical when analyzing these waveforms. Just as changing the gain and paper speed can help to identify abnormalities in an ECG tracing, changing one or both of the axis scales may be important in identifying abnormalities that would otherwise be missed. See Media Files 1 and 2 below.
 Media File 1. The expiratory breath flow does not return to baseline prior to the initiation of the next breath, suggesting the presence of auto–peak end-expiratory pressure (PEEP). This would have been missed if the flow axis scale had not been changed.
 Media File 2. The presence of ineffective breaths (missed triggers) can be easily missed in both the pressure and flow waveforms in the graphics with a long time axis. With a change in the time axis, these are more readily identifiable. Dividing a mechanical breath into the following 4 components helps to better identify and correct asynchronies: - Trigger phase
- Inspiratory phase
- Cycle phase and
- Expiratory phase
The mechanical breath may be initiated by the patient (patient trigger) or as a function of time (time trigger). The 2 common types of triggering available are pressure and flow.
In order to overcome the considerable effort that could be spent initiating a breath during a pressure trigger, as has been demonstrated in earlier studies,2 flow triggering was introduced. With this mode, the ventilator is triggered when a patient’s effort creates a difference between the inspiratory and expiratory base flow in the circuit. However, with the newer-generation ventilators, these modes are comparable.3, 4 The trigger phase can be divided into the following components: - Trigger pressure (TP) is the pressure that must be attained by patient effort to trigger the mechanical breath.
- Inspiratory trigger time (ITT) is the time elapsed from the patient’s effort to reach the TP set on the mechanical pressure. Thus, in patients with a low respiratory drive, this time could be prolonged and vice versa.
- Rise time to baseline pressure (RTBP) is the time elapsed from the mechanical breath trigger to attain the baseline pressure (trough airway pressure or the PEEP, if set). The patient does not receive any support until the circuit is pressurized to the baseline pressure, and considerable work could be spent during this time if not appropriately set. Inappropriate rise times can also affect the time for the pressure to rise to the peak airway pressure, as is described in the section on flow asynchrony below.
- Inspiratory delay time (IDT) is the total time elapsed from the initial patient effort to the pressurization of the circuit to baseline pressure. In other words, the IDT can be expressed by the following equation: IDT = ITT + RTBP.
Trigger-phase asynchrony
Trigger asynchrony can occur with any mode of mechanical ventilation. Common trigger problems include autotriggering, missed triggering, and double triggering. Appropriate valve sensitivity settings are required to avoid overtly sensitive settings that can lead to autotriggering and insensitive settings that can lead to missed triggering.
Autotriggering is a breath delivered by the ventilator in the absence of patient effort. Autotriggering may be caused by fluid in the circuit, circuit leaks, chest tube leaks, or vibration of the ventilator tubing (as might occur during insufflations and exsufflations of the lungs with poor compliance). Autotriggering can also occur in the following clinical settings: - Low respiratory rate, low respiratory drive, and apnea testing - Allow for low flow in the circuit such that any noise in the system (eg, cardiac oscillations) may trigger a breath
- High cardiac-output states, valvular heart disease,5, 6 and cardiomegaly
 Media File 3. Autotriggering: Representative flow, airway pressure (Paw), esophageal pressure (Pes), and arterial blood pressure (BP) waveforms from a patient who underwent mitral valve replacement and tricuspid annuloplasty for mitral stenosis, tricuspid regurgitation, and aortic regurgitation. With triggering sensitivity set at 1 L/min (left), pressure-support ventilation was activated between 2 synchronized intermittent mandatory ventilation breaths. When trigger sensitivity was changed to 4 L/min (right), pressure-support breaths disappeared and marked oscillation occurred in flow, Paw, and Pes. Cardiogenic oscillation was evaluated as the peak inspiratory flow fluctuation (A), amplitude in the flow oscillation (B), amplitude in airway pressure (C), and amplitude in esophageal pressure (D). Also note that the baseline of esophageal pressure was elevated when autotriggering occurred, suggesting hyperinflation of the lungs.
Ineffective triggering may occur as a manifestation of a primary asynchrony (eg, insensitive valve setting), secondary asynchronies (eg, auto-PEEP leading to missed triggering), or a combination of the two. Although more commonly seen during inspiration, ineffective triggering occurs during both inspiration and expiration. See Media File 4.
 Media File 4. Ineffective triggering: Flow and airway pressure tracings showing ineffective triggering (ie, wasted effort, defined as an airway pressure drop simultaneous to a flow decrease during the expiratory period and not followed by a ventilator cycle), indicating that the patient’s effort was not detected by the ventilator (arrows).
Mechanical ventilation can have a negative impact on the patient's respiratory drive, as has been shown by Kondili et al.7 See Media File 5. Thus, increasing ventilatory support can be associated with ineffective triggering.
 Media File 5. Graded increases in pressure support cause a decrease in respiratory drive (dP/dt), which is associated with considerable increase in the triggering time.
The application of external PEEP has been shown to decrease ineffective triggering in patients with high auto-PEEP.8 External PEEP in this setting reduces the work of breathing needed to trigger the ventilator.9, 10, 11 The most effective method for eliminating ineffective asynchrony in this setting is to reduce the level of ventilator support.12 This study12 also demonstrated that the application of external PEEP reduced but did not eliminate ineffective triggering. The following clinical conditions may predispose to ineffective triggering:
- Increasing pressure support - May be associated with a reduction in respiratory drive and ineffective triggering
- High tidal volumes
- Alkaline pH and increased bicarbonate levels
- Chronic obstructive pulmonary disease and dynamic hyperinflation
The use of in-line nebulizers may lead to ineffective triggering during flow triggering because of the interference with base flow. Thus, use of pressure-trigger settings during nebulizer treatment or the use of ultrasonic nebulizers may avoid this problem.
Double triggering is the delivery of 2 consecutive ventilator cycles separated by a very short expiratory time, with the first cycle being patient triggered. See Media File 6 below. Double triggering is commonly encountered in mechanically ventilated patients.1 Double triggering occurs when the patient’s ventilatory demand is high and the inspiratory time set on the ventilator is too short. That is, double triggering occurs when the ventilator inspiratory time is shorter than the patient’s inspiratory time and thus occurs more commonly in modes with fixed inspiratory flow times, such assist-control ventilation.1
 Media File 6. Flow and airway pressure recordings showing double triggering, defined as 2 consecutive ventilator cycles separated by an expiratory time less than one half the mean inspiratory time. Double triggering occurs when the ventilator inspiratory time is shorter than the patient’s inspiratory time. The patient’s effort is not completed at the end of the first ventilator cycle and triggers a second ventilator cycle.
Double triggering also occurs more commonly in patients whose PaO2/FiO2 (fraction of inspired oxygen) ratio is lower and whose peak inspiratory pressure is higher than in patients without this asynchrony. This situation is commonly seen in patients with acute lung injury or acute respiratory distress syndrome. Thus, as patients with acute lung injury or acute respiratory distress syndrome and low tidal volumes attempt to be successfully ventilated, double triggering can deliver volumes exactly double the intended prescription and thus can potentially result in worse outcomes.13
Decreasing the inspiratory time or increasing the tidal volumes may help with double triggering. If the patient has a variable respiratory drive such that setting a flow on a fixed mode of flow delivery is not adequate, changing to a variable flow (eg, pressure-control ventilation) or a dual-control mode may be helpful. Sedation adjustments may need to be made if all these measures fail. If the patient’s ventilatory need is high or has suddenly changed, it is important to determine the cause of this change (eg, stroke, pulmonary embolus) when making these adjustments.
During the inspiratory phase, the presence of inappropriate flow and patterns can be identified by close inspection of the flow and pressure graphics. Flow may be inadequate or excessive, both of which may contribute to patient-ventilator asynchrony. Importantly, realize that inappropriate flow rates and patterns may lead to a number of secondary asynchronies. For example, inappropriate flow may reduce the expiratory time, leading to auto-PEEP, which may lead to ineffective triggering. Flow can be delivered in the following 3 forms:
- Fixed flow (eg, assisted-controlled ventilation, synchronized intermittent mandatory ventilation)
- Variable flow (eg, pressure-control ventilation)
- Combined fixed and variable flows (dual modes, eg, volume-assured pressure support and pressure augmentation)
Evaluation of the inspiratory-phase asynchrony begins with identifying whether the flow is fixed or variable.
Fixed flow and asynchrony:
Fixed-flow asynchrony can be related to the flow rate, flow pattern, or a combination of the two. Common patterns of fixed-flow delivery include constant flow, descending ramp, or sinusoidal patterns. See Media File 7 below.
 Media File 7. Diagram showing the various flow patterns. Flow rate–related asynchrony
Adequate flow is represented by a smooth, rounded initial part of the pressure graphic with a plateau on the latter half of the pressure curve.
Inadequate flow
Inadequate flow can result in a significant increase in work of breathing. Inspection of the pressure and flow graphics show a "scooped-out" appearance of the pressure waveform and may also show an increase in flow in the direction of the flow if the patient’s effort draws in air through the demand valve. See Media File 8.  Media File 8. Patient on volume-controlled ventilation with a 2-breath sequence. The first breath shows adequate breath delivery. The second breath shows a scooping out of the pressure waveform, suggesting inadequate flow. An increase in flow in the flow waveform in the direction of flow suggests air has been sucked through the demand valve.
Inadequate flow rates may cause undue prolongation of the inspiratory time, leading to shortened expiratory times, which, in turn, may lead to auto-PEEP and ineffective triggering. See Media File 9.
 Media File 9. Airway pressure (Paw), flow, and esophageal pressure (Pes) waveforms from a patient with chronic obstructive pulmonary disease, ventilated with volume assist-control ventilation, with 2 inspiratory flow rates: 30 L/min and 90 L/min. With both flow rates, tidal volume was kept constant (0.55 L). Ineffective inspiratory efforts are indicated by arrows. Increasing the expiratory time (by increasing inspiratory flow at constant tidal volume) decreased dynamic hyperinflation, which reduced the number of ineffective inspiratory efforts, which increased the respiratory rate. During synchronized intermittent mandatory ventilation, the evaluation of the flow rate during the pressure-supported breath may help with setting of the flow rate during the mandatory breath. Excessive flow Excessive flow can be identified on the pressure waveform by the presence of acute "take off" of the ascending limp of the pressure curve along with a pressure spike at the beginning of the curve. See Media File 9 above. Sometimes, the presence of a continued strong patient’s inspiratory effort may give the illusion of excessive flows; however, close inspection of the ascending limb of the pressure curve shows a slow take off of the ascending limb, as opposed to an acute take off that would be expected with excessive flow rates. See Media File 10 below.
 Media File 10. Four-breath sequence of a patient on volume-control ventilation with a constant flow pattern. The first 2 pressure waveforms show a pressure spike at the beginning of the curve and an adequate rise in the pressures, suggesting excessive flow rates. Reduction of the flow rates results in the disappearance of the pressure spike and plateauing of the latter part of the pressure curve.
 Figure 11: Four-breath sequence of a patient on volume-control ventilation with a descending-ramp flow pattern. The first 2 breaths show a pressurelike spike at the beginning of the pressure curve. The last 2-breath delivery is accomplished with increased flow rate, resulting in a pressure pattern almost shadowing the flow pattern. This is an example how an inadequate flow rate in the descending flow pattern breath delivery can masquerade as an excessive flow rate.
Flow pattern–related asynchrony
Certain patterns of flow are used in certain clinical situations. For example, in patients with chronic obstructive pulmonary disease, the descending-ramp flow pattern or variable flow associated with pressure-control ventilation has been shown to be preferable.14, 15 However, care should be taken when switching between patterns, especially when prolonged expiratory times are required, because flow-pattern changes could be accompanied by a prolonging of inspiratory times with shortened expiratory times, resulting in auto-PEEP.
Variable flow and asynchrony
During pressure-control ventilation, the flow is variable. The flow depends on various variables such as respiratory system compliance, set target pressure, and patient effort. The time to target pressure is influenced by the rise-time setting, as is shown in Media File 12.16 When the rise time is adequate, the representative pressure waveform has a rounded front end and a plateau body.
 Media File 12. Flow, volume, airway pressure (Paw), esophageal pressure (Pes), and end-tidal carbon dioxide pressure waveforms recorded in a patient under spontaneous ventilation and under pressure-support ventilation. The pressure support level was 15 cm water. The slope was modulated so that the plateau pressure was reached after a time ranging from 0.1 second (T 0.1) to 1.5 seconds (T 1.5). Each reduction of the value of the pressure ramp slope was associated with an apparently dose-dependent progressive increase in Pes swings, while tidal volume appeared to remain constant.
Excessive rise time appears as a pressure overshoot at the front end of the pressure-time waveform, and inadequate rise time appears as a concave beginning to the pressure-time waveform. Rapid rise times may be associated with premature breath termination and may lead to double triggering if the patient’s effort is sufficient to trigger a mechanical breath.
Inadequate rise time, on the other hand, may lead to prolongation of inspiration, leading to neural asynchrony, and insufficient expiratory time, leading to auto-PEEP and even triggering asynchrony. This again is an example how a primary asynchrony may lead to multiple secondary asynchronies.
The cycling between inspiration and expiration in assist-control ventilation is a function of the preset inspiratory time and tidal volume. During pressure-control/pressure-support breathing, the cycling between inspiration and expiration is brought about by a drop in the flow rate and the breath cycles when the flow reaches a percentage of the peak flow. Other secondary cycling characteristics also are present as a safety precaution if the inspiratory time is unduly prolonged. During assist-control ventilation, prolongation of the inspiratory breath into neural breath termination and expiration can lead to cycle asynchrony. See Media File 13 below. Careful analysis of the pressure and flow curve reveals the presence of a spike at the terminal part of the pressure waveform and may have a “zero” flow in the flow waveform (eg, pressure-regulated volume control) or a sudden decrease in the flow. This spike may also be identified in the pressure volume curve.
 Media File 13. Waveforms of flow, airway pressure (Paw), and transverses abdominis electromyogram in a critically ill patient with chronic obstructive pulmonary disease receiving pressure support of 20 cm water. Expiratory muscle activity (vertical dotted line) began when mechanical inflation was only partly completed. Note the small airway pressure spike near the end of mechanical inflation, which coincides with the patient’s neural expiratory activity.
Overdistension of the lung with large tidal volumes may also show a similar pattern in the pressure-time curve and pressure-volume curve, and differentiation from neural synchrony can be confusing. Termination of insufflations during pressure-control/pressure-support ventilation is usually a function of the peak inspiratory flow or attainment of the peak pressure. Thus, premature termination may occur as a result of the following:
- Excessive rise time results in an initial pressure overshoot with the resultant termination of the breath because the target pressure is exceeded.
- Excessive rise time may also cause early breath termination because the cycling threshold, which is usually a percentage of the peak flow (eg, 25% of the peak flow in the Siemens 300C but varies between ventilator manufacturers and can easily be changed), is reached early and the breath cycles to expiration.
The media files below show the effects of varying rise time on breath cycling in a study by Tokioka et al.17
 Media File 14. Flow (V˙), volume (V), airway pressure (PAW), and esophageal pressure (PES) curves with termination criterion (TC) 5% and TC 35% during 10 cm water of pressure-support ventilation. With TC 5%, the breathing pattern was regular. Tidal volume was 391 mL and respiratory frequency was 17 breaths/min. The negative deflection of PES during inspiration was minimal. With TC 35%, tidal volume decreased to 281 mL and respiratory frequency increased to 23 breaths/min. The inspiratory flow terminated despite continuing negative deflection of PES. Work of breathing increased from 0.20 J/L with TC 5% to 0.32 J/L with TC 35%.
 Media File 15. Flow (V˙), volume (V), airway pressure (PAW), and esophageal pressure (PES) curves with termination criterion (TC) 5% and TC 45% during 10 cm water of pressure-support ventilation. With TC 5%, inspiratory flow terminated simultaneously with the cessation of the patient’s inspiratory effort estimated by PES. In contrast, premature termination with double breathing occurred with TC 45%. Work of breathing also increased from 0.42 J/L with TC 5% to 0.64 J/L with TC 45%.
Shortened expiatory time may lead to auto-PEEP. This can be seen in clinical conditions resulting in slow expiratory time constants, such as chronic obstructive pulmonary disease. In addition, as noted in the previous sections, it can result from other primary asynchronies and, by itself, can result in secondary asynchronies (eg, trigger asynchrony).
Managing auto-PEEP includes correction of any asynchronies, as is described in previous sections, and treatment of the underlying clinical problems.
| Media file 1:
The expiratory breath flow does not return to baseline prior to the initiation of the next breath, suggesting the presence of auto–peak end-expiratory pressure. This would have been missed if the flow axis scale had not been changed. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 2:
The presence of ineffective breaths (missed triggers) can be easily missed in both the pressure and flow waveforms in the graphics with a long time axis. With a change in the time axis, these are more readily identifiable. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 3:
Autotriggering: Representative flow, airway pressure (Paw), esophageal pressure (Pes), and arterial blood pressure (BP) waveforms from a patient who underwent mitral valve replacement and tricuspid annuloplasty for mitral stenosis, tricuspid regurgitation, and aortic regurgitation. With triggering sensitivity set at 1 L/min (left), pressure-support ventilation was activated between 2 synchronized intermittent mandatory ventilation breaths. When trigger sensitivity was changed to 4 L/min (right), pressure-support breaths disappeared and marked oscillation occurred in flow, Paw, and Pes. Cardiogenic oscillation was evaluated as the peak inspiratory flow fluctuation (A), amplitude in the flow oscillation (B), amplitude in airway pressure (C), and amplitude in esophageal pressure (D). Also note that the baseline of esophageal pressure was elevated when autotriggering occurred, suggesting hyperinflation of the lungs. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 4:
Ineffective triggering: Flow and airway pressure tracings showing ineffective triggering (ie, wasted effort, defined as an airway pressure drop simultaneous to a flow decrease during the expiratory period and not followed by a ventilator cycle), indicating that the patient's effort was not detected by the ventilator (arrows). |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 5:
Graded increases in pressure support cause a decrease in respiratory drive (dP/dt), which is associated with considerable increase in the triggering time. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 6:
Flow and airway pressure recordings showing double triggering, defined as 2 consecutive ventilator cycles separated by an expiratory time less than one half the mean inspiratory time. Double triggering occurs when the ventilator inspiratory time is shorter than the patient's inspiratory time. The patient's effort is not completed at the end of the first ventilator cycle and triggers a second ventilator cycle. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 7:
Diagram showing the various flow patterns. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 8:
Patient on volume-controlled ventilation with a 2-breath sequence. The first breath shows adequate breath delivery. The second breath shows a "scooping out" of the pressure waveform, suggesting inadequate flow. An increase in flow in the flow waveform in the direction of flow suggests air has been sucked through the demand valve. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 9:
Airway pressure (Paw), flow, and esophageal pressure (Pes) waveforms from a patient with chronic obstructive pulmonary disease, ventilated with volume assist-control ventilation, with 2 inspiratory flow rates: 30 L/min and 90 L/min. With both flow rates, tidal volume was kept constant (0.55 L). Ineffective inspiratory efforts are indicated by arrows. Increasing the expiratory time (by increasing inspiratory flow at constant tidal volume) decreased dynamic hyperinflation, which reduced the number of ineffective inspiratory efforts, which increased the respiratory rate. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 10:
Four-breath sequence of a patient on volume-control ventilation with a constant flow pattern. The first 2 pressure waveforms show a pressure spike at the beginning of the curve and an adequate rise in the pressures, suggesting excessive flow rates. Reduction of the flow rates results in the disappearance of the pressure spike and plateauing of the latter part of the pressure curve. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 11:
Four-breath sequence of a patient on volume-control ventilation with a descending-ramp flow pattern. The first 2 breaths show a pressurelike spike at the beginning of the pressure curve. The last 2 breath deliveries are accomplished with increased flow rate, resulting in a pressure pattern almost shadowing the flow pattern. This is an example how an inadequate flow rate in the descending flow pattern breath delivery can masquerade as an excessive flow rate. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 12:
Flow, volume, airway pressure (Paw), esophageal pressure (Pes), and end-tidal carbon dioxide pressure waveforms recorded in a patient under spontaneous ventilation and under pressure-support ventilation. The pressure support level was 15 cm water. The slope was modulated so that the plateau pressure was reached after a time ranging from 0.1 second (T 0.1) to 1.5 seconds (T 1.5). Each reduction of the value of the pressure ramp slope was associated with an apparently dose-dependent progressive increase in Pes swings, while tidal volume |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 13:
Waveforms of flow, airway pressure (Paw), and transverses abdominis electromyogram in a critically ill patient with chronic obstructive pulmonary disease receiving pressure support of 20 cm water. Expiratory muscle activity (vertical dotted line) began when mechanical inflation was only partly completed. Note the small airway pressure spike near the end of mechanical inflation, which coincides with the patient's neural expiratory activity. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 14:
Flow (V˙), volume (V), airway pressure (PAW), and esophageal pressure (PES) curves with termination criterion (TC) 5% and TC 35% during 10 cm water of pressure-support ventilation. With TC 5%, the breathing pattern was regular. Tidal volume was 391 mL and respiratory frequency was 17 breaths/min. The negative deflection of PES during inspiration was minimal. With TC 35%, tidal volume decreased to 281 mL and respiratory frequency increased to 23 breaths/min. The inspiratory flow terminated despite continuing negative deflection of PES. Work of breathing increased from 0.20 J/L with TC 5% to 0.32 J/L with TC 35%. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
| Media file 15:
Flow (V˙), volume (V), airway pressure (PAW), and esophageal pressure (PES) curves with termination criterion (TC) 5% and TC 45% during 10 cm water of pressure-support ventilation. With TC 5%, inspiratory flow terminated simultaneously with the cessation of the patient's inspiratory effort estimated by PES. In contrast, premature termination with double breathing occurred with TC 45%. Work of breathing also increased from 0.42 J/L with TC 5% to 0.64 J/L with TC 45%. |
 | View Full Size Image | |
Media type: Rhythm Strip
|
- Thille AW, Rodriguez P, Cabello B, Lellouche F, Brochard L. Patient-ventilator asynchrony during assisted mechanical ventilation. Intensive Care Med. 2006;32:1515–1522. [Medline].
- Sassoon CS, Gruer SE. Characteristics of the ventilator pressure- and flow-trigger variables. Intensive Care Med. Feb 1995;21(2):159-68. [Medline].
- Richard JC, Carlucci A, Breton L, Langlais N, Jaber S, Maggiore S, et al. Bench testing of pressure support ventilation with three different generations of ventilators. Intensive Care Med. Aug 2002;28(8):1049-57. [Medline].
- Calzia E, Lindner KH, Stahl W, Martin A, Radermacher P, Georgieff M. Work of breathing, inspiratory flow response, and expiratory resistance during continuous positive airway pressure with the ventilators EVITA-2, EVITA-4 and SV 300. Intensive Care Med. Sep 1998;24(9):931-8. [Medline].
- Imanaka H, Nishimura M, Takeuchi M, Kimball WR, Yahagi N, Kumon K. Autotriggering caused by cardiogenic oscillation during flow-triggered mechanical ventilation. Crit Care Med. 2000;28(2):402–407. [Medline].
- Hill LL, Pearl RG. Flow triggering, pressure triggering, and autotriggering during mechanical ventilation. Crit Care Med. 2000;28(2):579–581. [Medline].
- Kondili E, Prinianakis G, Georgopoulos D. Patient-ventilator interaction. Br J Anaesth. Jul 2003;91(1):106-19. [Medline].
- Nava S, Bruschi C, Rubini F, Palo A, Iotti G, Braschi A. Respiratory response and inspiratory effort during pressure support ventilation in COPD patients. Intensive Care Med. 195;21:871–879. [Medline].
- Smith TC, Marini JJ. Impact of PEEP on lung mechanics and work of breathing in severe airflow obstruction. J Appl Physiol. Oct 1988;65(4):1488-99. [Medline].
- Mancebo J, Albaladejo P, Touchard D, Bak E, Subirana M, Lemaire F, et al. Airway occlusion pressure to titrate positive end-expiratory pressure in patients with dynamic hyperinflation. Anesthesiology. 2000;93:81–90. [Medline].
- MacIntyre NR, Cheng KC, McConnell R. Applied PEEP during pressure support reduces the inspiratory threshold load of intrinsic PEEP. Chest. Jan 1997;111(1):188-93. [Medline].
- Chao DC, Scheinhorn DJ, Stearn- Hassenpflug M. Patient ventilator trigger asynchrony in prolonged mechanical ventilation. Chest. 1997;112:1592–1599. [Medline].
- The Acute Respiratory Distress Syndrome Network. Ventilation with Lower Tidal Volumes as Compared with Traditional Tidal Volumes for Acute Lung Injury and the Acute Respiratory Distress Syndrome. N Engl J Med. 2000;342:1301-1308. [Medline].
- Georgopoulos D, Kondili E, Prinianakis G. How to set the ventilator in asthma. Monaldi Arch Chest Dis. Feb 2000;55(1):74-83. [Medline].
- Yang SC, Yang SP. Effects of inspiratory flow waveforms on lung mechanics, gas exchange, and respiratory metabolism in COPD patients during mechanical ventilation. Chest. Dec 2002;122(6):2096-104. [Medline].
- Bonmarchand G, Chevron V, Menard JF, Girault C, Moritz-Berthelot F, Pasquis P, et al. Effects of Pressure ramp slope values on the work of breathing during pressure support ventilation in restrictive patients. Crit Care Med. 1999;27(4):715–722. [Medline].
- Tokioka H, Tanaka T, Ishizu T, Fukushima T, Iwaki T, Nakamura Y, et al. The effect of breath termination criterion on breathing patterns and the work of breathing during pressure support ventilation. Anesth Analg. Jan 2001;92(1):161-5. [Medline].
Ventilator Graphics excerpt Article Last Updated: Jun 11, 2008
|