Suture Care Introduction




Author: Margaret Terhune, MD, Consulting Staff, Department of Dermatology, Lexington Clinic East

Margaret Terhune is a member of the following medical societies: Phi Beta Kappa

Editors: Désirée Ratner, MD, Director of Dermatologic Surgery, George Henry Fox Assistant Clinical Professor, Department of Dermatology, Columbia Presbyterian Medical Center, New York Presbyterian Hospital; David F Butler, MD, Professor of Dermatology, Texas A&M University College of Medicine; Director, Division of Dermatology, Scott and White Clinic; Director Dermatology Residency Training Program, Scott and White Clinic; Jeffrey P Callen, MD, Professor of Medicine, Chief, Division of Dermatology, University of Louisville School of Medicine; Glen H Crawford, MD, Assistant Clinical Professor, Department of Dermatology, University of Pennsylvania School of Medicine; Chief, Division of Dermatology, The Pennsylvania Hospital; Dirk M Elston, MD, Director, Department of Dermatology, Geisinger Medical Center

Author and Editor Disclosure

Synonyms and related keywords: sutures, needles, staples, tapes, tissue adhesives, tissue healing, absorbable sutures, nonabsorbable sutures, multifilament sutures, monofilament sutures, natural sutures, synthetic sutures, surgical gut sutures, chromic gut sutures, fast-absorbing gut sutures, polydioxanone, polyglycolic acid, polyglactin, polytrimethylene carbonate, polyglecaprone, silk, nylon, polyester, polybutester, polypropylene, skin wound, laser soldering, laser welding, tissue laser soldering, tissue laser welding

Dermatologic wounds can heal by second intention, or they can be closed by a variety of methods. Although the skill and technique of the surgeon are important, so is the choice of wound closure materials. The purpose of these materials is to maintain wound closure until a wound is strong enough to withstand daily tensile forces and to enhance wound healing when the wound is most vulnerable.

In order to fully appreciate the essentials of a wound closure, it is first important to understand the process of wound healing. Healing occurs in 3 stages: inflammation, tissue formation, and matrix formation and remodeling. The inflammation stage begins within minutes after an injury. This stage is necessary to establish hemostasis and to begin mobilization of the immune system. The second stage is the tissue formation stage in which reepithelialization, angiogenesis, and fibroblast proliferation and migration predominate. During the final stage, the extracellular matrix, which is composed of fibronectin, hyaluronic acid, proteoglycans, and type III collagen, is deposited and constantly altered with the final accumulation of mature type I collagen. This stage may occur for as long as 6-12 months after wounding.

Wound strength gradually increases during the healing process. After 2 weeks, a wound has less than 10% of its final healed strength. By this time, most superficial or percutaneous closure materials are removed, and the resulting wound has little to rely on for strength unless additional support is available. Wound strength increases to 20% by 3 weeks and to 50% by 4 weeks. At 3-6 months, a wound achieves its maximum strength, which is 70-80% that of normal skin.



The choice of a particular suture material should be based on the patient, wound, tissue characteristics, and anatomic location. A surgeon's selection may not be specifically based on scientific data, but rather, on the preferences that he or she learned from mentors and/or in training.

Understanding the various characteristics of available suture materials is important to make an educated selection. No one suture possesses all desirable characteristics. The optimal suture should be easy to handle and have high tensile strength and knot security. Any tissue reaction should be minimal, and the material should resist infection and have good elasticity and plasticity to accommodate wound swelling. A low cost is preferred. Although some of the newer materials available have many of these properties, no one material is ideal and compromises must be made.

The physical characteristics of a suture material determine its utility; these characteristics include configuration, diameter, capillarity and fluid absorption, tensile strength, knot strength, elasticity, plasticity, and memory.

Configuration

The configuration of a suture is based on the number of strands of material used to fabricate it. A suture can be monofilament (ie, single-stranded) or multifilament (ie, multistranded). Multifilamentous sutures are twisted or braided.

Sizes

United States Pharmacopeia (USP) sizes are standardized and related to a specific diameter range (in millimeters) that is necessary to produce a certain tensile strength. These diameter ranges vary with the different categories of suture material. These categories include natural materials, synthetic absorbable materials, and synthetic nonabsorbable materials. Sizes are expressed with zeroes, such as 3-0, 4-0, 5-0, and 6-0; more zeroes indicate a smaller size.

Tensile strength

The tensile strength of a material is determined by the weight required to break a suture divided by its cross-sectional area. The rate of tensile strength loss is not the same as absorption and varies among suture materials. The implantation and tying of a suture decreases its strength. Dry, unused absorbable suture loses 4-13% of its initial strength after being soaked in sodium chloride solution for 24 hours. Knotted sutures have two-thirds the strength of unknotted sutures. In selecting suture material, the tensile strength of a suture does not need to exceed that of the tissue it is securing.

Knots

The knot is the weakest portion of the suture. Its strength is defined by the force necessary to cause slippage. The 2 types of knots used in dermatologic surgery are flat knots and sliding knots, with flat knots being more secure than sliding knots. Flat knots include square knots and surgeon's knots. A surgeon's knot differs from a square knot in the initial throw. Two wraps around the needle driver define the surgeon's knot. Although this initial throw adds no strength to the knot, it decreases the tendency of the wound to separate as the suture is tied. Three-throw flat knots are appropriate for use in dermatologic surgery; however, the suture material may affect the number of throws needed for security. Extra throws do not increase the strength of a properly tied knot, but they do add to its bulk and, therefore, to any potential tissue reaction, especially when buried.

Plasticity and elasticity

Plasticity is the ability of the suture to retain its new form and length after stretching. Plasticity allows a suture to accommodate wound swelling, thereby decreasing the risk of strangulated tissue and crosshatch marks. However, as swelling subsides, the suture retains its new size and may not continue to adequately approximate the wound edges.

Elasticity is the ability of a suture to regain its original form and length after stretching. After the swelling of a wound recedes, the suture returns to its original length and keeps the wound well approximated. Most sutures are elastic; few are plastic.

Memory is the ability of a suture to return to its original shape after deformation by tying. Memory is related to plasticity and elasticity. Sutures with a high degree of memory, particularly monofilament sutures, are stiff and difficult to handle; the knots are less secure and may require an extra throw to prevent loosening.

Handling characteristics

Handling characteristics of suture materials are defined by pliability and coefficient of friction. Pliability refers to the ease with which a suture can be bent. The coefficient of friction is a measure of the slipperiness of the suture. Sutures with a high coefficient of friction, generally multifilament sutures, are more difficult to pass through tissue, thereby causing a greater degree of tissue injury (and potential discomfort) during placement and removal. However, these sutures are more pliable, being easier to handle and manipulate for tying knots.

Tissue reaction

Different suture materials produce varying degrees of tissue reaction, specifically inflammation. Significant inflammation reduces the resistance to infection and delays the onset of wound healing. The type of material and size of the suture are thought to be the major factors contributing to this reaction. Natural materials are absorbed by proteolysis, which causes a prominent inflammatory response, while synthetic materials are absorbed by hydrolysis, which produces a minimal reaction.

Multifilamentous sutures have a high degree of capillarity, which is correlated with a tendency to absorb and retain both fluid and bacteria. These materials are associated with greater reactivity and may promote infection if bacterial contamination occurs during or shortly after surgery. Smit et al1 histologically evaluated tissue reactivity to different suture materials (ie, silk, plain and chromic catgut, polyester, nylon, Vicryl, Maxon, Prolene) in rats 7 days after surgery. Interestingly, the study did not reveal any significant differences in tissue reactions. This finding suggests that surgical trauma may have a greater influence on reactivity than the suture material.

The amount of suture placed in a wound, particularly with respect to the knot volume, affects inflammation. The suture size contributes more to knot volume than the number of throws. The volume of square knots is less than that of sliding knots, and knots of monofilament sutures are smaller than those of multifilament sutures.

Allergic reactions to suture materials are rare and have been specifically associated with chromic gut. Chromic acids may provoke a reaction in individuals who are sensitive to chromate.

For excellent patient education resources, visit eMedicine's Procedures Center. Also, see eMedicine's patient education article Suture Care.



Sutures are classified as absorbable or nonabsorbable, natural or synthetic, and multifilament or monofilament. Numerous companies manufacture sutures; however, Ethicon, Syneture (United States Surgical/Davis and Geck suture division of Tyco Healthcare), and Look (Surgical Specialties Corporation) manufacture most of the sutures used in dermatologic surgery. Most sutures are available in standard 18- and 27-inch lengths. Several manufacturers (eg, Delasco, Look) provide sutures in 8-, 9-, and 10-inch lengths. These shorter sutures are less expensive and used primarily for biopsy wounds or small wound closures.

Absorbable suture materials

Absorbable sutures are defined by the loss of most of their tensile strength within 60 days after placement. They are used primarily as buried sutures to close the dermis and subcutaneous tissue and reduce wound tension. The only natural absorbable suture available is surgical gut or catgut. Synthetic multifilamentous materials include polyglycolic acid (Dexon; Syneture) and Polyglactin 910 (Vicryl; Ethicon). Monofilamentous forms include polydioxanone (PDS; Ethicon), polytrimethylene carbonate (Maxon; Syneture), poliglecaprone (Monocryl; Ethicon), Glycomer 631 (Biosyn; Syneture), and Polyglytone 6211 (Caprosyn; Syneture).

Table 1. Characteristics of Absorbable Sutures
PropertyGutPolyglycolic AcidPolyglactinPolydioxanonePolytrimethylene CarbonatePoliglecaprone
HandlingFairFair-goodGoodPoorGoodExcellent
Knot securityPoorFair-goodFairPoorGoodGood
Tensile strength
Low

Proteolysis by 60-90 d,
unpredictable
High

Hydrolysis by 90-120 d
High

Hydrolysis by 60-90 d
Moderate

Hydrolysis by
180-210 d
High

Hydrolysis by 180-210 d
High

Hydrolysis by  90-120 d
Coefficient of frictionHighHighMediumLowLowLow
MemoryLowLowLowHighLowLow
Tissue reactivityHighLow-moderateLow-moderateLowLowLow
UsesMucosal tissues, vessel ligationBuriedBuried; Vicryl Rapide (Ethicon) in wounds requiring short-term dermal support Buried; in wounds requiring long-term dermal supportBuried; in wounds requiring long-term dermal supportBuried
Other
Low elasticity
Beige, violet, or green
Low elasticity
Clear or violet
Clear or violetClear or greenHigh elasticity
Clear

Surgical gut

Surgical gut or catgut was the first absorbable suture material available. It is a twisted fiber formed from the collagen of the intestines of sheep or cows. Surgical gut is packaged in alcohol to prevent it from drying and breaking. The 3 forms available are plain, chromic, and fast absorbing (Ethicon). Plain gut elicits a marked inflammatory reaction in tissue and maintains its tensile strength for only 7-10 days after implantation. Generally, it is completely absorbed by 70 days; however, loss of strength and absorption vary greatly.

Chromic gut is plain gut treated with chromium salts to slow absorption and decrease tissue reactivity. Its tensile strength is maintained for as long as 10-21 days, and complete absorption does not occur until at least day 90. Plain and chromic gut have limited uses in dermatologic surgery. This material is used in the closure of mucosal surfaces or as ligatures for blood vessels, among other uses.

Fast-absorbing gut is plain gut treated with heat to facilitate more rapid absorption. It was designed for percutaneous suturing and maintains its tensile strength for only 5-7 days. It is completely absorbed within 2-4 weeks. Fast-absorbing gut is useful for the percutaneous closure of facial wounds under low tension and for securing both split- and full-thickness skin grafts.

Polyglycolic acid

Polyglycolic acid (Dexon S; Syneture) was introduced in 1970 as the first synthetic absorbable suture. It has high tensile strength, with a retention of 60% at day 7, 35% at day 14, and only 5% at day 28. Polyglycolic acid is completely hydrolyzed by 90-120 days. This multifilamentous suture is uncoated and braided; therefore, it has good handling and knot security properties. However, its high coefficient of friction results in significant tissue drag. To minimize this drag, a polycaprolate-coated form is available (Dexon II; Syneture). This form slides readily through tissue and is easy to tie; however, 4 throws are recommended to ensure secure knots. The tissue reactivity associated with this material is relatively low, but the multifilament nature may potentiate infection. Polyglycolic acid is available as a clear or green suture.

A fast-absorbing polyglycolic acid suture (Polysyn FA; Surgical Specialties) is available that maintains tensile strength for 7-10 days and is completely hydrolyzed by 42 days. This suture is useful in wounds requiring short-term dermal support.

Polyglactin

Introduced in 1974, polyglactin was the second synthetic absorbable suture material available. Like polyglycolic acid, polyglactin is braided and has similar handling and knot security properties. Polyglactin 910 (Vicryl; Ethicon) is coated with Polyglactin 370, which facilitates knot tying and reduces tissue drag; however, this coating also reduces knot security and may require the use of extra throws.

The initial tensile strength of polyglactin is slightly greater than that of polyglycolic acid and is absorbed more quickly. Polyglactin retains 60% of its tensile strength at day 14 after implantation and only 8% of its original strength at day 28. It is completely hydrolyzed by 60-90 days. Tissue reactivity with polyglactin is low, less than that of polyglycolic acid. Both of these suture materials may be transepidermally eliminated if buried too superficially in the dermis. Although used primarily as a buried suture, polyglactin has been used for percutaneous closures without adverse outcomes and with resultant cost savings. Polyglactin is available as a clear or violet suture.

Two additional Polyglactin 910 sutures have been developed by Ethicon. Vicryl Rapide is Polyglactin 910 that has been ionized with gamma rays to speed its absorption. This product is useful as a buried suture in a wound requiring limited dermal support; it is completely absorbed in 35 days. The newest material is an antibacterial suture (Coated Vicryl Plus Antibacterial; Ethicon). The antibacterial agent used to coat the suture is triclosan.2 Biocompatibility and implantation studies have shown this to be noncytotoxic and nonirritating. Handling and wound healing characteristics and absorption profile are comparable to the original Polyglactin 910 material.

Ford et al3 showed that fewer pediatric patients treated with the antibacterial suture had pain on postoperative day 1 when compared with those treated with the original Polyglactin 910 suture (68% vs 89%). Both in vitro and in vivo studies have demonstrated that antibacterial Polyglactin 910 suture inhibits bacterial colonization with both methicillin-sensitive and methicillin-resistant Staphylococcus aureus and Staphylococcus epidermidis. This suture may be useful in wounds at increased risk of infection.

Polydioxanone

A synthetic monofilament absorbable suture, polydioxanone (PDS; Ethicon) was first available in 1982. Although its initial tensile strength is lower than that of the 2 synthetic multifilament sutures mentioned above, it retains its strength longer. At day 14 after implantation, it has 74% residual initial strength; at day 28, 58%; and at week 6, 41%. Complete hydrolysis occurs by 180-210 days. Polydioxanone is stiff and has poor handling and knot-tying properties. Knot security is likewise low and requires an additional throw. A newer product that has replaced the original product is PDS-II (Ethicon), which has improved handling capabilities.

As a monofilament suture, polydioxanone causes minimal tissue reaction. It is useful in contaminated wounds or wounds in locations at greater risk for infection. Polydioxanone is also useful as a buried suture in wounds that require prolonged dermal support. Dermal support of a wound may reduce the spreading of scars. Elliot and Mahaffey4 demonstrated that a 16% reduction in the spreading of scars can be accomplished by using dermal support for 3 weeks. They further showed a 38% reduction in spread with the use of dermal support for 6 months. Polydioxanone is available as a clear or violet suture. Polydioxanone is more expensive than polyglycolic acid or polyglactin.

Polytrimethylene carbonate

Polytrimethylene carbonate or polyglyconate (Maxon; Syneture) was introduced in 1985 as another synthetic monofilament absorbable suture. It has a high initial tensile strength (greater than that of polydioxanone), and it retains 81% of its strength at day 14, 59% at day 28, and 30% at week 6. It is completely hydrolyzed by 180-210 days. Polytrimethylene carbonate is easier to handle and has greater knot security than polydioxanone, polyglactin, or polyglycolic acid. Its tissue reactivity is comparable to that of polydioxanone, its uses are the same, and its cost are lower. Polytrimethylene carbonate is available as a clear or green suture.

Poliglecaprone

Poliglecaprone (Monocryl; Ethicon) is a synthetic absorbable material introduced in 1993. Poliglecaprone is very pliable despite its monofilament nature, and, as a result, its handling and knot strength are excellent. Among all absorbable monofilament sutures, poliglecaprone has the highest tensile strength; however, only 20-30% of its strength is retained at day 14 after implantation. Complete hydrolysis occurs by 90-120 days. Poliglecaprone is most useful as a buried suture in wounds in which prolonged dermal support is not essential. Similar to other monofilament sutures, poliglecaprone has minimal tissue drag and reactivity. It is available as a clear material. The cost of poliglecaprone is comparable to that of polydioxanone.

An antibacterial form of poliglecaprone is now available (Monocryl Plus Antibacterial; Ethicon). The antibacterial agent is triclosan, which has been shown to inhibit colonization of the suture by methicillin-sensitive and methicillin-resistant S aureus and S epidermidis, Escherichia coli, and Klebsiella pneumoniae, even after direct in vivo challenge with bacteria. The tensile strength and absorption profile are similar to that of untreated poliglecaprone.

Other

A monofilament synthetic polyester, Glycomer 631 (Biosyn; Syneture), is another absorbable suture. Rodeheaver et al5 compared it to Polyglactin 910 and found it to have greater tensile strength 4 weeks after implantation. The handling characteristics and knot security were also superior. Tissue drag and risk of bacterial infection were lower, likely a result of the monofilament construction. Glycomer 631 retains 75% of its tensile strength at day 14 and 40% at day 21. Absorption is complete between 90 and 110 days. This suture is available undyed or in violet.

One of the newest synthetic absorbable sutures is Polyglytone 6211 (Caprosyn; Syneture), which is a rapidly degraded monofilament polyester. Polyglytone 6211 has greater tensile strength, lower tissue reactivity, and improved handling characteristics compared with chromic gut suture. Its greatest advantage, however, is its rapid rate of absorption. It provides secure wound approximation for 10 days, and all tensile strength is lost by day 21. It is completely hydrolyzed in 56 days. Polyglytone 6211 is available undyed or in violet.

Nonabsorbable suture materials

Nonabsorbable sutures are defined by their resistance to degradation by living tissues. They are most useful in percutaneous closures. Surgical steel, silk, cotton, and linen are natural materials. Synthetic  nonabsorbable monofilament sutures are most commonly used in cutaneous procedures and include nylon, polypropylene, and polybutester. Synthetic nonabsorbable multifilament sutures composed of nylon and polyester are used infrequently in dermatologic surgery.

Table 2. Nonabsorbable Suture Characteristics
PropertiesSilkNylon, MonofilamentNylon, MultifilamentPolyesterPolypropylenePolybutester
HandlingExcellentPoorFair-goodGoodPoorGood
Knot securityExcellentPoorFair-goodGoodPoorFair-good
Tensile strengthLowHighHighHighModerateHigh
Coefficient of frictionHighLowHighHighVery lowVery low
MemoryLowHighMediumMediumHighLow
Tissue reactivityHighLowModerateLow-moderateLowLow
UsesMucosal tissues,  conjunctiva or intertriginous zones; to elevate or retract tissuesPercutaneous; buried if prolonged dermal support is neededMinimal use in dermatologic surgeryMinimal use in dermatologic surgeryPercutaneous; buried if prolonged dermal support is needed; running subcuticular closuresPercutaneous; running subcuticular closures
OtherBlackBlack, green, or clear
Green or clearHigh plasticity
Blue or clear
High elasticity
Blue or clear

Silk

Silk was first widely used as a suture material in the 1890s. It is a braided material formed from the protein fibers produced by silkworm larvae. Although silk is considered a nonabsorbable material, it is gradually degraded in tissue over 2 years. Silk has excellent handling and knot-tying properties and is the standard to which all other suture materials are compared. Its knot security is high, tensile strength low, and tissue reactivity high. Suture removal can be difficult and painful because the braided material becomes infiltrated with cells and encrusted with debris while implanted in the skin.

Silk is a soft, pliable suture material that is comfortable for patients and unlikely to tear through even delicate tissues. As a result, it is a good choice for use in mucosal tissues or intertriginous areas. Silk also is useful as a temporary suture to elevate or retract tissues for improved visibility during surgery. It is available in black.

Nylon

Introduced in 1940, nylon was the first synthetic suture available, and it is the most commonly used nonabsorbable material in dermatologic surgery. It is available in both monofilamentous and multifilamentous forms. Nylon has a high tensile strength, and, although it is classified as nonabsorbable, it loses tensile strength when buried in tissue. Multifilamentous forms retain no tensile strength after being in tissue for 6 months, whereas monofilamentous forms retain as much as two thirds of their original strength after 11 years. Monofilament nylon is stiff; therefore, handling and tying are difficult and knot security is low. The suture also may cut easily through thin tissue.

Multifilamentous forms have better handling properties but greater tissue reactivity and cost. They are used infrequently in dermatologic surgery. Monofilament nylon is relatively inexpensive and available as black, green, or clear. Although its greatest use is as a percutaneous suture, because of its low tissue reactivity, nylon (clear) can be used as a buried suture in situations in which prolonged dermal support is necessary.

Polyester

Polyester is a braided synthetic suture with uses similar to those of multifilamentous nylon. It has high tensile strength, with good handling, good knot security, and relatively low tissue reactivity. Polyester is available in a coated form that decreases tissue drag. It is a pliable, soft suture similar to silk and can be used in mucosal tissues or intertriginous areas; however, polyester is costly and is infrequently used in dermatologic surgery.

Polypropylene

Polypropylene is a monofilament synthetic suture that was introduced in 1962. Its tensile strength is lower than that of the other synthetic nonabsorbable sutures. Its handling, tying, and knot security are poor as a result of its stiff nature and high memory. An additional throw is needed for adequate knot security. A method to improve security is the use of thermocautery to fuse the knots or transform the ends into small beads. Tissue reactivity is extremely low for polypropylene, and, unlike nylon, gradual absorption does not occur if it is buried in tissue. As a result, polypropylene is an excellent choice for a buried suture for long-term dermal support.

Polypropylene has a low coefficient of friction and easily slides through tissue; this characteristic makes it the suture of choice for a running subcuticular closure. It is a plastic suture that accommodates tissue swelling; therefore, the likelihood of it cutting through the tissue and causing crosshatching is less than that of other materials. However, as swelling of a wound resolves, suture remains loose, and this looseness may affect wound approximation. The plasticity of polypropylene may actually improve knot security in some cases; the knot is tightened as the suture is stretched during wound swelling. Polypropylene is more expensive than nylon and is available as a clear or blue suture.

Polybutester

The newest monofilament nonabsorbable synthetic suture is polybutester (Novafil; Syneture). This suture combines many of the desirable characteristics of polypropylene and polyester. Polybutester has a high tensile strength with good handling qualities. Its memory is lower than that of polypropylene, and therefore, its knots are more secure. Polybutester is not a plastic suture, but it has unique elastic properties that allow it to optimally respond to wound edema. Like polypropylene, polybutester has a low coefficient of friction and is an excellent choice for a running subcuticular closure. Polybutester is available as a clear or blue suture. Its cost is comparable to that of polypropylene.



Needles are necessary for the placement of suture material in a wound. Needles should be made of high-quality stainless steel, sharp enough to penetrate tissue with minimal trauma, rigid enough to resist bending, and malleable enough to bend before breaking. The sharpness of the needle is determined by the method of sharpening. Hand-honed or electrohoned needles are sharper than those processed by machine grinding. Sharper needles pass more easily through tissue and create less trauma to the wound. They are ideal for fine cosmetic work. The cost of needles is based on the quality of the stainless steel and their sharpness. In choosing a needle for cutaneous surgery, the following should be considered: tissue thickness, tissue type, location, need for cosmesis, suture size, and cost.

Needle structure

The structure of a needle includes 3 standard portions: the shank or eye, the body, and the point. The shank is the portion of the needle that is attached to the suture. Swaged needles are preferred in dermatologic surgery. They have a hollow shank into which the manufacturer has secured the suture. This portion, the swage, is both the thickest and weakest part of the needle.

The body is the portion of the needle between the proximal end of the point and the contour change at the beginning of the swage. In cutaneous surgery, the body is usually curved. The curvature is based on a fraction of the arc of a circle; 3/8-circle and 1/2-circle curvatures are most often used. The choice of a particular needle length, width, and curvature depends on the size and depth of the wound and the thickness and type of tissue to be sutured. For example, in deep, narrow, or confined spaces, 1/2-circle needles are easier to handle. The cross-sectional shape of the body can be round, triangular, oval, rectangular, or trapezoidal. The body may be ribbed to improve the holder's grip on the needle.

The main functions of the body are to interact with the needle holder and to transmit penetrating force to the point. The needle holder should be placed approximately one third the distance from the shank to the tip of the point. Holding the needle over the shank may cause it to bend or break.

The point is the portion of the needle from the tip to the end of the taper at the maximum diameter of the body. Three types of needle points are available: round, conventional-cutting, and reverse-cutting. A round point is tapered and has no cutting edges; it passes through tissue by stretching it. This needle is used for suturing soft, elastic tissues (eg, fascia, muscle) but not skin.

Conventional-cutting and reverse-cutting needles are most commonly used in dermatologic surgery. They pass through tissue by cutting a path rather than stretching the tissue. Both have triangular shapes and 3 cutting edges. Two of the cutting edges are on opposing sides of the needle. The conventional-cutting needle has its third cutting edge on the inner curvature of the needle and has a tendency to cut through tissue if upward pressure is exerted. The reverse-cutting needle has its third cutting edge on the outer curvature. This needle is less likely to tear through tissue during suturing and is more often used for cutaneous procedures. A new quad-sided needle with diamond geometry (Dermaglide; Look) is available, and it is designed to minimize tissue trauma by retaining its sharpness longer.

Needle selection

The selection of needles is confusing because manufacturers use varying designations for similar needles (see Table 3). Ethicon needles, for example, include FS (for skin), PS (for plastic skin), P (for precision point), and PC (for precision cosmetic) types. Syneture needles include P or PC (plastic/cosmetic), C (cutaneous) and DX (DermaX premium x cutting). Look (Surgical Specialties) makes needles comparable to those made by the above manufacturers.

The FS and C types are large, reverse-cutting needles and are the least sharp and expensive. They are useful in the buried and percutaneous closure of wounds on the scalp, trunk, and extremities in which cosmesis is not crucial. The PS needle is similar to the FS needle in size and configuration, but it is sharper and better for cosmetic procedures in areas where the skin is thicker and tougher. The P needles are also reverse cutting, but they are sharper and smaller than the PS needle; therefore, they are an excellent choice for use in thin, delicate tissues when cosmesis is important.

The PC needles are conventional-cutting needles, but they have a long, narrow, and very sharp tip that is less likely to tear through tissue. As a result, they cause very little tissue trauma during suturing. PC needles are ideal for fine, detailed work.

Table 3. Needles
Needle StyleManufacturer
EthiconLookSyneture
11 mm, 3/8 circle, reverse cuttingP-1C1P-10
13 mm, 3/8 circle, reverse cuttingP-3C3P-13
13 mm, 3/8 circle, conventional cuttingPC-1CP1PC-10
19 mm, 3/8 circle, reverse cuttingPS-2PC-31P-12
19 mm, 3/8 circle, reverse cuttingFS-2C6C-13




Staples are formed from high-quality stainless steel and are available in regular and wide sizes. Staples are composed of (1) a cross-member that lays on the surface of the skin perpendicular to the wound, (2) legs that are vertically placed in the skin, and (3) tips that secure the staple parallel to the cross-member. Staples are relatively easy to place and may shorten the closure time by 70-80%. The primary utility of staples is in the closure of wounds under high tension on the trunk, extremities, and scalp. They are also used to secure split-thickness skin grafts. They are not used in delicate tissues or wounds in finely contoured areas, over bony prominences, or in highly mobile areas.

Several studies have been conducted to compare the use of staples and nylon sutures on the trunk, head, and neck; these revealed comparable cosmetic results. Advantages of staples include rapid speed of closure, a decreased risk of infection, improved wound eversion, and minimal tissue reactivity. Disadvantages include the need for a second operator to evert and reapproximate skin edges during staple placement, greater risk of crosshatch marking, and less precise wound approximation. The cost is usually more than that of suture material.

In dermatologic surgery, the staplers used are disposable and loaded with 5-35 staples, depending on the manufacturer. They are lightweight and have handles that are easy to grip and control. The width and height of the staples vary with the manufacturer. Most regular staples are 4-6 mm wide and 3.5-4 mm high. Wide staples for use in thicker skin are 6.5-7.5 mm wide and 4-5 mm high.

For staple placement, the stapler is gently held on the surface of the skin, perpendicular to the wound, and the handle is squeezed, plunging the staple into the skin to form an incomplete rectangle. The depth of penetration is based on the pressure exerted on the stapler against the skin. To disengage the staple, the handle is released. If the stapler has an ejector-spring release, it is lifted vertically off the skin. If not, the stapler must be moved anteriorly or posteriorly.

The correct placement of staples is critical to avoid complications such as tissue strangulation and crosshatch marking. Staples should be inserted at 45° or 60° angles. As a wound swells, a staple placed at an acute angle rotates into a vertical position, leaving a space between the cross-member and the skin surface to accommodate swelling. If placed at a 90° angle, the staple cannot move and is likely to strangulate the tissue during swelling.

Staplers that are used widely in dermatologic surgery include Appose (Autosuture, United States Surgical), Proximate (Ethicon), and Precise (3M). A unique stapler is the Multifire Premium stapler (Autosuture), which is disposable but can be reloaded for continued use on the same patient.

Staples are removed painlessly by using a specialized set of extractors.



Tapes are strips of microporous nonocclusive material (eg, paper, plastic, rayon fabric) backed by a thin film of acrylic polymer adhesive. They are useful as an adjunct to or a substitute for other wound closure materials. Although they are used most often to reinforce a wound after the removal of sutures or staples, they can also be used alone for wounds that are small, nonexudative, and under minimal tension.

The advantages of tapes include ease of use, comfort to the patient, and avoidance of tissue strangulation, infection, and crosshatch marks. Follow-up visits for removal are not necessary. Allergic reaction to the adhesive is uncommon. Disadvantages include limited wound eversion, imprecise wound-edge approximation, and inconsistent adhesion. Tapes have little usefulness in hairy or highly mobile areas. Moisture, soap, and wound exudate decrease the duration of tape adhesion.

To maximize adhesion, the use of a liquid adhesive is essential. Mastisol (Ferndale Laboratories) and tincture of benzoin are available for this purpose; Mastisol has superior adhesive strength. After the wound is grossly reapproximated, the area should be degreased with alcohol, acetone, or adhesive remover, and the liquid adhesive should be applied over the wound edges and the entire area to which the tapes will be placed. Once the adhesive has dried to a tacky feel, strips of tapes should be placed perpendicularly across the wound without overlapping one another. Additional tapes should not be used parallel to the wound to reinforce the tape edges because this application decreases adherence. Tapes that are kept dry and clean may remain adherent for up to 1-2 weeks.
 
Zempsky et al compared the cosmetic outcomes of facial lacerations closed with Steri-Strip Skin Closures (3M) or Dermabond tissue adhesive (Ethicon) in 97 patients. At short-term follow-up, more complications were noted in the tissue adhesive group (7) than in the tape group (1). Long-term follow-up revealed no cosmetic difference. Tape closure of simple facial lacerations may be a useful low-cost alternative to tissue adhesives and sutures.

Tapes are available in different widths (eg, one-eighth inch, quarter inch, half inch) and colors (eg, white, clear, flesh toned). Commonly available products include Steri-Strip Skin Closures (3M), Cover Strip II (Beiersdorf), and Proxi-Strip Skin Closures (Ethicon).



Cyanoacrylates for use in surgery have been available in Canada and Europe for 20 years, but products suitable for use in dermatologic surgery in the United States have not become available until recently. Octylcyanoacrylate (Dermabond; Ethicon) and N-butyl-2-cyanoacrylate (Indermil; Syneture) polymerize in an exothermic reaction on contact with fluid to form a 3-dimensional, strong, flexible bond, with uses comparable to those of 5-0 monofilament nylon. Both are available in single-use vials/ampules.

Octylcyanoacrylate and N-butyl-2-cyanoacrylate are useful for the closure of simple lacerations in children and uncooperative patients. They are also useful for the closure of incisions under casts or in cases in which follow-up is difficult. They are not for use in areas that are highly mobile or subject to friction (eg, over joints, hands, feet). Both cyanoacrylates are quick and easy to apply, requiring only one-tenth to one-fourth of the time required for suture placement. They provide an antimicrobial and waterproof coating, but repeated washing removes the adhesive in a few days. The cosmetic outcome is good, and no postoperative visit is required for removal.
 
Despite the fact that their use on the hands is discouraged, Sinha et al6 randomized 50 patients undergoing hand surgery to wound closure with sutures or N-butyl-2-cyanoacrylate. Five wound dehiscences occurred; 3 in the cyanoacrylate group and 2 in the suture group. The study demonstrated that N-butyl-2-cyanoacrylate may be as effective as suturing for cutaneous closure in hand surgery.

The cost per unit of cyanoacrylate is higher than that of comparable sutures; however, Osmond et al7 performed a cost analysis, comparing the use of sutures to tissue adhesives in an emergency department. They accounted for all equipment, health care worker time, and loss of income for follow-up visits and found that the use of the tissue adhesive was overall cost effective.

The use of cyanoacrylates requires that the wound be completely reapproximated before its application. In dermatologic surgery, most wounds require buried sutures prior to the application of the adhesive. If the adhesive seeps into the wound bed, healing is impaired.

Before application, the skin must be defatted with alcohol or acetone. Octylcyanoacrylate is applied in a thin layer over the entire wound and extending 5-10 mm beyond the wound edge. The formation of the bond produces heat that the patient can feel. Once the layer is dried (10-30 seconds), a second layer is applied. Three to 4 layers are necessary. N-butyl-2-cyanoacrylate requires only a single application by droplet that sets in 30 seconds and requires no direct contact with the wound. No additional bandaging is required, and the patient is advised to not perform wound care at home. By 7-14 days, most of the adhesive sloughs with the epidermis, and the remainder may be removed with soap and water or petroleum jelly. A chisel tip applicator is available for the octylcyanoacrylate vial to improve application. An accessory cannula tip can be obtained for use with the N-butyl-2-cyanoacrylate ampule for the same reason.

In addition to their indication for use as a surgical adhesive, both cyanoacrylates are approved by the US Food and Drug Administration for use as a barrier against common bacterial microbes, including certain staphylococci, pseudomonads, and E coli.

Recently, a new over-the-counter formulation of octylcyanoacrylate, Liquid Bandage (Johnson & Johnson), was introduced as an occlusive dressing for minor lacerations and abrasions. Martín-García et al8 studied this as a dressing over sutured facial excisions and found it to be a convenient and economical alternative to daily dressing changes.



The bonding of wounds using laser energy, known as laser welding, has been used on a limited basis as an alternative to traditional wound closure. Compared with previously mentioned closure materials, laser welding is faster, watertight, and avoids a foreign-body reaction. However, collateral thermal injury has prevented its regular clinical use. To minimize this injury, laser soldering was introduced. This process involves the application of a biologic "solder" (eg, bovine serum albumin) prior to temperature-controlled laser welding.

Animal studies performed by Simhon et al9 and Brosh et al10 demonstrated faster reepithelialization, reduced scar width, reduced dermal inflammation, and higher long-term tensile strength in laser-soldered incisions compared with sutured incisions. A clinical study by Kirsch et al11 examined laser soldering for hypospadias repair. Operative times and complication rates were reduced compared with standard suturing. Additional studies are necessary to determine the practical use of laser soldering in dermatologic surgery.



  1. Smit IB, Witte E, Brand R, Trimbos JB. Tissue reaction to suture materials revisited: is there argument to change our views?. Eur Surg Res. 1991;23(5-6):347-54. [Medline].
  2. Barbolt TA. Chemistry and safety of triclosan, and its use as an antimicrobial coating on Coated VICRYL* Plus Antibacterial Suture (coated polyglactin 910 suture with triclosan). Surg Infect (Larchmt). 2002;3 Suppl 1:S45-53. [Medline].
  3. Ford HR, Jones P, Gaines B, Reblock K, Simpkins DL. Intraoperative handling and wound healing: controlled clinical trial comparing coated VICRYL plus antibacterial suture (coated polyglactin 910 suture with triclosan) with coated VICRYL suture (coated polyglactin 910 suture). Surg Infect (Larchmt). 2005;6(3):313-21. [Medline].
  4. Elliot D, Mahaffey PJ. The stretched scar: the benefit of prolonged dermal support. Br J Plast Surg. Jan 1989;42(1):74-8. [Medline].
  5. Rodeheaver GT, Beltran KA, Green CW, Faulkner BC, Stiles BM, Stanimir GW, et al. Biomechanical and clinical performance of a new synthetic monofilament absorbable suture. J Long Term Eff Med Implants. 1996;6(3-4):181-98. [Medline].
  6. Sinha S, Naik M, Wright V, Timmons J, Campbell AC. A single blind, prospective, randomized trial comparing n-butyl 2-cyanoacrylate tissue adhesive (Indermil) and sutures for skin closure in hand surgery. J Hand Surg [Br]. Jun 2001;26(3):264-5. [Medline].
  7. Osmond MH, Klassen TP, Quinn JV. Economic comparison of a tissue adhesive and suturing in the repair of pediatric facial lacerations. J Pediatr. Jun 1995;126(6):892-5. [Medline].
  8. Martín-García RF, Janer AL, Rullán FV. Octyl-2-cyanoacrylate liquid bandage as a wound dressing in facial excisional surgery: results of an uncontrolled pilot study. Dermatol Surg. Jun 2005;31(6):670-3. [Medline].
  9. Simhon D, Brosh T, Halpern M, Ravid A, Vasilyev T, Kariv N, et al. Closure of skin incisions in rabbits by laser soldering: I: Wound healing pattern. Lasers Surg Med. 2004;35(1):1-11. [Medline].
  10. Brosh T, Simhon D, Halpern M, Ravid A, Vasilyev T, Kariv N, et al. Closure of skin incisions in rabbits by laser soldering II: Tensile strength. Lasers Surg Med. 2004;35(1):12-7. [Medline].
  11. Kirsch AJ, Cooper CS, Gatti J, Scherz HC, Canning DA, Zderic SA, et al. Laser tissue soldering for hypospadias repair: results of a controlled prospective clinical trial. J Urol. Feb 2001;165(2):574-7. [Medline].
  12. Austin PE, Dunn KA, Eily-Cofield K, Brown CK, Wooden WA, Bradfield JF. Subcuticular sutures and the rate of inflammation in noncontaminated wounds. Ann Emerg Med. Mar 1995;25(3):328-30. [Medline].
  13. Bang RL, Mustafa MD. Comparative study of skin wound closure with polybutester (Novafil) and polypropylene. J R Coll Surg Edinb. Aug 1989;34(4):205-7. [Medline].
  14. Bennett RA. Materials for wound closure. In: Fundamentals of Cutaneous Surgery. St. Louis, Mo: Mosby-Year Book; 1988.
  15. Bernstein G. Polybutester suture. J Dermatol Surg Oncol. Jun 1988;14(6):615-6. [Medline].
  16. Bezwada RS, Jamiolkowski DD, Lee IY, Agarwal V, Persivale J, Trenka-Benthin S, et al. Monocryl suture, a new ultra-pliable absorbable monofilament suture. Biomaterials. Oct 1995;16(15):1141-8. [Medline].
  17. Bourne RB, Bitar H, Andreae PR, Martin LM, Finlay JB, Marquis F. In-vivo comparison of four absorbable sutures: Vicryl, Dexon Plus, Maxon and PDS. Can J Surg. Jan 1988;31(1):43-5. [Medline].
  18. Chantarasak ND, Milner RH. A comparison of scar quality in wounds closed under tension with PGA (Dexon) and Polydioxanone (PDS). Br J Plast Surg. Nov 1989;42(6):687-91. [Medline].
  19. Clark RA. Cutaneous tissue repair: basic biologic considerations. I. J Am Acad Dermatol. Nov 1985;13(5 Pt 1):701-25. [Medline].
  20. Craig PH, Williams JA, Davis KW, Magoun AD, Levy AJ, Bogdansky S, et al. A biologic comparison of polyglactin 910 and polyglycolic acid synthetic absorbable sutures. Surg Gynecol Obstet. Jul 1975;141(1):1-10. [Medline].
  21. dos Santos LR, Freitas CA, Hojaij FC, Araújo Filho VJ, Cernea CR, Brandăo LG, et al. Prospective study using skin staplers in head and neck surgery. Am J Surg. Nov 1995;170(5):451-2. [Medline].
  22. Edlich RF, Becker DG, Thacker JG, Rodeheaver GT. Scientific basis for selecting staple and tape skin closures. Clin Plast Surg. Jul 1990;17(3):571-8. [Medline].
  23. Falanga V, Zitelli JA, Eaglstein WH. Wound healing. J Am Acad Dermatol. Sep 1988;19(3):559-63. [Medline].
  24. Ford HR, Jones P, Gaines B, Reblock K, Simpkins DL. Intraoperative handling and wound healing: controlled clinical trial comparing coated VICRYL plus antibacterial suture (coated polyglactin 910 suture with triclosan) with coated VICRYL suture (coated polyglactin 910 suture). Surg Infect (Larchmt). 2005;6(3):313-21. [Medline].
  25. Fosko SW, Heap D. Surgical pearl: an economical means of skin closure with absorbable suture. J Am Acad Dermatol. Aug 1998;39(2 Pt 1):248-50. [Medline].
  26. Gabel EA, Jimenez GP, Eaglstein WH, Kerdel FA, Falanga V. Performance comparison of nylon and an absorbable suture material (Polyglactin 910) in the closure of punch biopsy sites. Dermatol Surg. Aug 2000;26(8):750-2; discussion 752-3. [Medline].
  27. Hastings JC, Winkle WV, Barker E, Hines D, Nichols W. Effect of suture materials on healing wounds of the stomach and colon. Surg Gynecol Obstet. May 1975;140(5):701-7. [Medline].
  28. Johnson A, Rodeheaver GT, Durand LS, Edgerton MT, Edlich RF. Automatic disposable stapling devices for wound closure. Ann Emerg Med. Dec 1981;10(12):631-5. [Medline].
  29. Katz AR, Mukherjee DP, Kaganov AL, Gordon S. A new synthetic monofilament absorbable suture made from polytrimethylene carbonate. Surg Gynecol Obstet. Sep 1985;161(3):213-22. [Medline].
  30. Katz KH, Desciak EB, Maloney ME. The optimal application of surgical adhesive tape strips. Dermatol Surg. Sep 1999;25(9):686-8. [Medline].
  31. LaBagnara J Jr. A review of absorbable suture materials in head & neck surgery and introduction of monocryl: a new absorbable suture. Ear Nose Throat J. Jun 1995;74(6):409-15. [Medline].
  32. Markovchick V. Suture materials and mechanical after care. Emerg Med Clin North Am. Nov 1992;10(4):673-89. [Medline].
  33. Maw JL, Quinn JV, Wells GA, Ducic Y, Odell PF, Lamothe A, et al. A prospective comparison of octylcyanoacrylate tissue adhesive and suture for the closure of head and neck incisions. J Otolaryngol. Feb 1997;26(1):26-30. [Medline].
  34. Meiring L, Cilliers K, Barry R, Nel CJ. A comparison of a disposable skin stapler and nylon sutures for wound closure. S Afr Med J. Sep 4 1982;62(11):371-2. [Medline].
  35. Metz SA, Chegini N, Masterson BJ. In vivo and in vitro degradation of monofilament absorbable sutures, PDS and Maxon. Biomaterials. Jan 1990;11(1):41-5. [Medline].
  36. Mikhail GR, Selak L, Salo S. Reinforcement of surgical adhesive strips. J Dermatol Surg Oncol. Sep 1986;12(9):904-5, 908. [Medline].
  37. Moloney GE. The effect of human tissues on the tensile strength of implanted nylon sutures. Br J Surg. Mar 1961;48:528-30. [Medline].
  38. Morgan JE. Surgical tape. Plast Reconstr Surg. May 1981;67(5):685-91. [Medline].
  39. Moy RL, Kaufman AJ. Clinical comparison of polyglactic acid (Vicryl) and polytrimethylene carbonate (Maxon) suture material. J Dermatol Surg Oncol. Aug 1991;17(8):667-9. [Medline].
  40. Moy RL, Quan MB. An evaluation of wound closure tapes. J Dermatol Surg Oncol. Aug 1990;16(8):721-3. [Medline].
  41. Moy RL, Waldman B, Hein DW. A review of sutures and suturing techniques. J Dermatol Surg Oncol. Sep 1992;18(9):785-95. [Medline].
  42. Nordström RE, Nordström RM. Absorbable versus nonabsorbable sutures to prevent postoperative stretching of wound area. Plast Reconstr Surg. Aug 1986;78(2):186-90. [Medline].
  43. Pepicello J, Yavorek H. Five year experience with tape closure of abdominal wounds. Surg Gynecol Obstet. Oct 1989;169(4):310-4. [Medline].
  44. Pineros-Fernandez A, Drake DB, Rodeheaver PA, Moody DL, Edlich RF, Rodeheaver GT. CAPROSYN*, another major advance in synthetic monofilament absorbable suture. J Long Term Eff Med Implants. 2004;14(5):359-68. [Medline].
  45. Postlethwait RW. Long-term comparative study of nonabsorbable sutures. Ann Surg. Jun 1970;171(6):892-8. [Medline].
  46. Quinn J, Wells G, Sutcliffe T, Jarmuske M, Maw J, Stiell I, et al. A randomized trial comparing octylcyanoacrylate tissue adhesive and sutures in the management of lacerations. JAMA. May 21 1997;277(19):1527-30. [Medline].
  47. Quinn JV, Ramotar K, Osmond MH. The antimicrobial effects of a new tissue adhesive. Acad Emerg Med. 1996;3:536-7.
  48. Ray JA, Doddi N, Regula D, Williams JA, Melveger A. Polydioxanone (PDS), a novel monofilament synthetic absorbable suture. Surg Gynecol Obstet. Oct 1981;153(4):497-507. [Medline].
  49. Reiter D. Methods and materials for wound closure. Otolaryngol Clin North Am. Oct 1995;28(5):1069-80. [Medline].
  50. Rodeheaver GT, Nesbit WS, Edlich RF. Novafil. A dynamic suture for wound closure. Ann Surg. Aug 1986;204(2):193-9. [Medline].
  51. Rodeheaver GT, Powell TA, Thacker JG, Edlich RF. Mechanical performance of monofilament synthetic absorbable sutures. Am J Surg. Nov 1987;154(5):544-7. [Medline].
  52. Rothenburger S, Spangler D, Bhende S, Burkley D. In vitro antimicrobial evaluation of Coated VICRYL* Plus Antibacterial Suture (coated polyglactin 910 with triclosan) using zone of inhibition assays. Surg Infect (Larchmt). 2002;3 Suppl 1:S79-87. [Medline].
  53. Saxena AK, Willital GH. Octylcyanoacrylate tissue adhesive in the repair of pediatric extremity lacerations. Am Surg. May 1999;65(5):470-2. [Medline].
  54. Singer AJ, Clark RA. Cutaneous wound healing. N Engl J Med. Sep 2 1999;341(10):738-46. [Medline].
  55. Snow SN, Dortzbach R, Moyer D. Managing common suturing problems. J Dermatol Surg Oncol. Jun 1991;17(6):502-8. [Medline].
  56. Storch M, Perry LC, Davidson JM, Ward JJ. A 28-day study of the effect of Coated VICRYL* Plus Antibacterial Suture (coated polyglactin 910 suture with triclosan) on wound healing in guinea pig linear incisional skin wounds. Surg Infect (Larchmt). 2002;3 Suppl 1:S89-98. [Medline].
  57. Storch ML, Rothenburger SJ, Jacinto G. Experimental efficacy study of coated VICRYL plus antibacterial suture in guinea pigs challenged with Staphylococcus aureus. Surg Infect (Larchmt). 2004;5(3):281-8. [Medline].
  58. Thacker JG, Rodeheaver GT, Towler MA, Edlich RF. Surgical needle sharpness. Am J Surg. Mar 1989;157(3):334-9. [Medline].
  59. Toriumi DM, O'Grady K, Desai D, Bagal A. Use of octyl-2-cyanoacrylate for skin closure in facial plastic surgery. Plast Reconstr Surg. Nov 1998;102(6):2209-19. [Medline].
  60. Trimbos JB, Brohim R, van Rijssel EJ. Factors relating to the volume of surgical knots. Int J Gynaecol Obstet. Dec 1989;30(4):355-9. [Medline].
  61. van Rijssel EJ, Brand R, Admiraal C, Smit I, Trimbos JB. Tissue reaction and surgical knots: the effect of suture size, knot configuration, and knot volume. Obstet Gynecol. Jul 1989;74(1):64-8. [Medline].
  62. van Rijssel EJ, Trimbos JB, Booster MH. Mechanical performance of square knots and sliding knots in surgery: comparative study. Am J Obstet Gynecol. Jan 1990;162(1):93-7. [Medline].
  63. Zempsky WT, Parrotti D, Grem C, Nichols J. Randomized controlled comparison of cosmetic outcomes of simple facial lacerations closed with Steri Strip Skin Closures or Dermabond tissue adhesive. Pediatr Emerg Care. Aug 2004;20(8):519-24. [Medline].

Materials for Wound Closure excerpt