Author: Jack A Shohet, MD, Chairman of Otolaryngology, Hoag Hospital

Jack A Shohet is a member of the following medical societies: American Academy of Otolaryngology-Head and Neck Surgery, American Medical Association, American Neurotology Society, American Tinnitus Association, and California Medical Association

Editors: B Viswanatha, MBBS, MS, DLO, Professor of ENT, Sri Venkateshwara ENT Institute, Victoria Hospital, Bangalore Medical College and Research Institute, India; Francisco Talavera, PharmD, PhD, Senior Pharmacy Editor, eMedicine; Gerard J Gianoli, MD, Clinical Associate Professor, Department of Otolaryngology-Head and Neck Surgery, Tulane University School of Medicine; Vice President, The Ear and Balance Institute; Chief Executive Officer, Ponchartrain Surgery Center; Christopher L Slack, MD, Otolaryngology-Facial Plastic Surgery, Private Practice, Associated Coastal ENT; Medical Director, Treasure Coast Sleep Disorders; Arlen D Meyers, MD, MBA, Professor, Department of Otolaryngology-Head and Neck Surgery, University of Colorado School of Medicine

Author and Editor Disclosure

Synonyms and related keywords: implantable hearing devices, implantable hearing aids, implantable middle ear devices, implantable middle ear hearing devices, cochlear implants, auditory brainstem implants, ABIs, bone-anchored hearing aids, piezoelectric devices, electromagnetic hearing devices

Hearing loss affects up to 10% of the population in the United States. The prevalence increases with age and over one third of people older than 65 years have a significant hearing loss. Only approximately 20% of people with hearing loss seek assistance from hearing aids. Of these, as many as 16.2% do not wear their devices.

Although technical improvements and modifications have improved the fidelity of conventional aids, hearing aids still have many limitations. The aids may be difficult to maintain, requiring frequent cleaning, dehumidification, and battery changes. Some patients may perceive them as being uncomfortable because they simply cannot tolerate an object in the ear canal. Patients often complain of the occlusion effect of an object occupying the entire ear canal. Chronic otitis externa, canal exostoses, or frequent cerumen impactions make it difficult for some to wear hearing aids. Poorly fitting ear molds, faulty circuitry, or canal issues can lead to annoying feedback. Poor sound quality and problems hearing background noise are frequent complaints of those who wear hearing aids. Finally, some patients may perceive a social stigma associated with hearing-aid use.

For excellent patient education resources, visit eMedicine's Ear, Nose, and Throat Center. Also, see eMedicine's patient education article Hearing Loss.



Several types of devices can be considered implantable hearing devices. These include cochlear implants, auditory brainstem implants, and bone-anchored hearing devices.

Cochlear implants

A cochlear implant provides sound perception by means of an electrode surgically implanted into the cochlea. Candidates must have bilateral, profound hearing loss and meet strict audiologic criteria. These devices are covered in other articles of this journal (see Cochlear Implants, Indications and Cochlear Implants, Surgical Technique).

Auditory brainstem implants

Auditory brainstem implants (ABIs) were designed to be used in neurofibromatosis type 2 (NF-2) in which tumors involving complexes of both cranial nerve VII and VIII render the patient anacusic. These devices, which bypass the cochlea and cochlear nerve, are implanted into the lateral recess of the fourth ventricle adjacent to the cochlear nucleus to provide the patient with auditory perception. These devices are usually implanted after the tumor is resected, during the same operation.

House and Hitselberger implanted the first ABIs in 1979. The device was a single-channel percutaneous implant with a ball electrode, which they implanted in approximately 13 patients. The second-generation device incorporated the cochlear implant speech processor (3M/House) and was implanted in 25 patients with NF-2 starting in 1986. In 1992-1993, the first multichannel device (which had an 8-channel array) was developed. In 1994, a multicenter investigation was conducted to evaluate a multichannel ABI powered by a stimulator (Cochlear Corporation Nucleus 22) with transcutaneous signal transmission.

The latest ABI incorporates a digital speech processor (Nucleus 24). It offers 4 user-selectable programs, as well as programmable volume and sensitivity controls. The processor uses the SPEAK speech coding strategy and has the flexibility to add future strategies as they are developed. The device is indicated in patients with NF-2 aged 12 years or older. Implantation may occur during tumor removal on the first or second side or as a separate procedure. The patient should be medically and psychologically suitable, and no audiologic criteria are applied. Because of possible injury to the cochlear nucleus due to radiotherapy, prospective recipients who have undergone gamma-knife irradiation should be considered with extreme caution.

A study on 61 patients receiving the multichannel ABI showed that the device is effective and safe in providing useful auditory sensations in most patients with NF-2.1 The ABI provides enough auditory information to improve lip-reading abilities in most, and a few are able to achieve open-set (no lip-reading cues) speech understanding. Performance may improve for up to 8 years after implantation.

ABIs have been used with varying success in patients born with cochlear nerve aplasia, those with traumatic cochlear-nerve avulsion, and those in whom cochlear implantation is unsuccessful (for salvage treatment).

Bone-anchored hearing devices

The Baha device (Entific Medical Systems) is a percutaneous implantable device primarily used for conductive hearing loss, or more recently, for single-sided sensorineural hearing loss. Developed in Gothenburg, Sweden and used in Europe since 1977, the original device was designed to treat conductive or mixed losses and has become popular for hearing rehabilitation in patients with congenital ear malformations or refractory chronic ear disease. The Baha device can close the air-bone gap to within 10 dB of the preoperative bone-conduction thresholds in as many as 80% of patients and to within 5 dB in 60%.

An osseointegrated titanium fixture with a percutaneous abutment is implanted in the postauricular area. An external sound processor is attached to the abutment at will. A microphone in the processor, which vibrates the bone in the skull by means of the fixture, picks up sound. The sound is transmitted directly to the inner ear on the side with conductive hearing loss or better sensorineural hearing than the other.

The external processor is available in 3 sizes and is chosen depending on the bone thresholds of the ear to be aided. A body-worn external processor allows for the implantation for pure-tone-average bone thresholds as low as 70 dB HL. The 2 small processors are for pure-tone-average bone thresholds better than 45 dB HL. The Baha device can be placed bilaterally in patients with bilateral disorders to allow for sound localization and to improve speech recognition in noise.

The surgery is typically performed in a single stage in adults. About 3-4 months is required to allow for osseointegration before the external processor can be attached and its benefits realized. A 2-stage procedure is recommended in children in whom the fixture is placed into the bone at the first stage. After 3-6 months to allow for osseointegration, a second-stage operation is done to connect the abutment through the skin to the fixture. Complications are few and mostly limited to local infection and inflammation at the implant site and a failure of osseointegration.

The US Food and Drug Administration (FDA) approved the Baha device for use in children aged 5 years or older, but has been used successfully in Europe in children as young as 1.5 years. Indeed, one group with considerable experience recommends that the most suitable age is 2-4 years. A bone-augmentation technique may aid implantation in children of this age.

Before the FDA approved this device for single-sided deafness, the contralateral routing of signal (CROS) hearing aid was the only option available for rehabilitation. Poor performance and aesthetic considerations limited the use of CROS aids. The Baha device can be implanted on the side of the deaf ear, and it transmits the sound by means of bone conduction to the contralateral cochlea. This process eliminates the head-shadow effect and allows for hearing from both sides of the head. The Baha aid substantially improves speech recognition in quiet and in composite noise compared with the CROS aid.



The implantable middle-ear hearing devices discussed in this article were developed to treat conductive and sensorineural hearing loss. Extensive trials are in progress to determine their efficacy.

The rationale for development of implantable middle-ear devices is multifaceted. These devices improve fidelity by directly stimulating the ossicles, and they improve comfort by allowing the ear canal to remain open. In addition, most implantable middle-ear devices almost completely eliminate feedback, one of the most annoying adverse effects of conventional aids. Improved cosmesis by means of miniaturization and concealment of the components is another benefit. Finally, some devices may allow the patient to continue receiving amplification while swimming or bathing.



Implantable middle-ear devices are generally available in 2 types: piezoelectric and electromagnetic.

Piezoelectric Devices

Piezoelectric devices operate by passing an electric current into a piezoceramic crystal, which changes its volume and thereby produce a vibratory signal. The major disadvantage is that power output is directly related to the size of the crystal. Studies of early designs indicate that such an approach benefits only people with up to moderate (about 60 dB) hearing loss. These are the same individuals who benefit from the small completely-in-the-canal (CIC) type of hearing aid. A CIC hearing aid may be most useful if cosmetics are the primary concern. Piezoelectric transducers have the advantage of being inert in a magnetic field and therefore compatible with MRI.

Rion device, E-type

One of the earliest piezoelectric devices, the Rion Device E-type (RDE), has been used for both conductive and sensorineural losses. It is a partially implanted device composed of an external, ear-level microphone and amplifier, and an internal electromagnetic coil and vibrator element. The piezoelectric vibrator element is anchored to the squamous portion of the temporal bone with a titanium screw. It is attached to the stapes with a hydroxyapatite tube, which is interposed between the tip of the vibrator and the head of the stapes.

In 39 patients in Japan, an initial hearing improvement of 36 dB at 3 months after surgery eventually decreased to 21 dB with long-term follow-up.2 The reason for diminished performance was thought to be to a decrease in the sensitivity of the ossicular vibrator caused by aging and tissue reaction around the vibrator element. The authors reported that sensorineural hearing was not affected and that all patients preferred the device to a conventional aid.

Totally integrated cochlear amplifier

The Totally Integrated Cochlear Amplifier (TICA; Implex American Hearing Systems, now owned by Cochlear Corporation), is totally implantable. The microphone is implanted subcutaneously in the external ear adjacent to the tympanic membrane. A digitally programmable processor located subcutaneously on the mastoid bone processes the signal. A piezoelectric transducer is coupled to the body of the incus and drives the ossicular chain by vibratory actions. Many European and some North American research programs have used this approach.

The TICA received the C.E. mark in Europe in the late 1990s but have not undergone studies in the United States to date.

Middle ear transducer

The Middle Ear Transducer (MET; Otologics LLC, Boulder, CO) was first introduced as a semi-implantable device but was recently made totally implantable. The original device was composed of an implanted transducer mounted in a laser-drilled hole in the body of the incus. The transducer translates the electrical signals into a mechanical motion that directly stimulates the ossicles and enables the wearer to perceive sound. The transducer is coupled with an externally worn audio processor (Button processor), which contains the microphone, battery, and signal processor. The device has a CE mark for pan-European approval to market the device. To date, more than 300 patients have received the implant in Europe and in the United States.

The MET Fully-Implantable Ossicular Stimulator (Otologics) consists of 4 primary components: the implant, the programming system, the charger, and the remote control. The implant component consists of 2 main parts: the electronics capsule and the MET. The electronics capsule contains the microphone, battery, magnet, digital signal processor, and connector. A sensitive microphone located under the skin picks up sounds, which are amplified according to the wearer's needs, and converted into an electrical signal. The signal is sent down the lead and into the transducer, which is the same as that used in the semi-implantable version. The semi-implanted version can be upgraded to the fully implanted version in a single surgical procedure by using local anesthesia.

The programming system coil is placed over the implant site and held in place magnetically. The coil couples with the implant by means of a radiofrequency signal that is used to program the device in the same manner as a traditional digital hearing aid. The programming system also allows for extensive testing and diagnostics of the stimulator.

The charger system consists of the base station, charging coil, and charger body. To charge the implant, the wearer removes the charger body from the base station and places the coil on the skin, over the implant site. The charger body contains a clip that allows the charger to be attached to the belt of the wearer during charging. Typically, charging time will be about one hour if performed daily. While recharging the implant, the wearer can perform normal daily activities, turn the implant on and off, and adjust the volume.

A remote is used to control the stimulator when the device is not being charged. The remote allows the wearer to turn the implant on and off, and to adjust the volume. To use the remote control, the wearer holds the remote against the skin over the implant.

The US Phase I trial results yielded a 15-20 dB functional gain across audiometric frequencies in 20 patients. The pure-tone averages and monaural word recognition scores were better with the hearing aid in the same ear preoperatively, whereas the patients generally perceived more benefit in the postoperatively implant-aided conditions.3 

Envoy system

Another totally implantable piezoelectric device is the Esteem by Envoy Medical (originally St Croix Medical). This device uses the eardrum as the microphone, taking advantage of the natural acoustics of the ear canal without obstruction, interference, or any external devices. Therefore, the input signals are identical to those received by a person with normal hearing. This mechanical signal is detected from a piezoelectric transducer at the head of the incus (the sensor) and converted to an electrical signal by using existing transducer technology. The electrical signal is amplified, filtered, and converted back to a vibratory signal. The processed vibratory signal is then delivered by means of a piezoelectric transducer (the driver) attached to the stapes capitulum. The incus lenticular process is removed to prevent feedback to the sensor. The newly designed piezoelectric transducer can provide an output close to 110 dB sound pressure level (SPL).

An audiologist programs the implant using a device called the commander. After the device is programmed, patients are given a personal programmer that allows them to turn the device on or off, to adjust the volume, and to remotely modify background noise filters.

The advantages of such a device are notable. Without any appliance in the external auditory canal, the occlusion effect is eliminated. Uncoupling of the sensor and driver eliminate most feedback. Even more important for some is the fact that the device is completely concealed in the body.

The Envoy device faces some hurdles as it is further developed. Its battery life is an issue, as with any totally implantable device. The battery has an estimated life of 3-5 years depending on use and can be replaced by using local anesthetic. In addition, removal of a portion of the incus permanently alters the ossicular mechanism and prohibits full recovery of hearing to preimplantation baseline levels if the device fails or is in the off position. Modern ossicular reconstruction techniques may effectively restore hearing to within 10 dB. Finally, functional gain decreased by more than 3000 Hz in the phase I study of the Envoy device. This appears to have been substantially improved in the phase II studies.

The phase I clinical trial included 7 patients, 5 of whom had working implants in the 2-month postactivation period. The remaining 2 had no benefit from the device as a result in a breach of the hermetic sealing of the device, which allowed moisture to pass into the circuitry. This breach was corrected in the current version of the device. Overall, the patients with functioning devices perceived a benefit with the Envoy device compared with their hearing aid in terms of ease of communication in favorable conditions, background-noise reverberation, and aversiveness of sound. Speech discrimination was markedly improved over hearing-aid conditions by 17%. Functional gain and speech reception thresholds were similar for the Envoy device and for the hearing aids.

As of 2007, the US Phase II trial is underway. Over 70 patients have been implanted in this part of the trial. Overseas, the Esteem received the C.E. mark in 2006, and it is currently being used in several countries.

Electromagnetic Hearing Devices

Electromagnetic hearing devices function by passing an electric current into a coil, which creates a magnetic flux that drives an adjacent magnet. The small, 50-mg magnet is attached to one of the vibratory structures of the middle ear (eg, tympanic membrane, ossicles). To date, all of the research programs using this method use devices that are only partially implantable and that still require an external hearing-aid shell to house the electric coil. In some cases, the external coil can be housed in a CIC type of hearing-aid shell. The major disadvantage is that power is decreased by the square of the distance between the coil and the magnet; therefore, the coil and magnet must be close. A slight shift of coil position in the external ear results in unpredictable or insufficient power output. Furthermore, the anatomy of the middle ear space restricts the size of the magnet or coil.

Vibrant Soundbridge device

One example of an electromagnetic device is the Vibrant Soundbridge device. Originally developed by Symphonix Devices, Inc., the Soundbridge was the first FDA-approved implantable middle-ear hearing device to treat sensorineural hearing loss. It was marketed and implanted in the United States for a few years until the technology was purchased by Med-El of Austria. It is now marketed by Vibrant Med-El and implanted in Europe, but it is currently awaiting FDA approval for reintroduction to the US market. Over 1400 such devices have been implanted worldwide.

The Soundbridge device is a semi-implantable device composed of an external sound processor and amplifier, an audio processor, and an internal vibrating ossicular prosthesis (VORP). Sound passes into a microphone on the postauricular audio processor and is transmitted through the skin to an implanted receiver on the VORP. The VORP, which is implanted postauricularly (similar to a cochlear implant) conducts the sound to a magnet surrounded by a coil called the floating mass transducer (FMT). The transducer is attached to the long process of the incus and the magnet hugs the long axis of the stapes, which causes it to vibrate. One of the disadvantages of this device is that the confined space of the middle ear restricts the dimensions and crucial mass of the transducer, limiting power output.

The phase III FDA trial was completed in 2000, and results from the 53 patients submitted to the FDA were published in 2002.4 The device was safe, with no notable change in preoperative and postoperative bone thresholds. Furthermore, functional gain and word-recognition scores were improved with the Vibrant device compared with the patients' conventional hearing aids. Self-assessment inventories indicated that 94% of the patients believed that the overall sound quality of the Vibrant Soundbridge device was better than that with their conventional hearing aid.

As of 2007, the use of the Soundbridge has been expanded. It was successfully implanted on the round window membrane in patients with aural atresia and mixed hearing losses.5, 6, 7 The Soundbridge was also applied to the incus in a more traditional sense in patients with otosclerosis.8, 9

The audio processor has evolved from the original, analog Vibrant P unit to a digital, 3-channel Vibrant D unit to the current digital, 8-channel Vibrant Signia unit. The Signia device modestly increases functional gain and speech-in-noise understanding results compared with the Vibrant D device.

Soundtec direct system

The Soundtec implant was introduced to the US market in 2001 and voluntarily withdrawn in 2004. This semi-implantable device converts sound energy to electromagnetic energy to directly stimulate the ossicles. A surgically implanted neodymium-iron-boron (NdFeB) magnet is attached to the ossicular chain by positioning a collar around the neck of the stapes. An earmold coil assembly consisting of an acrylic skeleton mold with an embedded electromagnetic coil stimulates the magnet. The earmold coil assembly is inserted deeply into the ear canal, ideally approximately 2 mm away from the tympanic membrane. It is attached to a sound processor that is fit either in the canal or behind the ear, similar to a hearing aid.

Such a design offers several possible advantages. Because it works by electromagnetic energy through the ear canal, the Soundtec device does not require an acoustic seal, which may lead to the occlusion effect or alter the resonance qualities of the ear canal. Also, functional gain can be improved without necessarily precipitating feedback, a common problem with traditional aids that occurs when sound pressure escapes the ear canal and cycles back through the microphone.

A relative disadvantage is that the procedure requires separation and then reconstitution of the incudostapedial joint, a maneuver that may be responsible for as much as a 4.2-dB increase in air-conduction thresholds after surgery. Subsequent studies further demonstrated a loss of average bone conduction of pure tone after implantation. This effect may be a result of movement of the mobile stapes into the vestibule during disarticulation of the incudostapedial joint, which may cause sensorineural hearing loss.

Early reports indicated that, though the Soundtec did not provide a significant difference from optimal traditional amplification, it provided statistically significant high-frequency functional gain. Furthermore, patients' subjective reports indicated a cleaner, more natural sound without feedback than that achieved with traditional amplification.

The phase II FDA clinical trial included 103 patients with moderate to moderately severe sensorineural hearing loss who had previously worn hearing aids for at least 45 days.10 An average 7.9-dB increase in functional gain in the speech frequencies and a 9.6-dB increase in the high frequencies was reported, as was a statistically significant increase of 5.3% in speech discrimination. Furthermore, subjective measures of feedback, occlusive effect, perceived aided benefit, patient satisfaction, and device preference over the patient's optimally fit hearing aid were demonstrated.

A long-term follow-up retrospective review of 64 patients receiving the Soundtec device revealed a significant average functional gain of 26 dB.11 About 55% of patients complained of hearing the magnet move when the processor was not being worn. This effect was improved, but not completely eliminated, by further stabilizing the implant by placing adipose tissue between the implant collar and the neck of the stapes. A patient questionnaire to assess sound quality, speech in noise, and satisfaction with the Soundtec compared with conventional hearing aids failed to show a significant difference. The authors concluded that the ideal patient is one younger than 70 years with a moderate sensorineural hearing loss, speech discrimination scores equal or better than 60%, appropriately sized ear canals, and sufficient manual dexterity to insert the processor.

To date, around 600 Soundtec devices have been implanted, with the most users in the United States. The device was voluntarily withdrawn from the market in 2004 when the company identified ways to improve it and to eliminate the distortion some patients experienced. The distorted sound, which occurred primarily when the external processor was not used, occurred in as many as 7%. It is thought to occur from movement of the magnet around its single point of fixation with the ossicular chain. The anticipated new magnet will have an additional point of fixation onto the ossicular chain to further stabilize it. Furthermore, the company is looking into redesigning the sound processor with digital circuitry and making variable-sized coils, which may be interchangeable, to allow for tailoring of the coil length to the external auditory canal. The company expects the device to be back on the US market in approximately 1 year (personal communication, Dr J.V.D. Hough, May 15, 2005, Boca Raton, FL).

Additional studies of the Soundtec in a magnetic field indicate that it should be mechanically stable and nondestructive during 0.3-T open MRI with a modified MRI protocol.12, 13

Semi-implantable middle-ear electromagnetic hearing device

Another electromagnetic device is the semi-implantable middle-ear electromagnetic hearing device (SIMEHD). This device consists of an external unit, which is similar in appearance to a behind-the-ear hearing aid, and an internal device, which is anchored to the mastoid. The internal device contains a driving coil and a magnet cemented to the incus.



A few major biomechanical issues must be addressed in developing implantable middle-ear devices. First, the device should not affect normal function of the middle ear. In the optimal situation, the device should not alter air-conduction thresholds. If the device is unsuccessful, the added mass of the unit attached to the vibratory structure of the middle ear should not affect that structure's ability to vibrate.

Another important issue is the anchoring of the device to the ossicular chain. Even a little laxity at the interface between the prosthesis and bone could diminish the transmitted power enough to render the device ineffective. Long-term stability of the fixation must be considered as well. The mechanical forces acting at the interface could affect the life expectancy of the device.

Finally, the direction of the transducer's vibration must be coincident with the axis of normal sound transmission through the ossicular chain. In effect, the device must be attached to the tympanic membrane, the long process of the incus, or the head of the stapes. Otherwise, the transmitted force is reduced.



In general, candidates for implantable middle-ear devices should have tried conventional aids with no success. Some devices were designed for patients with mixed hearing loss. Candidates may also have a conductive hearing loss with an irreversibly disabled middle ear, as with severe tympanosclerosis or chronic eustachian tube dysfunction. The Japanese Rion device was designed for this group of patients.

Most current devices are designed for patients with mild-to-severe sensorineural hearing loss. Mild-to-moderate loss is usually adequately improved with conventional aids. Severe cases are limited by feedback. Given adequate output production, an implantable aid can amplify a severe loss and avoid feedback.



Many challenges remain in developing the ideal implantable hearing device. Limitations in the capacity and recharging cycles of available batteries necessitate transducer energy efficiency. Restrictions due to the size of the middle ear challenge the device to deliver enough gain to aid a severe hearing loss. Costs associated with the development of these devices, as well as surgical implantation, make these devices more expensive than conventional hearing aids. This cost may limit their widespread acceptance.

Another hurdle is in measuring the devices' perceived benefit. All implantable middle ear devices have shown benefit over hearing aids in the patient's perception of the sound fidelity, a very important quality. These implants rely on the Abbreviated Profile of Hearing Aid Benefit (APHAB), a subjective questionnaire of sound qualities that patients complete preoperatively with their hearing aid and postoperatively with the implant. However, quantitatively measuring the psychoacoustic qualities of sound that are so important to the patient is difficult.



Some devices are currently being studied in FDA trials, but the FDA has formally approved only the Vibrant Soundbridge and Soundtec direct systems so far. Many believe that these devices may represent a new era in hearing augmentation akin to the largely successful, but more narrowly indicated, cochlear implants. Although several hurdles still remain, the potential advantages and increasing number of people who may benefit from such devices continue to support their development.



Media file 1:  Auditory brainstem implant (Cochlear Corporation).
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 2:  Direct bone conduction with the Baha system.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 3:  Abutment of the Baha device.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 4:  External processor of the Baha device.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 5:  Totally Integrated Cochlear Amplifier (TICA; Implex American Hearing Systems, now owned by Cochlear Corporation).
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 6:  Processor and transducers of the Envoy device.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 7:  Envoy device implanted.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 8:  Vibrant Soundbridge device.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 9:  Soundtec Direct System.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 10:  Semi-implantable middle-ear transducer (Otologics).
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image

Media file 11:  Middle Ear Transducer (MET; Otologics), a fully implantable ossicular stimulator.
Click to see larger pictureClick to see detailView Full Size Image
Media type:  Image



  1. Otto SR, Brackmann DE, Hitselberger WE, Shannon RV, Kuchta J. Multichannel auditory brainstem implant: update on performance in 61 patients. J Neurosurg. Jun 2002;96(6):1063-71. [Medline].
  2. Yanagihara N, Sato H, Hinohira Y, Gyo K, Hori K. Long-term results using a piezoelectric semi-implantable middle ear hearing device: the Rion Device E-type. Otolaryngol Clin North Am. Apr 2001;34(2):389-400. [Medline].
  3. Jenkins HA, Niparko JK, Slattery WH, Neely JG, Fredrickson JM. Otologics Middle Ear Transducer Ossicular Stimulator: performance results with varying degrees of sensorineural hearing loss. Acta Otolaryngol. May 2004;124(4):391-4. [Medline].
  4. Luetje CM, Brackman D, Balkany TJ, Maw J, Baker RS, Kelsall D. Phase III clinical trial results with the Vibrant Soundbridge implantable middle ear hearing device: a prospective controlled multicenter study. Otolaryngol Head Neck Surg. Feb 2002;126(2):97-107. [Medline].
  5. Kiefer J, Arnold W, Staudenmaier R. Round window stimulation with an implantable hearing aid (Soundbridge) combined with autogenous reconstruction of the auricle - a new approach. ORL J Otorhinolaryngol Relat Spec. 2006;68(6):378-85. [Medline].
  6. Wollenberg B, Beltrame M, Schönweiler R, Gehrking E, Nitsch S, Steffen A. [Integration of the active middle ear implant Vibrant Soundbridge in total auricular reconstruction]. HNO. May 2007;55(5):349-56. [Medline].
  7. Colletti V, Carner M, Fiorino F, Sacchetto L, Miorelli V, Orsi A. Hearing restoration with auditory brainstem implant in three children with cochlear nerve aplasia. Otol Neurotol. Sep 2002;23(5):682-93. [Medline].
  8. Venail F, Lavieille JP, Meller R, Deveze A, Tardivet L, Magnan J. New perspectives for middle ear implants: first results in otosclerosis with mixed hearing loss. Laryngoscope. Mar 2007;117(3):552-5. [Medline].
  9. Dumon T. Vibrant soundbridge middle ear implant in otosclerosis: technique - indication. Adv Otorhinolaryngol. 2007;65:320-2. [Medline].
  10. Hough JV, Matthews P, Wood MW, Dyer RK Jr. Middle ear electromagnetic semi-implantable hearing device: results of the phase II SOUNDTEC direct system clinical trial. Otol Neurotol. Nov 2002;23(6):895-903. [Medline].
  11. Silverstein H, Atkins J, Thompson JH Jr, Gilman N. Experience with the SOUNDTEC implantable hearing aid. Otol Neurotol. Mar 2005;26(2):211-7. [Medline].
  12. Dyer RK Jr, Dormer KJ, Hough JV, Nakmali U, Wickersham R. Biomechanical influences of magnetic resonance imaging on the SOUNDTEC Direct System implant. Otolaryngol Head Neck Surg. Dec 2002;127(6):520-30. [Medline].
  13. Dyer RK Jr, Nakmali D, Dormer KJ. Magnetic resonance imaging compatibility and safety of the SOUNDTEC Direct System. Laryngoscope. Aug 2006;116(8):1321-33. [Medline].
  14. Chen DA, Backous DD, Arriaga MA, Garvin R, Kobylek D, Littman T. Phase 1 clinical trial results of the Envoy System: a totally implantable middle ear device for sensorineural hearing loss. Otolaryngol Head Neck Surg. Dec 2004;131(6):904-16. [Medline].
  15. Colletti V, Carner M, Miorelli V, Colletti L, Guida M, Fiorino F. Auditory brainstem implant in posttraumatic cochlear nerve avulsion. Audiol Neurootol. Jul-Aug 2004;9(4):247-55. [Medline].
  16. Colletti V, Fiorino FG, Carner M, Miorelli V, Guida M, Colletti L. Auditory brainstem implant as a salvage treatment after unsuccessful cochlear implantation. Otol Neurotol. Jul 2004;25(4):485-96; discussion 496. [Medline].
  17. Colletti V, Soli SD, Carner M, Colletti L. Treatment of mixed hearing losses via implantation of a vibratory transducer on the round window. Int J Audiol. Oct 2006;45(10):600-8. [Medline].
  18. Granstrom G, Tjellstrom A. Guided tissue generation in the temporal bone. Ann Otol Rhinol Laryngol. Apr 1999;108(4):349-54. [Medline].
  19. Huttenbrink KB. Current status and critical reflections on implantable hearing aids. Am J Otol. Jul 1999;20(4):409-15. [Medline].
  20. Jenkins HA, Atkins JS, Horlbeck D, Hoffer ME, Balough B, Arigo JV. U.S. Phase I preliminary results of use of the Otologics MET Fully-Implantable Ossicular Stimulator. Otolaryngol Head Neck Surg. Aug 2007;137(2):206-12. [Medline].
  21. Kochkin S, MarkeTrak V. Why my hearing aids are in the drawer: the consumer's perspective. Hear Journal. 2000;53:34-42.
  22. Lustig LR, Arts HA, Brackmann DE, Francis HF, Molony T, Megerian CA. Hearing rehabilitation using the BAHA bone-anchored hearing aid: results in 40 patients. Otol Neurotol. May 2001;22(3):328-34. [Medline].
  23. Maniglia AJ, Ko WH, Garverick SL, Abbass H, Kane M, Rosenbaum M. Semi-implantable middle ear electromagnetic hearing device for sensorineural hearing loss. Ear Nose Throat J. May 1997;76(5):333-8, 340-1. [Medline].
  24. Niparko JK, Cox KM, Lustig LR. Comparison of the bone anchored hearing aid implantable hearing device with contralateral routing of offside signal amplification in the rehabilitation of unilateral deafness. Otol Neurotol. Jan 2003;24(1):73-8. [Medline].
  25. Priwin C, Granstrom G. The bone-anchored hearing aid in children: a surgical and questionnaire follow-up study. Otolaryngol Head Neck Surg. Apr 2005;132(4):559-65. [Medline].
  26. Priwin C, Stenfelt S, Granström G, Tjellström A, Håkansson B. Bilateral bone-anchored hearing aids (BAHAs): an audiometric evaluation. Laryngoscope. Jan 2004;114(1):77-84. [Medline].
  27. Roland PS, Shoup AG, Shea MC, Richey HS, Jones DB. Verification of improved patient outcomes with a partially implantable hearing aid, The SOUNDTEC direct hearing system. Laryngoscope. Oct 2001;111(10):1682-6. [Medline].
  28. Snik AF, Cremers CW. First audiometric results with the Vibrant soundbridge, a semi-implantable hearing device for sensorineural hearing loss. Audiology. Nov-Dec 1999;38(6):335-8. [Medline].
  29. Snik AF, Dreschler WA, Tange RA, Cremers CW. Short- and long-term results with implantable transcutaneous and percutaneous bone-conduction devices. Arch Otolaryngol Head Neck Surg. Mar 1998;124(3):265-8. [Medline].
  30. Snik AF, Mylanus EA, Cremers CW. Implantable hearing devices for sensorineural hearing loss: a review of the audiometric data. Clin Otolaryngol. Oct 1998;23(5):414-9. [Medline].
  31. Todt I, Seidl RO, Ernst A. Hearing Benefit of Patients after Vibrant Soundbridge Implantation. ORL J Otorhinolaryngol Relat Spec. Aug 2 2005;67(4):203-206. [Medline].
  32. Wazen JJ, Caruso M, Tjellstrom A. Long-term results with the titanium bone-anchored hearing aid: the U.S. experience. Am J Otol. Nov 1998;19(6):737-41. [Medline].
  33. Zenner HP, Leysieffer H, Maassen M, Lehner R, Lenarz T, Baumann J. Human studies of a piezoelectric transducer and a microphone for a totally implantable electronic hearing device. Am J Otol. Mar 2000;21(2):196-204. [Medline].

Implantable Hearing Devices excerpt

Article Last Updated: Feb 14, 2008